Specifications
DISPLAY
Type
Height
Decimal Point
Overrange indication
Polarity
POWER REQUIREMENTS
AC Voltages
DC Voltages
Power Consumption
ACCURACY @25°C
F45 DC Process/Voltage
F35 DC Process/Voltage
F45 DC Current
F35 DC Current
F45 AC Voltage/Current
F35 AC Voltage/Current
ENVIRONMENTAL
Operating Temperature
Storage Temperature
Relative Humidity
Warm-up Time
NOISE REJECTION
NMRR
CMRR
A TO D CONVERSION
Technique
Rate
MECHANICAL
Bezel
Depth
Panel Cutout
Case Material
Weight
Temperature Coefficient
F35 AC/TRMS (Voltage/Current)
F35 DC (Voltage/Current/Process)
F45 AC (Voltage/Current)
F45 DC Current
F45 DC (Voltage/Process)
7-segment, red LED
0.56" (14.2mm)
user-programmable,
internally or on the terminal block
most significant digit = "1"; other digits
blank
Automatic, with "-" indication,
"+" indication implied
120 or 220VAC, ±10% 50/60Hz
9-32DCV, ±1%
F35: 3VA, F45: 2VA
±0.02% of reading ± 1 count
±0.1% of reading ± 1 count
±0.05% of reading ±1 count
2A ± 5 counts 5A ± 5 counts
±0.1% of reading ± 1 count
2A ± 5 counts, 5A ± 5 counts
±0.5% of reading ± 35 counts
±1% of reading ± 5 counts (45Hz-1KHz)
0 to 55°C
-10 to 60°C
0 to 85% non-condensing
Less than 15 minutes
F35=50dB, 50/60Hz; F45=60dB, 50/60Hz
(w/1K unbalanced @ 60Hz) 90dB min.
Dual slope integration
F35=3 samples per second, nominal;
F45=2.5 samples per second, nominal
3.78" x 1.89" x .51"
2.94"
3.62" X 1.77"
94V-1, UL rated Noryl®
9.0oz (255.1g)
(±0.1% ± 0.5 count)/°C
(±0.01% ± 0.05 count)/°C
(±0.05% ± 0.5 count)/°C
(±0.01% ± 0.1 count)/°C
(±0.05% ± 0.1 count)/°C
A
7
• 1/8 DIN Indicator
• 3-1/2 or 4-1/2 digit bright red LED display
• Front panel pops off for easy decimal
point setting and display scaling
• Only 3.12” (79mm) required behind panel
• Optional excitation output
• NEMA 4X enclosure optional
• Din Rail Adapter available
(page C1)
Mounting Requirements
The Falcon series 1/8 DIN indicators require a panel cutout of
1.77" (45mm) high by 3.62" (92mm) wide. To install the Falcon
into a panel cutout, remove the clips from the side of the meter.
Slide the meter through your panel cutout, then slide the mount-
ing clips back on the meter. Press evenly to ensure a proper fit.
Tighten screws.
Dimensions
4.04”
3.59”
92 mm
3.78
.51
2.94
1.89
1.74
2.28
more >>
A
8
Specifications for F35 Frequency Meters
DISPLAY
Type
Height
Overrange indication
POWER REQUIREMENTS
AC Voltages
Power Consumption
ACCURACY @25°C
200 Hz:
2 KHz:
INPUT LEVEL
7-segment, red LED
0.56" (14.2mm)
Most significant digit = "1"; other digits
blank
120 or 220VAC, ±10% 50/60Hz
2.5VA min./4VA max.
±0.2% of input ±0.2 Hz
±0.2% of input ± 2Hz
500mV to 250V RMS at 1.0M imped-
ance OR 5V to 24V Square Wave (DC
offset 2V maximum)
200Hz = 0.1Hz
2kHz =1Hz
0 to 55°C
-10 to 60°C
0 to 85% non-condensing @ 40°C
(±0.05% of input ± 0.5 count)/°C
Less than 15 minutes
Frequency-to-voltage
3 samples per second, nominal
Specifications for F45 Temperature Meters
DISPLAY
Type
Height
Decimal Point
Overrange indication
Polarity
POWER REQUIREMENTS
AC Voltages
DC Voltages
Power Consumption
ENVIRONMENTAL
Operating Temperature
Storage Temperature
Relative Humidity
Warm-up Time
INPUTS
Thermocouple
RTD
Millivolt
Input Impedance
Conversion Rate
Open Thermocouple Detection
Temperature
Sensor Type
E
J
K
T
RTD Pt 100
R, S
mV
mV
Temperature Range
-200 to 1000°C
-328 to 1832°F
-200 to 1200°C
-328 to 2192°F
-200 to 1372°C
-328 to 2501°F
-200 to 400°C
-328 to 752°F
-200 to 850°C
-328 to 1562°F
-50 to 1768°C
-58 to 3214°F
Voltage Range
±84.00mV
Voltage Range
-19.999mV
to +84.000mV
±.02%
of rdg
±
1 count
±.02%
of rdg
±
1 count
1 Degree
Automatic
0.01 mV
0.001mV
±0.1%
of rdg
± 1°C
±0.1%
of rdg
± 1.8°F
0.1 or 1 Degree
User Selectable
Accuracy
Resolution
7-segment, red LED
0.56" (14.2mm)
Jumper-selectable 2-position (correspon-
ding to resolution desired)
Most significant digit = "1"; other digits blank
Automatic, with "-" indication,
"+" indication implied
120 or 220 ACV, ±10% 50/60Hz
9-32 DCV, ±1%
3VA
0 to 55°C
-10 to 60°C
0 to 85% non-condensing
Less than 20 minutes
J, K, E, T, R, and S
Platinum 100 (.00385 alpha), 2, 3 or 4 wire
±84mV reading of uncompensated mV
7M (typical)
2-1/2 times per second
-1 on display, -40nA bias on thermocouple
Resolution:
ENVIRONMENTAL
Operating Temperature
Storage Temperature
Relative Humidity
Temperature Coefficient
Warm-up Time
CONVERSION
Technique
Rate
Inputs
DC Voltage
Range
200mV
2V
20V
200V
Max Input
Resolution Input Impedance Resolution Input Impedance (Unfused)
100uV
1mV
10mV
100mV
>100MEG
>10MEG
>10MEG
>9.9MEG
10uV
100uV
1mV
10mV
>100MEG
>10MEG
>9.9MEG
>9.8MEG
100V
250V
250V
250V
F35
F45
DC / AC TRMS Current
Range
F35
200uA
2mA
20mA
200mA
2A
5A
AC TRMS Voltage
Range
200mV
2V
20V
200V
Max Input
Resolution Input Impedance Resolution Input Impedance (Unfused)
100uV
1mV
10mV
100mV
>100MEG
>1MEG
>10MEG
>9.9MEG
10uV
100uV
1mV
10mV
>100MEG
>1MEG
>1MEG
>1MEG
100V
250V
250V
250V
4 to 20mA
1 to 5Vdc
0 to 10Vdc
F35
F45
DC Process
Range
F35
F45
Max Input
Resolution Input Impedance Resolution Input Impedance (Unfused)
10uA
10mV
10mV
NA
>10MEG
>10MEG
1uA
1mV
1mV
NA
>9.9MEG
>9.9MEG
200mV
NA
NA
100nA
1uA
10uA
100uA
1mA
10mA
Resolution
F45
10nA
100nA
1uA
10uA
100uA
1mA
200mV
Voltage Drop
Max Input
(Unfused)
10mA
40mA
100mA
500mA
2.2A
5.2A
A
9
Ordering Information
Falcon Indicators can be con gured by making an entry in each section. Example: F35-1-52-0.
Basic
Unit
Selection
Description
Power
Supply
Function/
Range
Excitation
Output
Selection
Display Scaling
(temp. units only)
Description
Basic Unit
F35
F45
1
2
3
3 1/2 digit LED
4 1/2 digit LED
61
62
63
64
65
66
Function/Range cont’d
200 ACµA TRMS •
2 ACmA TRMS •
20 ACmA TRMS •
200 ACmA TRMS •
2 ACA TRMS •
5 ACA TRMS •
Power Supply
120 ACV
220 ACV
9-32 DCV†
Function/Range
11
12
13
14
200 DCmV
2 DCV
20 DCV
200 DCV
200 DC A
2 DCmA
20 DCmA
200 DCmA
2 DCA
5 DCA
200 ACmV
2 ACV
20 ACV
200 ACV
200 AC A
2 ACmA
20 ACmA
200 AC mA
2 ACA
5 ACA
200 ACmV TRMS •
2 ACV TRMS •
20 ACV TRMS •
200 ACV TRMS •
71
72
73
4-20 DCmA Process
1-5 DCV Process
0-10 DCV Process
21
22
23
24
25
26
31
32
33
34
41
42
43
44
45
46
51
52
53
54
80 (F45)
81 (F45)
82 (F45)
83 (F45)
84 (F45)
85 (F45)
86 (F45)
90 (F45)
91
92
93
94
(F35)
(F35)
(F35)
(F35)
J T/C
K T/C
S T/C
T T/C
E T/C
R T/C
DCmV
RTD Pt100 Ohm
20-199.9 Hz RMS
20-1999 Hz RMS
20-199.9 Hz Sq. Wave
20-1999 Hz Sq. Wave
Excitation Output
(not available with AC, temperature, or frequency inputs)
0
1
2
None
12 DCV @ 25mA max. current
24 DCV @ 25mA max. current
Display Scaling
(Temperature units only)
C
F
°C
°F
† Not available for use with frequency meters
• Only available with F35