Selecting the Right Signal Conditioning Series for Your
Application
DOCUMENTATION
Application Notes
•
AN-528: Design Considerations for Designing a 5B Series
System
•
AN-529: 5B Series Signal Conditioning Subsystems in RFI/
EMI Fields
•
AN-530: Narrow or Wide Bandwidth 5B Series Input
Modules: Which is the Right Choice?
•
AN-533: Applying 5B Series Backplanes and Mounting
Cards
•
AN-535: Digital Input/Output Subsystems
Data Sheet
•
5B47: Isolated, Linearized Thermocouple Input Data Sheet
User Guides
•
5B Series Manual
DESIGN RESOURCES
•
5B47 Material Declaration
•
PCN-PDN Information
•
Quality And Reliability
•
Symbols and Footprints
DISCUSSIONS
View all 5B47 EngineerZone Discussions.
SAMPLE AND BUY
Visit the product page to see pricing options.
TOOLS AND SIMULATIONS
•
5B I/O Subsystem Software Tools
TECHNICAL SUPPORT
Submit a technical question or find your regional support
number.
DOCUMENT FEEDBACK
Submit feedback for this data sheet.
This page is dynamically generated by Analog Devices, Inc., and inserted into this data sheet. A dynamic change to the content on this page will not
trigger a change to either the revision number or the content of the product data sheet. This dynamic page may be frequently modified.
5B47
GENERAL DESCRIPTION
The 5B47 is a single-channel signal conditioning module that
interfaces, amplifies and filters J, K, T, E, R, S, and B-type
thermocouple. The module provides an isolated and protected
precision output of 0 to + 5 V.
The 5B47 internally linearizes its input signal to provide an
output that is linear with temperature. Accuracy for each type of
sensor is specified in the 5B47 model table. (See Model 5B37 for
non-linearized conditioning of thermocouple signals.)
Accurate and System-Ready
- Internal cold-junction
compensation largely corrects errors arising from parasitic
thermocouples formed by thermocouple connection to the
input screw terminals, providing an accuracy of +0.25
o
C @
+25
o
C ambient temperature. The module generates a
predictable upscale signal to indicate an open thermocouple; for
a downscale response, connect a 20MΩ, 0.25W resistor across
screw terminals 1 and 3.
The 5B47 protects the computer side from damage due to field-
side over-voltage faults. All models withstand 240V rms at their
input terminals without damage, thereby shielding computer-
side circuitry from field-side over-voltage conditions. Further,
the 5B47 is mix-and-match and hot-swappable with other 5B
Series modules, so can be inserted or removed from any socket
in the same backplane without disrupting system power.
Inside the 5B47 Module
– A chopper-stabilized input amplifier
provides low drift and stable gain. At the amplifier input, a
.
stable, laser-trimmed zero-scale input voltage is subtracted from
the input signal to set the zero-scale value. For user
convenience, the zero can be optionally factory-set to meet
custom needs. This allows suppression of a zero-scale input
value many times larger than the total span for precise
expanded-scale measurements.
Internal multi-pole low-pass filtering with a 4-Hz cutoff (-3 dB)
provides 60 dB of normal-mode rejection (noise on signal) and
enhancement of common-mode rejection (noise on signal
return) at 60 Hz, enabling accurate measurement of small
signals in high electrical noise.
Signal isolation by transformer coupling uses a proprietary
modulation technique for linear, stable and reliable
performance. The differential input circuit on the field side is
fully floating, eliminating the need for any input grounding. A
demodulator on the computer side of the signal transformer
recovers the original signal, which is then filtered and buffered
to provide a low-noise, low-impedance output signal. The
output common must be kept within +3 V of power common.
A series output switch eliminates the need for external
multiplexing in many applications. The switch is turned on by
an active-low enable input. If the switch is to be on at all times,
the enable-input should be grounded to power common as it is
on the 5B01 and 5B08 backplanes.
.
Figure 2
Rev. 0 | Page 2 of 8
5B47
5B47 Models Available
Model
5B47-J-01
5B47-J-02
5B47-J-03
5B47-K-04
5B47-K-05
5B47-T-06
5B47-T-07
5B47-E-08
5B47-R-09
5B47-S-10
5B47-B-11
5B47-Custom
1
Input Type
Type J
Type J
Type J
Type K
Type K
Type T
Type T
Type E
Type R
Type S
Type B
Type J, K, T, E, R, S, B
Input Range
0°C to +760°C (+32°F to +1400°F)
-100°C to +300°C (-148°F to +572°F)
0°C to +500°C (+32°F to +932°F)
0°C to +1000°C (+32°F to +1832°F)
0°C to +500°C (+32°F to +932°F)
-100°C to +400°C (-148°F to +752°F)
0°C to +200°C (+32°F to +392°F)
0°C to +1000°C (+32°F to +1832°F)
+500°C to +1750°C (+932°F to +3182°F)
+500°C to +1750°C (+932°F to +3182°F)
+500°C to +1800°C (+932°F to +3272°F)
*
Output Range
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
0 V to +5 V
*
Accuracy
1
±1.1°C
±0.5°C
±0.6°C
±1.3°C
±0.6°C
±1.4°C
±0.5°C
±1.7°C
±2.5°C
±2.4°C
±5.1°C
The CJC sensor accuracy should be added to the module accuracy for a system accuracy.
* Custom Input/Output ranges are available. Refer to configuration guide.
5B47 Specifications
Description
Input Ranges
Standard Ranges
Custom Ranges
Output Ranges (R
L
> 50 k
Ω)
Accuracy
1
Initial @ +25°C
Input Offset vs. Temperature
2
Output Offset vs. Temperature
Gain vs. Temperature
See Model Table
±1 µV/°C
±20 µV/°C
±0.0025% of Reading/°C
Cold Junction Compensation
Initial Accuracy @ +25°C
3
Accuracy, +5°C to +45°C
Input Bias Current
Input Resistance
Power On
Power Off
Overload
Noise
Input, 0.1 Hz to 10 Hz Bandwidth
0.2 µV rms
5M
Ω
40 k
Ω
40 k
Ω
±0.25°C (±0.75°C, maximum)
±0.5°C (±0.0125°C/°C)
-25 nA
Refer to Model Table
±5 mV to ±500 mV
+1 V to +5 V or 0 V to +5 V
Model 5B47
Rev. 0 | Page 3 of 8
5B47
Output, 100 kHz Bandwidth
Bandwidth, -3 dB
Output Rise Time, 10% to 90% Span
500 µV rms
4 Hz
200 ms
Common-Mode Voltage (CMV)
Input-to-Output, Continuous
Output-to-Power, Continuous
4
Transient
1500 V rms, maximum
±3 V, maximum
ANSI/IEEE C37.90.1-1989
Common Mode Rejection (CMR)
1 k
Ω
Source Imbalance, 50/60 Hz
Normal Mode Rejection (NMR), 50/60 Hz
160 dB
60 dB
Input Protection
Continuous
Transient
Output Resistance
Voltage Output Protection
Output Selection Time
240 V rms, maximum
ANSI/IEEE C37.90.1-1989
50
Ω
Continuous Short to Ground
6 µs @ C
load
= 0 to 2,000 pF
Output Enable Control
Max Logic "0"
Min Logic "1"
Max Logic "1"
Input Current "0"
Open Input Response
Open Input Response Time
Power Supply Voltage
Power Supply Current
Power Supply Sensitivity, RTI
Mechanical Dimensions
+1 V
+2.5 V
+36 V
0.4 mA
Upscale
10 seconds
+5 V ±5%
30 mA
±2 µV/% of V
s
2.275" x 2.375" x 0.595"
(57.8 mm x 59.1 mm x 15.1 mm)
Environmental
Temperature Range
Rated Performance
Operating
Storage
Relative Humidity
RFI Susceptibility
-25°C to +85°C
-40°C to +85°C
-40°C to +85°C
0 to 93% @ +40°C non-condensing
±0.5% Span error @ 400 MHz, 5 Watt, 3 ft
Includes the combined effects of repeatability, hysteresis, and nonlinearity and assumes R
L
> 50 k
Ω
. Loads heavier than 50 k
Ω
will degrade nonlinearity and gain temperature
coefficient.
2
±1 µV/°C is equivalent to 0.02°C/°C for Type J thermocouples; 0.025°C/°C for Type K and T thermocouples; 0.016°C/°C for Type E thermocouples; 0.168°C/°C for Type R and S
thermocouples.
3
When used with the model AC1361 CJC sensor (see Accessories section), which is provided on each channel of 5B Series backplanes and mounting cards.
4
The output common must be kept within ±3 V of power common.
In order to deepen the understanding of 51 single-chip microcomputer instructions for beginners, the process of instruction execution is now described in detail here, hoping to inspire you! The proces...
Electronic Information Network IntroductionMyoelectric signals are the temporal and spatial superposition of motor unit action potentials (MUAPs) in muscles, while surface electromyographic signals ar...
[i=s]This post was last edited by jishuaihu on 2015-1-27 19:44[/i] There is a requirement like this in my work. The waveform I sampled is as shown in the figure below. The waveform is continuous, and ...
[i=s]This post was last edited by zhang1gong on 2022-1-5 22:24[/i]1 Unpacking
On the afternoon of the last day of 2021, I finally received the evaluation board from Silicon Labs: BG22-EK4108A.
I opene...
Recently, the first Digital China Summit was held in Fuzhou, attracting many domestic ICT industry giants to participate in the exhibition. The long queues outside the venue, the noisy and exciting...[Details]
International Solar Photovoltaic Network News: Chinese inverter company Sungrow announced that it will build a 3GW photovoltaic inverter manufacturing plant in Bangalore, the capital of Karnataka, ...[Details]
After searching a lot of information, I finally understood the time base timer. I did not use any library functions but operated the registers directly. The following introduces the systick in STM32...[Details]
stm32cubeMX graphical configuration content STM32CubeMX is part of the original STMCube initiative from st microelectronics, and STM32Cube includes STM32CubeMX. STM32CubeMX is a graphical software co...[Details]
background After the hardware engineer completed the PCB and schematic design, the routine was modified according to the tutorial of Zhengdian Atom to verify the correctness of the hardware. STM32F...[Details]
Since the second half of 2017,
the market share of
quantum dot
displays has dropped significantly, with global sales falling from 3.42 million in 2016 to 2 million in 2017, a 42% decrease; d...[Details]
When designing a product, if the application environment's requirements for power isolation are not considered, the product will be unstable due to improper design, and even high voltage may damage t...[Details]
As many companies listed with
the concept of
touch screen
begin to "transform and upgrade", the official voice of the TP industry is getting smaller and smaller. At the same time, the list o...[Details]
The ARM11 (S3C6410) processor is selected as the hardware platform, the embedded Linux operating system is used as the software platform, the background difference method and the frame difference...[Details]
From the functional classification, it can be divided into: 1. Active control ADAS: ACC/AEB/LKS, etc. 2. Warning ADAS: FCW/LDW/PCW, etc. 3. Other auxiliary ADAS: BSD/ADB/panora...[Details]
Since the 20th century, our data
storage
has evolved from media such as magnetic tapes, floppy disks and CDs to sophisticated semiconductor
storage
chips that can store data in coun...[Details]
Site conditions: 1. AD acquisition, DMA transmission; AD is set to single-channel continuous mode, DMA is set to single-shot mode. The upper threshold of the AD analog watchdog setting is less than t...[Details]
The role of the watchdog: Since the operation of the microcontroller is often interfered by the external electromagnetic field, the program may run away and fall into an infinite loop. The normal ope...[Details]
ADC: 1. The ADC module inside STM32 has three ADC1, ADC2, and ADC3, which are independent of each other, so synchronous sampling can be performed. 2ADC input clock must not exceed 14MHz, it is genera...[Details]