For proper and optimal use of the product, please read this
manual thoroughly before using.
Keep this manual for future reference.
INSTRUCTION MANUAL
Ver.2.1
Important Information and Warnings
①Non-corrosive gases should be used as pressure media for PS60.
②The maximum applicable pressure for the PS60-102R/302R at the time of vacuum break is 500kPa.
③Always carry out wiring work with the power off.
④Press the button more than 3 seconds in the Operation Mode, the panel lock function will be completed and disable to the
key operation. Please refer to the following“Panel lock function”and cancel the pane lock function.
⑤For stability, use a regulated direct current power supply.
Surge absorbing devices diodes, varistors, etc.) necessary if inductive loads such as relays and solenoids are connected to
(
are
the same power line as the PS60. Do not wire in parallel to high voltage cables or power lines, or use the same cable ducts
which contain high voltage cables or power lines.
⑥Check fluctuations in power voltage so that the power input cannot exceed the rating. Also please do not give a rapid voltage
fluctuation like intercepting the eenergization immediately after starting and during setting operation. Memory data may
disappear, and whitch results in a defect of operation/
⑦Be careful not to apply force to the display area of the main body during piping.
⑧Use pH neutral detergents to clean the body. Do not use lacquer thinner and other solvents for cleaning.
⑨Do not use pointed objects such as pens to press the setting buttons on the display panel. Doing so may damage the setting
buttons by piercing them.
⑩Do not put a piece of wire or other long thin object from pressure port. Doing so may damage the internal diaphragm to cause
malfunctioning.
⑪Do not use the product in a place where much steams
Power supply unit
Surge absorbing circuit
and/or dust exist or the product may be subjected to
+V
direct water or oil splash.
⑫
【Recommended measures against noise interference】
PS60
AC input
DC output
It is recommended to use noise absorbing components
(line filter, surge absorber, etc.) in the power supply
FG
terminal of the PS60.
Earthing of FG terminal
For more detailed information please ask for the nearest distributor or the following sales center.
Nishi-Shinjuku Kimuraya Bidg. , 7-5-25
Nishi-Shinjuku Shinjuku-ku Tokyo 160-0023 , Japan Phone.:(03) 3364-7055
Specifications
Model
Type
(Pressure reference)
Rated pressure range
Maximum pressure
Break-down pressure
Acceptable media
Power supply
Current consumption
-100∼100kPa
200kPa
500kPa
102R
PS60
302R
Gauge pressure
-100∼300kPa
600kPa
1.0MPa
Non-corrosive gases
12V∼24VDC±10%, ripple P-P 10% or less
30mA maximum
103R
-0.10∼1.00MPa
1.5MPa
2.0MPa
NPN
(Two outputs) (Two outputs)Transistor, open collector
,PNP
Switch rating: 30VDC, 100mA maximum
Residual voltage: 1.2V maximum
(NPN) 2.2V maximum
/
(PNP) 100mA
at
Switch outputs
Hysteresis
Repeatability
Response
Short circuit protection
Pressure indication
Accuracy
Switch status indication
IP protection
Operating temperature
Operating humidity
Operating conditions Vibration resistance
Shock resistance
EMC
Thermal error
Pressure port
Pressure receiving area material
Net weight
Accessories
2
0∼30 count setting (adjustable)
±0.3%FS
5ms maximum
Included
Signed 3 digits, 7-segment-LED indication
(sampling cycle: approx. 4 times per second)
±1%FS±1 digit
Output 1
(P1)
and output 2
(P2) LED
,
(red)
light up when switch outputs are ON.
Meets IP40 of IEC
−10∼50℃
(storage−20∼70℃)
35∼85 %RH
10∼500Hz, amplitude 1.5mm/98.1
m
/
S
, three directions, two hours each
2
490
m
/
S
, three directions, three times each
EMI: EN55011 Group1,ClassB:1998
EMS: EN61326-1:1997/A-1:1998/The permissible variations in display
counts and set value of switch output not exceed±5%FS.
±3%FS
(0∼50℃, reference temp. 25℃)
M5 female screew
Single crystal silicon
Approx. 50g
(included 1.5m cable)
Connector with cable
(1.5m) DIN rail adapter
,
Output Electrical Diagram(Wire colors correspond to I.E.C standards)
NPN
(Two outputs type)
Display
PNP
(Two output type)
2
(brown)
LED
LED×
2
Power supply
DC10.8∼26.4V
Display
LED
LED×
2
Load
Load
Sensor
4
(black)Switch
Output1
3
(White)Switch Output2
1
(biue)
COM
2
(brown)
(black)Switch Output1
4
(White)Switch Output
2
Output2
3
Load
Load
Power supply
DC10.8∼26.4V
Main circuit
Main circuit
Ditails of the front panel
(SW1 L D
)E
P L D
1E
D
OWN button
PS60
P1
P2
(SW2) E
LD
P L D
2E
Sensor
1
(biue)
COM
3 digits LED
Display unit
½P½
MO E button
D
U button
P
1
Error Messages
■ If the following error messages are displayed, follow the procedures in the table:
Error message
Problem
Solution
Overload current.
(Brinking of SW1 or SW2 indicates
Disconnect the power, then check the load condition.
excessive current on SW1 or SW2.)
Pressure detected when adjusting the zero point.
Applied pressure is higher than the maximum value
of the pressure display range.
Press the button and reset the display. Release the applied pressure
in the pressure port (opened to the atmosphere) and adjust the zero point again.
Check the applied pressure.
Functions
Initial LED Display
PS60
P1
P2
½P½
PS60
P1
P2
½P½
All LED flash once.
The Operations Mode(pressule
detection) activated.
is
Non-display mode(Low power mode)
●When you do not operate any buttons for about 10 seconds, the system will automatically select non-display mode and the LED indicator section will go off.
Pressing any key will cause the LED indicator section to come on back again.
(Note 1)
The decimal point shown in the figure on the right blinks during non-display mode.
(Note 2)
Switch outputs and switch LEDs are operable even during non-display mode.
(Note 3)
Error messages will appear during non-display mode.
※ For how to select non-display mode, see the description of the initial setting mode.
PS60
P1
P2
½P½
Conversion factor
●You can select a conversion factor from the options shown in the table on the right.
(Note 1)
Slashed box: No factors options are available due to inappropriate resolution
and the number of digits for display.
※ For how to set the conversion factor, see the description of the initial setting mode.
Number
selected
( k Pa)
(MPa)
Pressure range
102R
−100∼100
─
−75∼75
−1.00∼1.00
−14.5∼14.5
29.5∼0.0
302R
−100∼300
─
−75∼225
−1.00∼3.00
−14.5∼43.5
29.5∼0.0
103R
─
−0.10∼1.00
─
−1.0∼10.0
−14∼145
─
( -”
“ symbol does not display) ( -”
“ symbol does not display)
Switch working mode
●You can select switch working mode from the options shown in the table below.
(Note 1) the Separate Mode, SW1 and SW2 work separately.
In
(Note 2) the Window Comparator Mode, the minimum value for SW1 and SW2 corresponds to Setting 1 and the maximum value to Setting 2.
In
※ For how to set the switch output, see the description of the initial setting mode.
There are four operation modes. There are shown in the diagrams below.
Output
SW1
SW2
Separate Window Comparator Separate Window Comparator
Mode
H
L
B
L
B
A
A
Operation H
○
○
○
○
○
○
○
○
○
○
○
○
○
○
○
○
Setting 1
Maximum: Setting 2
Minimum: Setting 1
Separate Mode
(HI operation)
ON
OFF
-Pr
H
H
P1:SW1
P2:SW2
Window Comparator Mode
(A operation)
ON
OFF
P1
P2
ON
Digital filter
●Two different digital filters 25ms and 250ms) available.
(
are
The digital filters are useful when it is hard to take readings due to too great fluctuations in pressure.
(Note 1)
Any selected digital filter will be reflected on the pressure display and switch action.
※ For how to set the digital filter, see the description of the pressure setting mode.
Number selected
Pr
-Pr
H
H
Pr
(LO operation)
ON
OFF
Pr
H
H
OFF
-Pr
(B operation)
ON
P1:SW1
P2:SW2
P1
P2
OFF
Pr
-Pr
H
H
Setting 2
Maximum: Setting 2
Minimum: Setting 1
P1≦P2 or P1≧P2
P1≦P2−2H
P1: Setting 1, P2: Setting 2, H: Hysteresis
2
Operational Procedures
Press the and buttons simultaneously.
Press the button more than one second.
Initial
Setting
Mode
Non-display setting
(two selections)
Factor setting
(five selections)
Switch working mode setting
(eight combinations)
Operations
Mode
Press the and buttons simultaneously.
Press the button more than one second.
Pressure
Setting
Mode
P1 setting
P2 setting
Hysteresis 0∼30 counts)
(
Filter setting
(F0∼F2)
Initial Setting Mode
This mode is used to set non-display mode, magnification and switch outputs.
Entering Initial Setting Mode
PS60
P1
P2
½P½
Press the and buttons simultaneously in OperationsMode.
After switched to Initial Setting Mode, the third digit will blink.
(The values for the 102R / 302R model and for the103R have been set in the factory.)
Making initial setting
PS60
P1
P2
½P½
Press the button to move to the next digit. The value of the digitmay be set when the LED below
the digit blinks.
The number will change every time the or button is pressed.
Press the button to move to the next digit.
Display mode
Factor
102R/302R
Switch working mode
:
Standard Mode
:
Non-display
Mode
:
Separate Mode
103R
:
Window Comparator Mode
3
Pressure Setting Mode
This mode is used to make Setting 1, Setting 2, hysteresis and digital filter setting.
Entering Pressure Setting Mode
PS60
P1
P2
½P½
Press the and buttons simultaneously in Operations Mode.
After switched to Pressure Setting Mode, P1 LED should be
blinking to indicate the value for Setting 1.
Setting pressure value
Setting 1
(P1)
PS60
P1
P2
The P1 LED should be blinking.
½P½
Use the or button to select a value for Setting 1
(P1)
.
(The values for 102R and for 103R have been set in the factory.)
(Note 1)
The setting can only be made within the allowable display range.
(Note 2)
The setting should meet P1≦ P2 - 2H when Window Comparator Mode is selected.
Setting 2
(P2)
PS60
P1
P2
The P2 LED should be blinking.
½P½
Use the or button to select a value for Setting 2
(P2)
.
(The values for 102R and for 103R have been set in the factory.)
(Note 1)
The setting can only be made within the allowable display range.
(Note 2)
The setting should meet P1≦ P2 - 2H when Window Comparator Mode is selected.
Hysteresis H)
( setting
PS60
P1
P2
The P1 and P2 LEDs should be blinking.
½P½
Use the or button to set the hysteresis H)
( .
(The values for 102R / 302R and for 103R have been set in the factory.)
(Note 1)
The setting should be 30 counts or less.
(Note 2)
The setting should meet P1≦ P2 - 2H when Window Comparator Mode is selected.
Digital filter setting
PS60
P1
P2
½P½
Use the or button to set the digital filter.
: No filter, : 25ms filter, : 250ms filter
(The value has been set in the factory.)
Zero Point Adjustment
Adjust the pressure indication at the time of pressure release in the pressure port to "zero".
Zero-point adjustment
Adjustr the pressure indication to“zero”when the pressure port is released.
・Open the pressure port to the atmospheric pressure first.
PS60
½P½
P1
P2
・Press and buttons simultaneously in the Operation mode.
・When finger is released from each button, blinks twice.