Agilent 87075C
75-Ohm Multiport Test Sets
Data Sheet
Agilent 87075C
3 MHz to 1.3 GHz
This document describes the performance
and features of Agilent Technologies 87075C
75-Ohm multiport test sets, both as standalone
units and when combined with Agilent 75-Ohm
8712ET, 8712ES, 8714ET, or 8714ES network
analyzers. The following options are available:
• Option 006 (6 ports)
• Option 012 (12 ports)
For more information about these test sets,
please read the following documents:
• Agilent 87075C brochure,
literature number 5968-4766E
• Agilent 87075C Configuration Guide,
literature number 5968-4768E
Introduction
All specifications and characteristics apply over a
25 °C ± 5 °C range (unless otherwise stated) and
30 minutes after the instrument has been turned on.
Uncorrected (raw) performance:
Indicates instrument
performance without error correction. The uncor-
rected (raw) performance affects the stability of a
calibration.
System performance:
Performance of a complete
multiport test system, which includes an 87075C
test set and a 75-Ohm 8712ET/ES or 8714ET/ES
network analyzer.
Test Set Cal:
The calibration of a multiport test sys-
tem, requiring the connection of known calibration
standards to all of the ports that will be used for
measurements.
SelfCal:
An automated system calibration that uses
calibration standards internal to the test set and
the most recent Test Set Cal data to calibrate the
test system.
Environmental specifications:
Environmental specifi-
cations bound the external conditions for which
the specifications are valid. The environmental
specifications also bound the external conditions
the test set may be subject to without permanently
affecting performance or causing physical damage.
Definitions
Specification:
Warranted performance. Specifications
include guardbands to account for the expected
statistical distribution, measurement uncertainties,
and changes in performance due to environmental
conditions.
Characteristic:
A performance parameter that the
product is expected to meet before it leaves the
factory, but is not verified in the field and is not
covered by the product warranty. A characteristic
includes the same guardbands as a specification.
Typical:
Expected performance of an average unit.
A typical does not include guardbands. It is not
covered by the product warranty.
Nominal:
A general, descriptive term that does not
imply a level of performance. It is not covered by
the product warranty.
Supplemental information:
May include typical, nomi-
nal, or characteristic values.
Calibration:
The process of measuring known stan-
dards from a calibration kit to characterize the sys-
tematic (repeatable) errors of a network analyzer.
Corrected (residual) performance:
Indicates perfor-
mance after error correction (calibration). It is
determined by the quality of calibration standards
and how well “known” they are, plus system
repeatability, stability, and noise.
Table of contents
System performance, two-port calibration
System performance, T/R calibration
System performance, uncorrected
System performance, general
Test set input/output performance
Test set general information
Physical dimensions
Block diagrams
System features
3
5
8
9
12
13
14
15
16
2
System performance
Two-port calibration (user)
Agilent 85036B/E Type-N calibration kit
Specification1 (dB)
Parameter
Directivity
Source match
Load Match
Reflection tracking
Transmission tracking
3 MHz to 1.3 GHz
47
37
47
±0.1
±0.1
Transmission uncertainty (typical)2
Magnitude
Phase
Reflection uncertainty (typical)2
Magnitude
1.
Phase
These specifications apply under the following conditions: measurement uses the
“Fine” (15 Hz) bandwidth, no averaging, and isolation cal “on”; Test Set Cals use
the “Fine” (15 Hz) bandwidth, 16 averages, and assume an isolation calibration
has been performed; the test set must be used with a 75-Ohm 8712ES or 8714ES
network analyzer with firmware revision E.06.00 or later; the test set and the ana-
lyzer must have had their performance verified within the last year; both instru-
ments must have warmed up for at least 30 minutes after turn-on; measurements
are made at an environmental temperature of 25 °C ± 5 °C and within ± 1 °C of
the last valid Test Set Cal.
2. These uncertainty curves only include the effects of the test port(s) within the
measurement path. The effect of the uncorrected match of test ports outside
the measurement path is ignored, and is dependent on the isolation between the
ports of the DUT that are within the measurement path and ports of the DUT that
are outside the measurement path.
3
System performance
Two-port calibration (user)
Agilent 85039B Type-F calibration kit
Specification1,2 (dB)
Parameter
Directivity
Source match
Load Match
Reflection tracking
Transmission tracking
3 MHz to 1.3 GHz
35
28
35
±0.1
±0.2
Transmission uncertainty (typical)
3
Magnitude
Phase
Reflection uncertainty (typical)
3
Magnitude
Phase
1. These specifications apply under the following conditions: measurement uses the
“Fine” (15 Hz) bandwidth, no averaging, and isolation cal “on”; Test Set Cals use
the “Fine” (15 Hz) bandwidth, 16 averages, and assume an isolation calibration
has been performed; the test set must be used with a 75-Ohm 8712ES or 8714ES
network analyzer with firmware revision E.06.00 or later; the test set and the
analyzer must have had their performance verified within the last year; both
instruments must have warmed up for at least 30 minutes after turn-on; measure-
ments are made at an environmental temperature of 25 °C ± 5 °C and within
± 1 °C of the last valid Test Set Cal.
2. Measurements made using a DUT with center pins of the Type-F connectors
meeting the 0.77 to 0.86 mm limits.
3. These uncertainty curves only include the effects of the test port(s) within the
measurement path. The effect of the uncorrected match of test ports outside
the measurement path is ignored, and is dependent on the isolation between
the ports of the DUT that are within the measurement path and ports of the
DUT that are outside the measurement path.
4
System performance
Transmission/reflection (T/R) calibration (user)
Agilent 85036B/E Type-N calibration kit
Specification
1
(dB)
Parameter
Directivity
Source match
Load match (reflection calibration)
Load match (transmission calibration)
Reflection tracking
Transmission tracking
3 MHz to 1.3 GHz
40
35
20
15
±0.1
±0.1
Typical
2
(dB)
3 MHz to 1.3 GHz
—
—
25
20
—
—
1. These specifications are valid for reflection (one-port) and transmission
(enhanced-response) Test Set Cals, and apply under the following conditions:
measurement uses the “Fine” (15 Hz) bandwidth, no averaging, and isolation cal
“on”; Test Set Cals use the “Fine” (15 Hz) bandwidth, 16 averages, and assume an
isolation calibration has been performed; the test set must be used with a
75-Ohm 8712ET, 8712ES, 8714ET, or 8714ES network analyzer with firmware revi-
sion E.06.00 or later; the test set and the analyzer must have had their perfor-
mance verified within the last year; both instruments must have warmed up for at
least 30 minutes after turn-on; measurements are made at an environmental
temperature of 25 °C ± 5 °C and within ± 1 °C of the last valid Test Set Cal.
2. Corrected system performance is changed to typical when the current tempera-
ture has drifted beyond ± 1 °C of the last valid Test Set Cal. Typicals are valid only
when the current temperature is within 25 °C ± 5 °C, and within ± 1 °C of the
most recent SelfCal.
5