ME20
20W Single Output External Power
Medical Series
Features
Meets UL/EN/IEC60601-1-2, 4th ed. for EMC
*
Approved to EN/IEC/UL60601-1, 3rd edition
with isolation levels which satisfy the 2 MOPP
requirements.
Meets DoE Efficiency Level VI Requirements
No load input power
Average Efficiency
Up to 20W of AC-DC Power
Universal Input 90-264Vac Input Range
Desktop and Wall-Plug versions
Meets EN55011/CISPR11, FCC Part 15.109
Class B Conducted & Radiated Emissions, with
6db margin
E-cap life of >10 years
>1,000,000 Hours MTBF
3 Year Warranty
IP22 Rated Enclosure
LPS
Description
A high performance AC to DC external power supply family designed for medical applications. The ME20A Medical Series
low power external AC-DC power supplies are approved to safety EN/IEC/UL60601-1, 3rd edition with isolation levels
which satisfy the 2 MOPP requirements and designed to UL/EN/IEC60601-1-2, 4th edition for EMC
*
. The ME20A Series
models will operate at universal input range of 90 to 264Vac over the wide temperature range of -20°C to +70°C,
delivering full rated output power up to +40°C and applicable output power derating up to 70°C. These models are
available in desktop and wall-plug versions; include an IP22 rating per IEC60529 for the enclosure, and output cable
terminated at a variety of output connectors. These models use only high quality electrolytic capacitors, providing greater
than 10 years life operating at rated output conditions.
*
Consult Factory for Table 9 compliance information.
Model Selection
Model
Number
ME20A0503F01
ME20A0603F01
ME20A0703F01
ME20A0903F01
ME20A1203F01
ME20A1503F01
ME20A1803F01
ME20A2403F01
ME20A4803F01
ME20A0503N01
ME20A0603N01
ME20A0703N01
ME20A0903N01
ME20A1203N01
ME20A1503N01
ME20A1803N01
ME20A2403N01
ME20A4803N01
www.slpower.com
Volts
5.0V
5.9V
7.5V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
5.0V
5.9V
7.5V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
Output
Current
3.00A
2.50A
2.00A
2.00A
1.50A
1.20A
1.00A
0.83A
0.42A
3.00A
2.50A
2.00A
2.00A
1.50A
1.20A
1.00A
0.83A
0.42A
Output
Power
15W
15W
15W
18W
18W
18W
18W
20W
20W
15W
15W
15W
18W
18W
18W
18W
20W
20W
Ripple &
Noise¹
75mV pk-pk
75mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
75mV pk-pk
75mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
Line
Regulation
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
Load
Regulation
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
Output
Connector
Input
Configuration
2.5 x 5.5 x 9.5mm
Straight Barrel Type,
center positive
Class I Desktop,
IEC60320 C14
Receptacle
2.5 x 5.5 x 9.5mm
Straight Barrel Type,
center positive
Class II Desktop,
IEC60320 C8
Receptacle
ME20 Rev 2.8
9-NOV-17
1
ME20
20W Single Output External Power
Medical Series
Model Selection
(continued)
Model
Number
ME20A0503Q01
ME20A0603Q01
ME20A0703Q01
ME20A0903Q01
ME20A1203Q01
ME20A1503Q01
ME20A1803Q01
ME20A2403Q01
ME20A4803Q01
ME20A0503B01
ME20A0603B01
ME20A0703B01
ME20A0903B01
ME20A1203B01
ME20A1503B01
ME20A1803B01
ME20A2403B01
ME20A4803B01
ME20A0503C01
ME20A0603C01
ME20A0703C01
ME20A0903C01
ME20A1203C01
ME20A1503C01
ME20A1803C01
ME20A2403C01
ME20A4803C01
Notes:
Volts
5.0V
5.9V
7.5V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
5.0V
5.9V
7.5V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
5.0V
5.9V
7.5V
9.0V
12.0V
15.0V
18.0V
24.0V
48.0V
Output
Current
3.00A
2.50A
2.00A
2.00A
1.50A
1.20A
1.00A
0.83A
0.42A
3.00A
2.50A
2.00A
2.00A
1.50A
1.20A
1.00A
0.83A
0.42A
3.00A
2.50A
2.00A
2.00A
1.50A
1.20A
1.00A
0.83A
0.42A
Output
Power
15W
15W
15W
18W
18W
18W
18W
20W
20W
15W
15W
15W
18W
18W
18W
18W
20W
20W
15W
15W
15W
18W
18W
18W
18W
20W
20W
Ripple &
Noise¹
75mV pk-pk
75mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
75mV pk-pk
75mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
75mV pk-pk
75mV pk-pk
75mV pk-pk
90mV pk-pk
120mV pk-pk
150mV pk-pk
180mV pk-pk
240mV pk-pk
480mV pk-pk
Line
Regulation
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
±1%
Load
Regulation
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
±5%
Output
Connector
Input
Configuration
2.5 x 5.5 x 9.5mm
Straight Barrel Type,
center positive
Class II Desktop,
IEC60320 C18
Receptacle
2.5 x 5.5 x 9.5mm
Straight Barrel Type,
center positive
Class II Wall-Plug,
Interchangeable
Blades
(North American
Blade included)²
2.5 x 5.5 x 9.5mm
Straight Barrel Type,
center positive
Class II Wall-Plug,
Fixed North American
Blades³
1. Measured at the output connector, with noise probe directly across output and load terminated with 0.1μF ceramic and 10μF low ESR capacitors. For 5V and
6V models, values listed are typical, 100mV pk-pk maximum with 0.1μF ceramic and 47μF low ESR capacitors used at measurement point.
2. Order blade kit KT-1027K for other blades (EU. UK, Australia)
3. For EU fixed blades, replace “C” in the model number with “M”, for UK blades, replace “C” with “G”, for Australia blades, replace “C” with “H”.
4. For Input Class I models: For AC GND connected to output common (-), insert a “B” in the part number where the “A” is located (ME20B0503F01).
5. All specifications are typical at nominal input, full load, at 25°C ambient unless noted.
General Specifications
Parameter
AC Input
Input Current
Inrush Current
Input Fuses
Earth Leakage
Current
(Input to Ground)
Patient Leakage
Current
(Output to Earth)
Efficiency
Specification
100-240Vac, ±10%, 47-63Hz, 1∅
100Vac: 0.5A, 240Vac: 0.2A
264Vac, cold start: will not exceed 40A
F1, F2: 3.15A, 250Vac fuses (line & neutral
lines) provided on all models
<500μA@264Vac, 60Hz, NC
<1mA@264Vac, 60Hz, SFC
<100μA@264Vac, 60Hz, NC
<500µA@264Vac, 60Hz, SFC
>87%, typical
Parameter
Turn On Time
Hold-up Time
Overtemperature
Protection
Overload Protection
Short Circuit
Protection
Overvoltage
Protection
Isolation
Specification
Less than 700mS @115Vac, full load
20mS min., at full Load, 100Vac input
Will shutdown upon an overtemperature
condition, auto-recovery.
130 to 180% of rating, Hiccup Mode
Hiccup Mode, auto recovery.
130 to 150% of output voltage, hiccup mode
Input-Output: 2 MOPP
Input-Ground: 1 MOPP
Output-Ground: 1500Vac
ANSI/AAMI ES60601-1:2005/(R)2012, CSA
CAN/CSA-C22.2 NO, 60601-1:14; IEC60601-
1:2005+CORR.2:2007+AM:2012; EN 60601-
1:2006/A11:2011, EN60601-1: 2006/A12:2014,
EN60601-1:2006/A1:2013,
Output Power
No Load Input
Power
Ripple and Noise
www.slpower.com
15 to 20W continuous – See models chart for
specific voltage model ratings.
<0.1W per DoE Efficiency Level VI
Requirements
See models chart on pg 1.
Safety Standards
Operating
Temperature
Storage Temperature
-20°C to +70°C. See curve for derating
-40°C to +85°C
ME20 Rev 2.8
9-NOV-17
2
ME20
20W Single Output External Power
Medical Series
General Specifications
(continued)
Parameter
Output Voltage
Transient
Response
Regulation
Specification
See models chart on pg 1.
500µs response time for return to within 0.5%
of final value for any 50% load step over the
range of 5% to 100% of rated load,
∆i/∆t<
0.2A/µs. Max. voltage deviation is +/-3.5%.
See models chart on pg 1.
Operating: 0.003g/Hz, 1.5grms overall, 3
axes, 10 min/axis, 1-500Hz.
Non-Oper.: random waveform, 3 minutes per
axis, 3 axes and Sine waveform, Vib.
frequency/acceleration: 10-500Hz/1g, sweep
rate of 1 octave / minutes, Vibration time of 10
sweeps / axes, 3 axes
See outline drawings
Parameter
Altitude
Relative Humidity
Drop Test
Specification
Operating: to 5000m.
Non-operating: -500 to 40,000 ft.
5% to 95%, non-condensing
1.4m from table top to wooden platform, 4
faces.
Operating: Half-sine, 20gpk, 10mS, 3 axes,
6 shocks total
Non-Operating: Half-sine waveform, impact
acceleration of 100G, Pulse duration of 6
mS, Number of shocks: 3 for each of the
three axis
>1.000,000 hours, full load, 110 & 220Vac
input, 25°C amb., per Telcordia 332 Issue 6
(stress method).
>7 year life based on calculations at
115Vac/60Hz & 230Vac/50Hz, ambient 25°C
at 24 hrs per day, 365 days/year, 6 power up
cycles per day.
Vibration
Shock
Dimensions
MTBF
Weight
150g
E-Cap Life
All specifications are typical at nominal input, full load, at 25°C ambient unless noted.
EMI/EMC Compliance
Parameter
Conducted Emissions:
Radiated Emissions:
Common Mode Noise:
Electro-Static Discharge (ESD)
Immunity on Power ports:
Radiated RF EM Fields Susceptibility
Electrical Fast Transients (EFT) /Bursts:
Surges, Line to Line (Diff Mode) and Line
to GND (CMN Mode)
Conducted Disturbances induced by RF
Fields
Rated Power frequency magnetic fields
Specification
EN55011/CISPR22 Class B, FCC Part 15.107, Class B: 6db margin typ, at 115 and 230Vac
EN55022/CISPR22 Class B, FCC Part 15.109, Class B: 3db margin typ, at 115 and 230Vac
High Frequency (100kHz-20MHz): <40mA pk-pk
EN55024/IEC61000-4-2, Level 4: +/- 8kV contact, +/- 15kV air, Criteria A
IEC60601-1-2, 4
th
Edition, Table 4
EN55022/EN61000-4-3, 10V/m, 80MHz-2.7GHz, 80% AM at 1kHz
IEC60601-1-2, 4
th
Edition, Table 4
EN55024/IEC61000-4-4, Level 4, +/-4kV, 100Khz rep rate, 40A, Criteria A
IEC60601-1-2, 4
th
Edition, Table 5
EN55024/IEC61000-4-5, Level 4, +/-2kV DM, +/-4kV CM, Criteria A
Surpasses IEC60601-1-2, 4
th
Edition requirements.
EN55022/IEC61000-4-6, 3.6V/m – Level 4, 0.15 to 80Mhz; and 12V/m) in ISM and amateur
radio bands between 0.15Mhz and 80Mhz, 80% AM at 1KHz
IEC60601-1-2, 4
th
Edition, Table 5
EN55024/IEC1000-4-8, Level 4: 30 A/m, 50/60 Hz
IEC60601-1-2, 4th Edition, Table 4
EN55024/IECEN61000-4-11: --100% dip for 10 mS, at 0, 45, 90, 135, 180, 225,
270 and 315 degrees, 100% dip for 20mS, 0 deg., Criteria A
--100% dip for 5000mS (250/300 cycles), Criteria B
--60% dip for 100mS, Criteria B
--30% dip for 500mS, Criteria A
IEC60601-1-2, 4th Edition, Table 5
EN55011/EN61000-3-2, Class A
EN61000-3-3
Voltage Interruptions, Dips, Sags &
Surges
Harmonic Current Emissions
Flicker Test
All specifications are typical at nominal input, full load, at 25°C ambient unless noted. Consult factory for information regarding testing for or usage under special
environments.
Note: Performance criteria are based are defined as following:
A – Normal performance during and after the test
B – Temporary degradation, self-recoverable
C – Temporary degradation, operator intervention required to recover the operation
D – Permanent damage
www.slpower.com
ME20 Rev 2.8
9-NOV-17
3