EEWORLDEEWORLDEEWORLD

Part Number

Search

GPC41-5

Description
41 WATT GLOBAL PERFORMANCE SWITCHERS
File Size194KB,2 Pages
ManufacturerETC2
Download Datasheet Compare View All

GPC41-5 Online Shopping

Suppliers Part Number Price MOQ In stock  
GPC41-5 - - View Buy Now

GPC41-5 Overview

41 WATT GLOBAL PERFORMANCE SWITCHERS

GPC41 Commercial/GPM41 Medical
41 WATT GLOBAL PERFORMANCE SWITCHERS
GLOBAL PERFORMANCE SWITCHERS
Summary:
• Ultra small (2.00” x 4.00” x 1.50”)
• Wide-range ac input 85-264 Vac
• 2-year warranty
• Conducted EMI exceeds FCC Class B and CISPR
22 Class B (Commercial models) and CISPR 11
Class B (Medical models)
• Commercial Approved to UL1950, IEC950,
EN60950 and CSA22.2-234 L3
• Medical Approved to UL2601-1, IEC601-1 and
CSA22.2 No. 601
• RoHS Compliant (G suffix)
marked to LVD
SPECIFICATIONS
Ac Input
85-264 Vac, 47-63 Hz single phase.
Input Current
Maximum input current at 120 Vac, 60 Hz with full rated output load: 1.3 A
Hold-Up Time
20 ms minimum from loss of ac input at full load, nominal line (115 Vac).
Output Power
40 W continuous, 50 W peak. Peak ratings are for 60 s maximum duration,
10% duty cycle. During peak load condition, output regulation may exceed
total regulation limits.
Overload Protection
Fully protected against short circuit and output overload.
Short circuit protection is cycling type power limit.
Recovery after fault is automatic. Factory set to begin power limiting at ap-
proximately 55 W. See output ratings chart for additional notes or conditions.
Overvoltage Protection.
See output ratings chart.
Efficiency
69-77% at full rated load, nominal input voltage, depending
on model.
Turn-on Time
Less than 1 s at 120 Vac, 25°C (inversely proportional to input voltage).
Input Protection
Requires external ac line fusing. A 2.5 A, 250 V rated fuse with I2t > 20 is
recommended.
Inrush Current
Inrush is limited by internal thermistor. Inrush at 240 Vac under cold start
conditions will not exceed 34 A.
Temperature Coefficient
0.03%/°C typical on all outputs.
Environmental
Designed for 0 to 50°C operation at full rated output
power; derate output current and total output power by
2.5% per °C above 50°C. See Environmental and
Packaging Specifications on next page.
Output Noise
0.5% rms, 1% pk-pk, 20 MHz bandwidth, differential mode. Measured with noise
probe directly across output terminals of the power supply.
Transient Response
Main output—500 μs typical response time for return to within 0.5% of final value
for a 50% load step change. ∆i/ ∆t<0.2 A/μs. Maximum voltage deviation is 3.5%.
Startup/ shutdown overshoot less than 3%.
EMI/EMC Compliance
All models include built-in EMI filtering to meet the
following emissions requirements:
EMI SPECIFICATIONS
Conducted Emissions GPC41
Conducted Emissions GPM41
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Surge Susceptibility
Commercial Leakage Current
60 μA single fault; 60 Hz input.
Commercial Safety
Approved to UL1950, CSA22.2 No. 234 Level 3, IEC950 and EN60950. UL file
#E135803 commercial; CSA #LR46516 all models. All dc outputs are SELV under
normal and single fault conditions.
Medical Leakage Current
35 μA , 254 Vac 60 Hz input.
Medical Safety
Approved to UL2601, CSA22.2 No. 601 Level 3, IEC601.
UL file E116994; CSA #LR46516. The output(s) are intended for safety earthed Sig-
nal Output and Intermediate Circuits only. The output(s) are not acceptable for
patient connection without additional isolation. All dc outputs are SELV under
normal and single fault conditions.
COMPLIANCE LEVEL
EN55022 Class B; FCC Class B
EN55011 Class B; FCC Class B
EN61000-4-2, 6 kV contact, 8 kV air
EN61000-4-3, 3 V/meter
EN61000-4-4, 2 kV, 5 kH
EN61000-4-5, 1 kV diff., 2 kV com.
DC Power Supplies
All specifications are typical at nominal input, full load at 25
O
C unless otherwise stated

GPC41-5 Related Products

GPC41-5 GPC41-12 GPC41-15 GPC41-24 GPC41-28 GPM41-5
Description 41 WATT GLOBAL PERFORMANCE SWITCHERS 41 WATT GLOBAL PERFORMANCE SWITCHERS 41 WATT GLOBAL PERFORMANCE SWITCHERS 41 WATT GLOBAL PERFORMANCE SWITCHERS 41 WATT GLOBAL PERFORMANCE SWITCHERS 41 WATT GLOBAL PERFORMANCE SWITCHERS

EEWorld
subscription
account

EEWorld
service
account

Automotive
development
circle

Robot
development
community

Index Files: 2336  1579  352  2820  1686  48  32  8  57  34 
Datasheet   0 1 2 3 4 5 6 7 8 9 A B C D E F G H I J K L M N O P Q R S T U V W X Y Z
Room 1530, 15th Floor, Building B, No. 18 Zhongguancun Street, Haidian District, Beijing Telephone: (010) 82350740 Postal Code: 100190
Copyright © 2005-2026 EEWORLD.com.cn, Inc. All rights reserved 京ICP证060456号 京ICP备10001474号-1 电信业务审批[2006]字第258号函 京公网安备 11010802033920号