EEWORLDEEWORLDEEWORLD

Part Number

Search

19U030PG9K

Description
GAGE, PEIZORESISTIVE PRESSURE SENSOR, 0-5Psi, 0-50mV, CYLINDRICAL, THROUGH HOLE MOUNT
CategoryThe sensor   
File Size627KB,8 Pages
ManufacturerHoneywell
Websitehttp://www.ssec.honeywell.com/
Download Datasheet View All

19U030PG9K Overview

GAGE, PEIZORESISTIVE PRESSURE SENSOR, 0-5Psi, 0-50mV, CYLINDRICAL, THROUGH HOLE MOUNT

19U030PG9K Preview

19 mm Series
Low Cost, Stainless Steel,
Isolated Pressure Sensors
DESCRIPTION
Honeywell’s stainless steel 19C, 19U, and 19 Vacuum Gage
Series sensors were developed for pressure applications that
involve measurement of hostile media in harsh environments
compatible with 316 stainless steel. The special Vacuum Gage
Series sensors are specifically designed for applications that
can be exposed to vacuum.
The 19C Series are calibrated and temperature compensated.
They are available for use with either a constant voltage or
current source.
FEATURES
Low cost
Rugged, isolated stainless steel package
Small size
Reliable semiconductor technology
Calibrated and temperature compensated
Absolute and gage pressures
Vacuum compatible, isolated sensors
0 psi to 3 psi to, 0 psi to 500 psi
POTENTIAL APPLICATIONS
Industrial controls
Process control systems
Industrial automation and flow control
Pressure calibrators
They feature a variety of pressure connections to allow use in
a wide range of OEM (Original Equipment Manufacturer)
equipment. The 19U Series is uncompensated for applications
that use their own specialized circuit designs.
The 19C and 19U Series sensors are rugged and reliable
transducers for use in a wide variety of pressure sensing
applications where corrosive liquids or gases are monitored.
19 mm Series
PRESSURE SENSOR CHARACTERISTICS (All Devices)
Environmental Specifications
Compensated temperature range
Operating temperature range
Storage temperature range
Vibration
Shock
Life
Insulation resistance
Maximum Ratings
Voltage version “K”
Voltage version “L”
0 °C to 82 °C [32 °F to 180 °F]
-40 °C to 125 °C [-40 °F to 257 °F]
-40 °C to 125 °C [-40 °F to 257 °F]
10 G at 20 Hz to 2000 Hz
100 G for 11 ms
1 million cycles min.
100 mΩ at 50 Vdc
supply voltage V
S
= +15 Vdc
supply current I
S
= +2.0 mA
PRESSURE RANGE SPECIFICATIONS (19C, 19U)
Catalog Listing
19(C,U) 003P G (1…9) (K/L)
19(C,U) 005P G (1…9) (K/L)
19(C,U) 010P G (1…9) (K/L)
19(C,U) 015P (A,G,V) (1…9) (K/L)
19(C,U) 030P (A,G,V) (1…9) (K/L)
19(C,U) 050P (A,G,V) (1…9) (K/L)
19(C,U) 100P (A,G,V) (1…9) (K/L)
19(C,U) 200P (A,G,V) (1…9) (K/L)
19(C,U) 300P (A,G,V) (1…9) (K/L)
19(C,U) 500P (A,G,V) (1…9) (K/L)
Pressure Range
0 psig to 3 psig
0 psig to 5 psig
0 psig to 10 psig
0 psi to 15 psi
0 psi to 30 psi
0 psi to 50 psi
0 psi to 100 psi
0 psi to 200 psi
0 psi to 300 psi
0 psi to 500 psi
(1)
Proof Pressure
9 psig
15 psig
30 psig
45 psi
90 psi
150 psi
300 psi
600 psi
900 psi
1200 psi
(7)
Burst Pressure
15 psig
25 psig
50 psig
75 psi
150 psi
250 psi
500 psi
1000 psi
1500 psi
2400 psi
(8)
19C SERIES PERFORMANCE CHARACTERISTICS
Characteristic
Zero pressure offset
(2)
Full-scale span
(2)
Full-scale span (3 psi and 5 psi “K” units only)
(3)
Pressure non-linearity
(3)
Pressure hysteresis
Repeatability
(4)
Temp. effect on span
(4)
Temp. effect on offset
(4)
Temp. effect on span (0 psi to 3 psi and 0 psi to 5 psi only)
(4)
Temp. effect on offset (0 psi to 3 psi and 0 psi to 5 psi only)
Thermal hysteresis (0 °C to 82 °C [32 °F to 180 °F])
(5)
Long-term stability of offset and span
(6)
Response time
(9)
Common mode voltage (voltage version “K”)
Input impedance (current version “L”)
Input impedance (voltage version “K”)
Output resistance
Minimum
-2
98
48
0.5
2.0
8.0
3.0
Typical
0
100
50
±0.1
±0.015
±0.010
±0.5
±0.5
±0.1
±0.1
0.1
1.3
4.5
25
4.5
Maximum
2
102
52
±0.25
±0.030
±0.030
±1.0
±1.0
±2.0
±2.0
±0.3
±0.3
2.0
8.0
50
6.0
Unit
mV
mV
mV
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
ms
Vdc
2
www.honeywell.com/sensing
Low Cost, Stainless Steel, Pressure Sensors
19U SERIES PERFORMANCE CHARACTERISTICS
(1)
Characteristic
Zero pressure offset
(2)
Full-scale span
(2)
Full-scale span (3 psi and 5 psi only)
(3)
Pressure non-linearity
(3)
Pressure hysteresis
Repeatability
(10)
Temp. coefficient of resistance (0 °C to 82 °C [32 °F to 180 °F])
(10)
Temp. coefficient of offset (0 °C to 82 °C [32 °F to 180 °F])
(10)
Temp. coefficient of span (0 °C to 82 °C [32 °F to 180 °F])
Thermal hysteresis (0 °C to 82 °C [32 °F to 180 °F])
(5)
Long-term stability of offset and span
(6)
Response time
Input resistance
Output resistance
Minimum
-7.5
30
20
2700
360
4.00
4.00
Typical
1.0
40
30
±0.1
±0.015
±0.010
3420
30
720
±0.1
±0.1
±0.1
4.75
4.75
(1)
Maximum
7.5
60
40
±0.25
±0.030
±0.030
4500
1620
±0.3
±0.3
6.0
6.0
Unit
mV/V
mV/V
mV/V
%FSS
%FSS
%FSS
ppm/°C
μV/V/°C
ppm/°C
%FSS
%FSS
ms
19C VACUUM GAGE COMPENSATED SERIES PERFORMANCE CHARACTERISTICS
Characteristic
Zero pressure offset (at 0 psig)
(2)
Full-scale span (0 psig to positive full-scale)
(3)
Pressure non-linearity
(3)
Pressure hysteresis
Repeatability
(4)
Temp. effect on span
(4)
Temp. effect on offset
Thermal hysteresis (0 °C to 82 °C [32 °F to 180 °F])
(5)
Long-term stability of offset and span
(6)
Response time
(9)
Common mode voltage (voltage version “K”)
Input impedance (current version “L”)
Input impedance (voltage version “K”)
Output resistance
Minimum
-2
98
0.5
1.5
8.0
3.0
Typical
0
100
±0.1
±0.015
±0.010
±0.5
±0.5
±0.1
±0.1
0.1
1.3
4.5
25
4.5
(1)
Maximum
2
102
±0.25
±0.030
±0.030
±1.0
±1.0
±0.3
±0.3
2.0
8.0
50
6.0
Unit
mV
mV
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
%FSS
ms
Vdc
19U VACUUM GAGE UNCOMPENSATED SERIES PERFORMANCE CHARACTERISTICS
Characteristic
Zero pressure offset
(2)
Full-scale span (0 psig to positive full-scale)
(3)
Pressure non-linearity
(3)
Pressure hysteresis
Repeatability
(10)
Temp. coefficient of resistance (0 °C to 82 °C [32 °F to 180 °F])
(10)
Temp. coefficient of offset (0 °C to 82 °C [32 °F to 180 °F])
(10)
Temp. coefficient of span (0 °C to 82 °C [32 °F to 180 °F])
Thermal hysteresis (0 °C to 82 °C [32 °F to 180 °F])
(5)
Long-term stability of offset and span
(6)
Response time
Input resistance
Output resistance
Minimum
-7.5
30
2700
360
4.00
4.00
Typical
1.0
40
±0.1
±0.015
±0.010
3420
30
720
±0.1
±0.1
±0.1
4.75
4.75
Maximum
7.5
60
±0.25
±0.030
±0.030
4500
1620
±0.3
±0.3
6.0
6.0
Unit
mV/V
mV/V
%FSS
%FSS
%FSS
ppm/°C
•V/V/°C
V/V/°C
ppm/°C
%FSS
%FSS
ms
Honeywell Sensing and Control
3
19 mm Series
Notes:
1. Reference conditions (unless otherwise noted): TA = 25 °C [77 °F]; 19C Supply VS = 10 Vdc ±0.01 Vdc or
IS = 1.5 mA ±0.0015 mA; 19U Supply VS = 5 Vdc ±0.01 Vdc or I = 1.0 mA ±0.0015 mA.
2. Full-scale span (FSS) is the algebraic difference between the output voltage at full-scale pressure and the output at zero
pressure. FSS is ratiometric to supply voltage.
3. Pressure non-linearity is based on best-fit straight line from zero to the full-scale pressure. Pressure hysteresis is the
maximum output difference at any point within the operation pressure range for increasing and decreasing pressure.
4. Maximum error band of the offset voltage or span over the compensated temperature range, relative to the 25 °C [77 °F]
reading.
5. Long-term stability over a six-month period.
6. Response time for a 0 psi to FSS pressure step change, 10% to 90% rise time.
7. The maximum pressure that can be applied without changing the transducer’s performance or accuracy.
8. The maximum pressure that can be applied to a transducer without rupture of either the sensing element or transducer
9. Common mode voltage as measured from output to ground. For higher levels of common mode voltage, contact factory.
10. The error band resulting from a maximum deviation of a transducer’s output parameter (offset, span, or resistance) as
temperature is varied 25 °C [77 °F] to any other temperature within the specified range (0 °C to 82 °C [32 °F to 180 °F]). This
parameter is not 100% tested and is guaranteed by process design and tested on a sample basis only. Temperature
coefficient of span is evaluated using a 1 mA constant current excitation source.
PHYSICAL DIMENSIONS (For reference only: mm [in].)
Package 1 (see note): Cell Package
Note:
Non-concentricity effects at the diaphragm weld area may cause runout of up to ±0.006 in between the upper and lower portions of the
sensor body. (It is recommended to use a counter bore to mate with this device to allow for this non-concentricity).
Package 2: Flush Mount Package
Package 3: Flush Mount with Flange
4 Honeywell
Sensing and Control
Low Cost, Stainless Steel, Pressure Sensors
Package 4: Male 1/8-27 NPT
Package 5: Male 1/4-18 NPT
Honeywell Sensing and Control
5

Technical ResourceMore

EEWorld
subscription
account

EEWorld
service
account

Automotive
development
circle

Robot
development
community

Index Files: 1709  780  2044  2510  1736  35  16  42  51  43 
Datasheet   0 1 2 3 4 5 6 7 8 9 A B C D E F G H I J K L M N O P Q R S T U V W X Y Z
Room 1530, 15th Floor, Building B, No. 18 Zhongguancun Street, Haidian District, Beijing Telephone: (010) 82350740 Postal Code: 100190
Copyright © 2005-2026 EEWORLD.com.cn, Inc. All rights reserved 京ICP证060456号 京ICP备10001474号-1 电信业务审批[2006]字第258号函 京公网安备 11010802033920号