40W AC-DC Reliab le Green Medical Adaptor
GSM40B
series
(
for 5~9V)
EN60601-1
ANSI/AAMI ES60601-1
IEC60601-1
+ +
Applications
Mobile clinical workstation
Oral irrigator
Portable hemodialysis machine
Breath Machine
Medical computer monitor
Features
2 pole AC inlet IEC320-C8
, Class
Ⅱ
power unit
Medical safety approved (2 x MOPP) accroding to
ANSI/AAMI ES60601-1 and IEC/EN60601-1
Extremely low leakage current
No load power consumption<0.1W
Energy efficiency level VI and meet CoC Version 5
(Except 5~9V for Level V)
-30~+70
℃
wide range working temperature
Protections: Short circuit / Overload / Over voltage
LED indicator for power on
Lifetime > 90 K hours
3 years warranty
Description
GSM40B is a highly reliable, 40W desktop style single-output green medical adaptor series. This product
is equipped with a 2-pin (no FG) standard IEC320-C8 power plug, adopting the input range from 80VAC to
264VAC. The entire series supplies different output voltages between 5VDC and 48VDC that can satisfy
the demands for various kinds of medical electrical devices. The circuitry design meets the international
medical standards (2*MOPP), having an ultra low leakage current (<50
μ
A), fitting the medical devices in
direct electrical contact with the patients.
With the efficiency up to 91% and the extremely low no-load power consumption below 0.1W,
GSM40B is
compliant with USA EISA 2007/DoE, Canada NRCan, Australia and New Zealand MEPS, EU ErP,and meet
Code of Conduct (CoC) Version 5. The supreme feature allows the adaptor to
save the energy when it is either
under the operating mode or the standby mode. The entire series utilizes the 94V-0 flame retardant plastic
case, providing the double insulation that effectively prevents electrical shock. GSM40B is approved with
the international medical safety certificates.
Model Encoding
GSM 40 B 05 - P1J
DC plug type
Output voltage
IEC320-C8 AC inlet
Rated wattage
Series name
{
P1J: Standard model,2.1
ψ
x5.5
ψ
x11mm,C+,tuning fork type
Other options available by customer requested (see Page 4~5)
File Name:GSM40B-SPEC 2018-01-12
40W AC-DC Reliab le Green Medical Adaptor
SPECIFICATION
ORDER NO.
SAFETY MODEL NO.
DC VOLTAGE
RATED CURRENT
CURRENT RANGE
RATED POWER (max.)
OUTPUT
GSM40B
GSM40B15
15V
2.67A
0 ~ 2.67A
40W
100mVp-p
GSM40B18
18V
2.22A
0 ~ 2.22A
40W
120mVp-p
GSM40B24
24V
1.67A
0 ~ 1.67A
40W
150mVp-p
series
GSM40B05-P1J GSM40B07-P1J GSM40B09-P1J GSM40B12-P1J GSM40B15-P1J GSM40B18-P1J GSM40B24-P1J GSM40B48-P1J
GSM40B05
Note.2
5V
5A
0 ~ 5A
25W
GSM40B07
7.5V
5.34A
0 ~ 5.34A
40W
80mVp-p
GSM40B09
9V
4.45A
0 ~ 4.45A
40W
100mVp-p
GSM40B12
12V
3.34A
0 ~ 3.34A
40W
100mVp-p
GSM40B48
48V
0.84A
0 ~ 0.84A
40W
150mVp-p
RIPPLE & NOISE (max.)
Note.3
80mVp-p
VOLTAGE TOLERANCE
Note.4
±5.0%
LINE REGULATION
Note.5
±1.0%
LOAD REGULATION
SETUP, RISE TIME
HOLD UP TIME (Typ.)
VOLTAGE RANGE
FREQUENCY RANGE
EFFICIENCY (Typ.)
AC CURRENT (Typ.)
INRUSH CURRENT (Typ.)
LEAKAGE CURRENT(max.)
OVERLOAD
±5.0%
±5.0%
±1.0%
±5.0%
±5.0%
±1.0%
±5.0%
±3.0%
±1.0%
±3.0%
±3.0%
±1.0%
±3.0%
±3.0%
±1.0%
±3.0%
±2.5%
±1.0%
±2.5%
±2.5%
±1.0%
±2.5%
Note.6
1000ms, 30ms / 230VAC
1500ms, 30ms / 115VAC at full load
50ms / 230VAC
24ms / 115VAC at full load
Note.7
80 ~ 264VAC 113 ~ 370VDC
47 ~ 63Hz
81%
1A / 115VAC
85.5%
0.5A / 230VAC
86%
88%
88.5%
89%
90%
91%
INPUT
Cold start 30A / 115VAC
60A / 230VAC
Touch current < 50
μ
A/264VAC
105 ~ 160% rated output power
Protection type : Hiccup mode, recovers automatically after fault condition is removed
5.2 ~ 7,0V
7.8 ~ 10.2V
9.4 ~ 12.2V
12.6 ~ 16.2V 15.7 ~ 20.3V
18.9 ~ 24.3V
Protection type : Shut down o/p voltage, re-power on to recover
-30 ~ +70
℃
(Refer to "Derating Curve")
20% ~ 90% RH non-condensing
-40 ~ +85
℃
, 10 ~ 95% RH non-condensing
PROTECTION
OVER VOLTAGE
WORKING TEMP.
WORKING HUMIDITY
ENVIRONMENT STORAGE TEMP., HUMIDITY
TEMP. COEFFICIENT
VIBRATION
25.2 ~ 32.4V
50.4 ~ 64.8V
±
0.03% /
℃
(0 ~ 50
℃
)
10 ~ 500Hz, 2G 10min./1cycle, period for 60min. each along X, Y, Z axes
OPERATING ALTITUDE
Note.8
3000 meters
IEC60601-1, EN60601-1/ EN60601-1-11, ANSI/AAMI ES60601-1 / ES60601-1-11(3.1 version),
SAFETY STANDARDS
CAN/CSA-C22.2 No. 60601-1:14 - Edition 3
,
EAC TP TC 004
approved
ISOLATION LEVEL
WITHSTAND VOLTAGE
ISOLATION RESISTANCE
Primary-Secondary: 2xMOPP
I/P-O/P:4KVAC
I/P-O/P:100M Ohms / 500VDC / 25
℃
/ 70% RH
Parameter
Standard
EN55011 (CISPR11),
FCC PART 15 / CISPR22,
Conducted
emission
CAN ICES-3(B)/NMB-3(B)
EN55011 (CISPR11),
FCC PART 15 / CISPR22,
Radiated emission
CAN ICES-3(B)/NMB-3(B)
EN61000-3-2
Harmonic current
Voltage flicker
EN55024 , EN60601-1-2, EN61204-3
Parameter
ESD
EN61000-3-3
Standard
EN61000-4-2
EN61000-4-3
EN61000-4-4
EN61000-4-5
EN61000-4-6
EN61000-4-8
EN61000-4-11
Test Level / Note
Class B
Class B
Class A
-----
Test Level / Note
EMC EMISSION
SAFETY &
EMC
(Note 9)
RF field susceptibility
EMC IMMUNITY
Level 4, 15KV air ; Level 4, 8KV contact
Level 3, 10V/m( 80MHz~2.7GHz )
Table 9, 9~28V/m( 385MHz~5.78GHz )
Level 3, 2KV
EFT bursts
Surge susceptibility
Conducted susceptibility
Magnetic field immunity
Voltage dip, interruption
Level 3, 1KV/Line-Line
Level 3, 10V
Level 4, 30A/m
100% dip 1 periods, 30% dip 25 periods,
100% interruptions 250 periods
OTHERS
MTBF
DIMENSION
PACKING
PLUG
CABLE
740K hrs min. MIL-HDBK-217F(25
℃
)
125*50*31.5mm (L*W*H)
0.29Kg; 40pcs/12.6Kg/1.05CUFT
See page 4~5 ; Other type available by customer requested
See page 4~5 ; Other type available by customer requested
CONNECTOR
NOTE
1. All parameters are specified at 230VAC input, rated load, 25
℃
70% RH ambient.
2. DC voltage: The output voltage set at point measure by plug terminal & 50% load.
3. Ripple & noise are measured at 20MHz by using a 12" twisted pair terminated with a 0.1 f & 47 f capacitor.
μ
μ
4. Tolerance: includes set up tolerance, line regulation, load regulation.
5. Line regulation is measured from low line to high line at rated load.
6. Length of set up time is measured at first cold start. Turning ON/OFF the power supply may lead to increase of the set up time.
7.
Derating may be needed under low input voltages. Pleas check the derating curve for more details.
8. The ambient temperature derating of 3.5
℃
/1000m with fanless models and of 5
℃
/1000m with fan models for operating altitude higher than 2000m(6500ft).
9. The power supply is considered as an independent unit, but the final equipment still need to re-confirm that the whole system complies with the
EMC directives.
For guidance on how to perform these EMC tests, please refer to “EMI testing of component power supplies.”
(as available on http://www.meanwell.com)
File Name:GSM40B-SPEC 2018-01-12
40W AC-DC Reliab le Green Medical Adaptor
GSM40B
Pin Assignment
Output
PIN No.
+Vo
1
-Vo
2
3
-Vo
4
+Vo
Pin Assignment
Output
PIN No.
+Vo
1
-Vo
2
3
-Vo
4
+Vo
Pin Assignment
Output
PIN No.
-Vo
1
-Vo
2
3
+Vo
-Vo
4
5
+Vo
Pin Assignment
Output
PIN No.
1
+Vo
series
Min. DIN 4 Pin with Lock (male)
2 3
1 4
KYCON
KPPX
-
4P equivalent
Type No.
R7B
Min. DIN 4 Pin with Lock (female)
Type No.
23
14
KYCON KPJX-CM-4S equivalent
R7BF
DIN 5 Pin (male)
Type No.
3
1
4 2 5
R1B
Stripped and tinned leads
L
(red)
Type No.
1
2
L1
(black)
by customer
Length of Land L1 by request
( MW’s standard length, L: 25 mm, L1: 5 mm)
2
-Vo
Installation Manual
Please refer to : http://www.meanwell.com/manual.html
File Name:GSM40B-SPEC 2018-01-12