Instruction Manual
board. The link and 3 of the components fit
under
the 7107 IC. Then attach the IC sockets, switches
and remaining components. Watch the orientation of
the electrolytic capacitors and the two IC's. The two
terminal blocks which are next to each other slide
together in a tongue & groove arrangement. Put
them together before soldering them. We have
supplied resistor RY1 of 10M resistance. Because the
input impedance of the 7107 is so high this will make
sure that the display does not pick up stray charge
and will calibrate correctly when you first turn it on.
Leave this resistor in place except when you use the
input voltage divider as explained below.
The last step in construction is to attach the two
PCB's together using the 27 pin post header
connector. The short, 90
o
pins fit into the top of the
mother board. Solder it into place first. Then fit on
the display board to the long straight pins. Make
sure you get the display board the correct way
around. Look for the V+ and the G pins. These are
marked on both PCB's. On the display board it is
marked on the bottom side under the solder mask.)
Make sure you match them up: V+ to V+, G to G.
Connect +5V to the board and turn it on. Put the
switch in the Normal position. When the TEST pin 37
of the 7107 is connected to V
CC
then the display
should read -1888. You can easily do this by
touching together the two wires you previously
soldered into the two pads at TEST. This will tell you
that the IC and the display are working correctly.
WHAT TO DO IF IT DOES NOT WORK
Poor soldering is the most likely reason. Check all
solder joints carefully under a good light. Next, check
that all components are in their correct position on
the PCB. Thirdly, follow the track with a voltmeter to
check the potential differences at various parts of the
circuit. Check that -5V is going to pin 26 of the 7107.
CIRCUIT DESCRIPTION
As you can see from the schematic on the previous
page most of the connections are between the 7107
and the LED's. The 7107 contains a number of inputs
which can be varied to do different things. Of prime
importance is the reference voltage, VREF, which is
set by the 10K trimpot and R3. These will be
discussed below.
-5V Generation.
With the addition of only two
capacitors the 7660 IC performs the complete supply
voltage conversion from positive to negative for an
input range of 1.5 to 10V, resulting in
complementary output voltages of -1.5V to -10V.
7107.
The heart of the meter is the a/d converter
built into
the 7107. It uses a dual slope conversion technique.
It relies on the charging and discharging of an
integrating capacitor and having a counter count
GSK-61 General
Purpose 3 ½ Digit LED
Panel Meter
The GSK-61 is a general purpose 3 1/2 digit panel
meter (DPM), which uses the 7107 IC and an LED
display for its output. The advantages for using the
7107 are because it is powered by +5V and -5V,
which can accept inputs in the range +4V to -4V,
and 2) when a digital reading of zero is desired for a
V
IN
not equal to zero as in, for example, temperature
and weighing systems with a variable tare, both the
offset voltage and the transducer voltage can be
conveniently input directly to the 7107.
Kit 61 has been designed as a versatile 3 1/2 digit
panel meter with these considerations in mind. Also
in this Kit we have used small, monoblock capacitors
rather than the much bigger Mylar or metalized
types. This saves space as well as making the meter
look better. Some components are under the IC to
save even more space.
Voltage is the most frequently measured electrical
quantity. In DPM’s to measure temperature, current,
wind speed, resistance, lux etc., what is actually
being measured is voltage. After calibrating the
meter for its particular purpose the voltage
measured will give an accurate digital reading of the
analogue quantity being measured.
Two circuit boards have been used. The 4 LED
displays mount on one board which is soldered to the
mother board at right angles. This allows low profile
end viewing.
The kit is constructed on single-sided printed circuit
boards.
ASSEMBLY INSTRUCTIONS
Check the components in the Kit against the
Component listing. Make sure you identify every
component.
Display PCB.
Fit the 4 LED displays to the display
PCB by following the overlay and solder. Do
not
attach
to the mother board yet. That will be the last
thing to do.
Main PCB.
It is easiest to solder the lowest height
components first - the diodes, resistors monoblock
capacitors and the single link. Use some of the leads
cut off from the resistors to attach four wires coming
out of the TEST, REF HI & REF LO pads. These will be
used later.
Note
that R5 R6 R7
stand up
on the
www.globalspecialties.com
when the capacitor voltage is above a set value.
Since the capacitor discharge is linear the counter
reading is proportional to the input voltage. There
are three phases to the process:
Phase 1.
Auto Zero. The autozero capacitor is
charged to the integrators offset voltage. This
voltage is subtracted from the input signal during
phase 2. The integrator thus appears to have zero
offset voltage.
Phase 2.
Signal Integrate. The signal input is
averaged for 1000 clock pulses.
Phase 3.
Reference Integrate. Input low is internally
connected to Common (which may be an offset
voltage.) VREF is averaged back to either zero volts
or the offset voltage over another 1000 clock pulses.
The number of clock pulses counted to return to this
value is a digital measure of V
IN
.
System Timing.
This is determined by the
components connected to pins 38, 39 & 40. Values
are unchanged for all ranges measured. The internal
oscillator runs at 48kHz, or 3 readings per second.
Decimal Point.
A jumper selects the decimal point
position in the displays.
Analog Section.
C1 is the reference capacitor and is
unchanged for all ranges measured. IN LO is tied to
the analogue COMMON pin 32 by the Normal position
of the switch except when an Offset voltage is input.
See below.
The integration capacitor C5 is suitable for all ranges
measured but the value of the integration resistor R1
should be increased to 470K for a V
REF
of 1V.
Auto-Zero Capacitor.
This is C4 connected to pin
29. It has some influence on the noise of the system
and recovery from overload input. On the 2V scale a
0.047uF capacitor may give better results.
Voltage Measurement.
Since the maximum value
which can be displayed is 1999, voltmeters with full
scale readings of 199.9mV, 1.999V, 19.99V etc. can
be made. The user must decide their own need. Then
a reference voltage and maybe an input attenuator
must be selected.
To use the meter to measure 0 - 199.9mV the
trimpot is adjusted so that the reference voltage
between pins 35 & 36 is 100mV. And to set the
meter for 0 - 1.999V, V
REF
must be set to 1.0V.
Measuring higher voltages and non-standard
voltages will be discussed below.
The relationship between full scale input voltage and
the reference voltage is:
V
IN
= 2 x V
REF
And V
REF
must be in the range 100mV to 1.0V.
Components supplied in this Kit are for V
REF
of
100mV. For a V
REF
of 1V two components should be
changed (as mentioned above) to maintain
sensitivity and recovery from over-voltage. The 10K
trim pot and resistor R3 will allow adjustment for
either value, and for intermediate values when
required (discussed below.) In the following
discussion we will assume that we are using a V
REF
of
100mV.
Calibration is done by attaching a (preferably digital)
multimeter to REF HI and REF LO and adjusting the
trimpot to read 100 mV. This is why wires were put
into these pads during construction. Now the meter
is calibrated to read 0 - 199.9mV.
Voltage Divider.
To measure voltage greater than
0.2V an input voltage divider is required. See the
figure below. This is the purpose of the space for 4
resistors on the main circuit board. The general
relation for full scale sensitivity is now:
V
IN
(full scale) = 2V
REF
x R
Y
/ ( R
X
+ R
Y
)
For example, a 0 - 20V range (when V
REF
is 0.1V) can
be obtained using a 100:1 voltage divider. This can
be done by making RX = 1M and RY = 10K. The
decimal point jumper is placed at position '2' so a full
scale display of 19.99V is indicated. Similarly, a 0 -
200V range can be obtained with RX = 1M and RY =
1K.
If V
REF
is 1V a similar pattern of voltage divider
resistors can be determined. Remember that if no
input divider is used to put the 10M resistor back in
RY1 across the input leads.
Input an Offset Voltage.
A major advantage of a
7107-based meter over a 7106-based meter is that
an offset voltage may be read directly by the 7107.
The same function with a 7106 meter requires
additional electronics to be built. To use the Offset
input, first move the switch the 'Use Offset' position.
The offset voltage is input to pin 32, Offset, and pin
30, In Lo, while the transducer is connected as
normal using Input 2 and Input 3. Refer to the
schematic.
Non-standard Voltage Input.
In many
applications it is required that the output of a
transducer is converted by a scale factor into some
meaningful result. For example, a load cell of a
weighing system may have an output voltage of
0.682V when it has 2.0 Kg weight on it. You want
the meter to read the range 0 - 1.99 Kg directly.
It is an easy matter to adjust V
REF
to 0.341V (half the
output voltage), put the decimal point in the correct
position by moving the jumper and the panel meter
now reads off 0 - 1.99 Kg directly from the display.
Current Measurement.
Currents up to 2A can be
easily measured using the space on the board for a
5W shunt resistor, R. The current is converted into a
voltage by the shunt resistor. The voltage divider
resistors RX and RY (including RY1) are not used.
The principal is shown in Figure 2 below.
If R = 0.1 ohms then 200mV will be developed when
the current through it is 2A. This voltage is applied
to the meter which is set up for the 200mV range.
www.globalspecialties.com
(That is, V
REF
is set to 100mV.) Power dissipation at
the maximum reading is I
2
R which is 0.4W, well
COMPONENTS
Resistors 5%, 1/4W:
47K
R1 yellow violet orange
100K
R2 brown black yellow
15K
R3 brown green orange
1M
R4 brown black green
470R
R5 R6 R7 yellow violet brown
10M
RY1 brown black blue
10K trimpot 103
1N4148 diode
Capacitors:
0.1uF 104 monoblok C1
0.01uF 103 monoblok C2
100pF 101 monoblok C3
0.47uF 474 monoblok C4
0.22uF 224 monoblok C5
100uF 16V miniecap C6
10uF mini ecap C7 C8
LED CA display KW1-561ASA
7107 IC1
7660 IC2
40 pin IC socket
8 pin IC socket
K61 main PCB
K61 display PCB
2 pole terminal block
SPDT pcb-mounted switch
Jumper
3 post dual row header
27 pin single-in-line 90
o
header
1
1
1
1
3
1
1
2
1
1
1
1
1
1
2
4
1
1
1
1
1
1
3
2
1
1
1
Figure 1
within the 5W rating of the resistor. See Table below.
To measure a full scale of 200mA then R should be
1.0 ohms in order to generate 200mV input to the
meter. For a 20 mA meter then R = 10 ohms. Note
that because of wide tolerances in the shunt resistors
it may be necessary to adjust the reference voltage
in order to get the correct reading. Therefore, further
adjustment of V
REF
using a known current may be
required.
WHAT TO LEARN FROM THIS KIT
The Kit shows how much of electronics today can be
contained in a single chip. Commercial low to
medium cost digital multimeters are nothing more
than this kit, some switches and passive components
and a plastic case. The main reason today for the
failure of meters is more likely due to switch contact
and mechanical failure rather than failure of the
electronics itself.
See our full line of kits at www.globalspecialties.com
For more technical information about using the 7106
and 7107 IC’s go to
http://www.intersil.com/data/fn/fn3082.pdf
http://www.intersil.com/data/an/an023.pdf
Figure 2
www.globalspecialties.com
www.globalspecialties.com