Rev C
DCMOP06 SERIES
6 Watt Medical DC/DC Power
Converters
FEATURES
•
2µA Patient Leakage Current
•
Single & Dual Outputs
•
Under Voltage Protection
•
High Efficiency up to 89%
•
2:1 Wide Input Voltage Ranges
•
Built-in EMI Class A Filter
Size:
1.25 x 0.80 x 0.40 inches
(31.8 x 20.3 x 10.2 mm)
Applications:
•
Medical Equipment
•
Telecom/Datacom
•
Industry Control Systems
•
Semiconductor Equipment
•
PV Power Systems
•
IGBT Gate Drivers
•
Low Stand-by Power
Consumption
•
6 Watts Output Power
•
Reinforced Insulation for 300VAC Working Voltage
•
Clearance and Creepage Distance: 6.6mm/2MOOP
•
3000VAC Input to Output 2MOOP Isolation
•
Short Circuit, Over Voltage, and Over Load Protection
•
CE Marked
•
Compliant to RoHS II & REACH
•
ANSI/AAMI ES60601-1, EN60601-1, IEC60601-1 3
rd
Edition,
UL60950-1, EN60950-1, & IEC60950-1 Safety Approvals
•
Optional Remote ON/OFF Control and Trim Pin
DESCRIPTION
The DCMOP06 series of medical DC/DC power converters provides 6 Watts of output power in a 1.25”
x 0.80” x 0.40” DIP package. This series consists of single and dual output models with 2:1 wide input
voltage ranges of 4.5-9VDC, 9-18VDC, 18-36VDC, and 36-75VDC. Some features include high
efficiency up to 89%, 3000VDC I/O (2 MOOP) isolation, and low stand-by power consumption. These
converters are also protected against under voltage, short circuit, over voltage, and over load
conditions. All models are RoHS compliant and have ANSI/AAMI ES60601-1, EN60601-1, IEC60601-1
3
rd
Edition, UL60950-1, EN60950-1, and IEC60950-1 safety approvals. Remote ON/OFF and Trim
functions are also available for this series.
MODEL SELECTION TABLE
Model Number
(1)
DCMOP06-5S33x
DCMOP06-5S05x
DCMOP06-5S12x
DCMOP06-5S15x
DCMOP06-5S24x
DCMOP06-12S33x
DCMOP06-12S05x
DCMOP06-12S12x
DCMOP06-12S15x
DCMOP06-12S24x
DCMOP06-24S33x
DCMOP06-24S05x
DCMOP06-24S12x
DCMOP06-24S15x
DCMOP06-24S24x
DCMOP06-48S33x
DCMOP06-48S05x
DCMOP06-48S12x
DCMOP06-48S15x
DCMOP06-48S24x
Model Number
(1)
DCMOP06-5D05x
DCMOP06-5D12x
DCMOP06-5D15x
DCMOP06-12D05x
DCMOP06-12D12x
DCMOP06-12D15x
DCMOP06-24D05x
DCMOP06-24D12x
DCMOP06-24D15x
DCMOP06-48D05x
DCMOP06-48D12x
DCMOP06-48D15x
Input Voltage
Range
5 VDC
(4.5 - 9 VDC)
Output Voltage
3.3 VDC
5 VDC
12 VDC
15 VDC
24 VDC
3.3 VDC
5 VDC
12 VDC
15 VDC
24 VDC
3.3 VDC
5 VDC
12 VDC
15 VDC
24 VDC
3.3 VDC
5 VDC
12 VDC
15 VDC
24 VDC
Output Voltage
±5 VDC
±12 VDC
±15 VDC
±5 VDC
±12 VDC
±15 VDC
±5 VDC
±12 VDC
±15 VDC
±5 VDC
±12 VDC
±15 VDC
Output
Current
1800mA
1200mA
500mA
400mA
250mA
1800mA
1200mA
500mA
400mA
250mA
1800mA
1200mA
500mA
400mA
250mA
1800mA
1200mA
500mA
400mA
250mA
SINGLE OUTPUT MODELS
Output
Ripple & Noise
30mVp-p
30mVp-p
40mVp-p
40mVp-p
50mVp-p
30mVp-p
30mVp-p
40mVp-p
40mVp-p
50mVp-p
30mVp-p
30mVp-p
40mVp-p
40mVp-p
50mVp-p
30mVp-p
30mVp-p
40mVp-p
40mVp-p
50mVp-p
Output
Ripple & Noise
30mVp-p
40mVp-p
40mVp-p
30mVp-p
40mVp-p
40mVp-p
30mVp-p
40mVp-p
40mVp-p
30mVp-p
40mVp-p
40mVp-p
12 VDC
(9 - 18 VDC)
24 VDC
(18 - 36 VDC)
48 VDC
(36 - 75 VDC)
Input Voltage
Range
5 VDC
(4.5 - 9 VDC)
12 VDC
(9 - 18 VDC)
24 VDC
(18 - 36 VDC)
48 VDC
(36 - 75 VDC)
Output
Current
±600mA
±250mA
±200mA
±600mA
±250mA
±200mA
±600mA
±250mA
±200mA
±600mA
±250mA
±200mA
DUAL OUTPUT MODELS
No Load
Input Current
10mA
10mA
15mA
15mA
20mA
10mA
10mA
10mA
10mA
10mA
6mA
6mA
6mA
6mA
6mA
4mA
4mA
4mA
4mA
4mA
No Load
Input Current
25mA
25mA
25mA
10mA
10mA
10mA
6mA
6mA
6mA
4mA
4mA
4mA
Output Power
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
Output Power
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
6W
Efficiency
81.5%
86%
86%
87.5%
87%
83.5%
86%
89%
88.5%
88.5%
83%
86%
89%
89%
88.5%
82.5%
86.5%
88%
88.5%
88%
Efficiency
84%
86.5%
87.5%
85%
89%
88%
85%
88.5%
88.5%
85%
88%
87%
Maximum
Capacitive Load
2100µF
1500µF
260µF
210µF
75µF
2100µF
1500µF
260µF
210µF
75µF
2100µF
1500µF
260µF
210µF
75µF
2100µF
1500µF
260µF
210µF
75µF
Maximum
Capacitive Load
±860µF
±150µF
±110µF
±860µF
±150µF
±110µF
±860µF
±150µF
±110µF
±860µF
±150µF
±110µF
5/21/2019
Wall Industries, Inc. • Tel: 603-778-2300 • Toll Free: 888-597-9255 •
website:
www.wallindustries.com
• e-mail:
sales@wallindustries.com
Page 1 of 5
Rev C
DCMOP06 SERIES
6 Watt Medical DC/DC Power
Converters
SPECIFICATIONS: DCMOP06 SERIES
All specifications are based on 25°C, Nominal Input Voltage, and Maximum Output Current unless otherwise noted.
We reserve the right to change specifications based on technological advances.
SPECIFICATION
INPUT SPECIFICATIONS
Input Voltage Range
TEST CONDITIONS
5VDC nominal input models
12VDC nominal input models
24VDC nominal input models
48VDC nominal input models
5VDC nominal input models
12VDC nominal input models
24VDC nominal input models
48VDC nominal input models
5VDC nominal input models
12VDC nominal input models
24VDC nominal input models
48VDC nominal input models
5VDC nominal input models
12VDC nominal input models
24VDC nominal input models
48VDC nominal input models
No Load
Referenced to –INPUT pin
Nominal Vin
Nominal Vin
DC/DC ON
DC/DC OFF
Min
4.5
9
18
36
Typ
5
12
24
48
Max
9
18
36
75
4.5
9
18
36
Unit
VDC
Start-Up Voltage
VDC
Shutdown Voltage
Input Surge Voltage (3sec,
max.)
Input Current
Input Filter
Remote ON/OFF Control
(Only for “B” type pin
connection models)
16
25
50
100
See Table
Pi type
Open or 0 ~ 1.2VDC
2.2 ~ 12 VDC
-0.5
2.5
1
4
8
16
33
VDC
VDC
Input Current of CTRL Pin
Remote OFF Input Current
Output Voltage
Voltage Accuracy
Line Regulation
Load Regulation
Cross Regulation
Voltage Adjustability
(Only for “B” type pin
connection models)
OUTPUT SPECIFICATIONS
mA
mA
%
%
%
%
%
%
Low line to high line at full load
No load to full load
Asymmetrical load 25%/100% FL
Single Output Models
Dual Output Models
Output Power
Output Current
Maximum Capacitive Load
Ripple & Noise (20MHz BW)
Transient Response
Recovery Time
Start-Up Time
Temperature Coefficient
Single Output Models
Dual Output Models
Single Output Models
Dual Output Models
Dual Output Models
3.3V, 5V, 12V Output Models
15V, 24V Output Models
±5V, ±12V, ±15V Output Models
-1.0
-0.2
-0.5
-0.2
-1.0
-5.0
-10
-10
-10
Minimum input and constant resistive load
Measured with a 10µF/25V X7R MLCC 3.3V, 5V Output Models
Measured with a 10µF/25V X7R MLCC 12V, 15V Output Models
Measured with a 4.7µF/50V X7R MLCC 24V Output Models
25% load step change
Constant resistive load
Power Up
Remote On/Off
See Table
+1.0
+0.2
+0.5
+0.2
+1.0
+5.0
+10
+20
+10
See Table
See Table
See Table
30
40
50
250
30
30
mVp-p
µs
ms
PROTECTION
-0.02
+0.02
%/°C
Short Circuit Protection
Over Load Protection
% of rated Iout; hiccup mode
Single Output Models
3.3V Output Models
5V Output Models
12V Output Models
15V Outputs Models
24V Output Models
5V Output Models
12V Output Models
15V Output Models
Continuous, automatic recovery
150
3.7
5.6
13.5
18.3
29.1
5.6
13.5
17.0
5
7.0
16
22.0
34.5
7.0
18.2
22.0
%
Over Voltage Protection
Continuous
clamp
VDC
Dual Output Models
5/21/2019
Wall Industries, Inc. • Tel: 603-778-2300 • Toll Free: 888-597-9255 •
website:
www.wallindustries.com
• e-mail:
sales@wallindustries.com
Page 2 of 5
Rev C
DCMOP06 SERIES
6 Watt Medical DC/DC Power
Converters
SPECIFICATIONS: DCMOP06 SERIES
SPECIFICATION
All specifications are based on 25°C, Nominal Input Voltage, and Maximum Output Current unless otherwise noted.
We reserve the right to change specifications based on technological advances.
GENERAL SPECIFICATIONS
Efficiency
Switching Frequency
Isolation Voltage
Isolation Capacitance
Leakage Current
Clearance/Creepage
TEST CONDITIONS
Min
Typ
Max
Unit
kHz
VAC
pF
µA
mm
°C
°C
°C/W
% RH
Nominal input voltage and full load
1 minute
240VAC, 60Hz
Input to Output
225
3000
See Table
250
275
12
17
2
ENVIRONMENTAL SPECIFICATIONS
Operating Ambient Temperature
Storage Temperature Range
Thermal Impedance
Relative Humidity
Thermal Shock
Vibration
MTBF
Weight
6.6
-40
+88
-55
5
18
Without derating
With derating
Natural convection (20LFM)
+88
+105
+125
95
MIL-STD-810F
MIL-STD-810F
4,718,000 hours
PHYSICAL SPECIFICATIONS
Dimensions (L x W x H)
Case Material
Base Material
Potting Material
MIL-HDBK-217F Full Load
SAFETY & EMC CHARACTERISTICS
Safety Approvals (pending)
EMI (See Note 2)
ESD
Radiated Immunity
Fast Transient (See Note 3)
Surge (See Note 3)
Conducted Immunity
Power Frequency Magnetic Field
0.48oz (14g)
1.25x0.80x0.40 inches
(31.8x20.3x10.2mm)
Non-conductive black plastic
Non-conductive black plastic
Silicon (UL94-V0)
ANSI/AAMI ES60601-1, IEC60601-1, EN60601-1, UL60950-1
(6)
, EN60950-1, IEC60950-1
EN55011, EN55022, and FCC Part 18
Class A, Class B
Air ±8kV
EN61000-4-2
Perf. Criteria A
Contact ±6kV
EN61000-4-3
10 V/m
Perf. Criteria A
EN61000-4-4
±2kV
Perf. Criteria A
EN61000-4-5
±2kV
Perf. Criteria A
EN61000-4-6
10 Vrms
Perf. Criteria A
EN61000-4-8
100A/m continuous; 1000A/m 1 second
Perf. Criteria A
1. The “x” in the model number represents the Pin Connection type. It can be “A” for pin connection type A or “B” for pin connection type B. See
mechanical drawings on page 4 for more information.
2. The DCMOP06 series meets EMI Class A without an external filter added. This series can only meet EMI Class B with external components
added. Please contact factory for more information.
3. An external input filter capacitor is required if the module has to meet EN61000-4-4, EN61000-4-5.
• For 5VDC nominal input models we recommend connecting an aluminum electrolytic capacitor (Nippon Chemi-con KY series, 1000µF/25V)
and a reverse diode (Vishay V10P45) in parallel.
• For 12VDC & 24VDC nominal input models we recommend connecting an aluminum electrolytic capacitor (Nippon Chemi-con KY series,
470µF/50V) in parallel.
• For 48VDC nominal input models we recommend connecting an aluminum electrolytic capacitor (Nippon Chemi-con KY series, 330µF/100V)
in parallel.
4. Remote ON/OFF control is optional and is only available for “B” type pin connection models. To order the converter with remote ON/OFF add the
suffix “-P” to the model number (Ex: DCMOP06-48S12B-P).
5. Trim function is optional and is only available for “B” type pin connection models. To order the converter with Trim pin add the suffix “-T” to the
model number (Ex: DCMOP06-48S12B-T).
6. This product is Listed to applicable standards and requirements by UL.
CAUTION:
This power module is not internally fused. An input line fuse must always be used.
NOTES
*Due to advances in technology, specifications subject to change without notice.
5/21/2019
Wall Industries, Inc. • Tel: 603-778-2300 • Toll Free: 888-597-9255 •
website:
www.wallindustries.com
• e-mail:
sales@wallindustries.com
Page 3 of 5
Rev C
DCMOP06 SERIES
6 Watt Medical DC/DC Power
Converters
CHARACTERISTIC CURVES
DCMOP06-48S05A Derating Curve
DCMOP06-48S05A Efficiency vs Input Voltage
DCMOP06-48S05A Efficiency vs Output Load
MECHANICAL DRAWINGS
A Type Pin Connection (Suffix “A”)
PIN
1
11
12
13
15
23
24
PIN CONNECTIONS
SINGLE
+INPUT
NO PIN
-OUTPUT
+OUTPUT
NO PIN
-INPUT
-INPUT
+INPUT
COMMON
NO PIN
-OUTPUT
+OUTPUT
-INPUT
-INPUT
DUAL
1. Dimensions in inch (mm)
2. Tolerance: x.xx±0.02 (x.x±0.5)
x.xxx±0.01 (x.xx±0.25)
3. Pin Pitch Tolerance: ±0.01 (0.25)
4. Pin Dimension Tolerance: ±0.004 (0.1)
B Type Pin Connection (Suffix “B”)
PIN
1
2
10
11
14
16
22
23
SINGLE
CTRL (Optional)
-INPUT
TRIM (Optional)
**NO PIN/NC
+OUTPUT
-OUTPUT
+INPUT
+INPUT
PIN CONNECTIONS
DUAL
CTRL (Optional)
-INPUT
TRIM (Optional)
-OUTPUT
+OUTPUT
COMMON
+INPUT
+INPUT
**For single output models, Pin 11 is “NO PIN” with the trim
pin option (Suffix “-T”) and “NC” without the trim pin option.
1. Dimensions in inch (mm)
2. Tolerance: x.xx±0.02 (x.x±0.5)
x.xxx±0.01 (x.xx±0.25)
3. Pin Pitch Tolerance: ±0.01 (0.25)
4. Pin Dimension Tolerance: ±0.004 (0.1)
Output can be externally trimmed by using the method shown
below. ( ) for dual output trim
EXTERNAL OUTPUT TRIMMING
5/21/2019
Wall Industries, Inc. • Tel: 603-778-2300 • Toll Free: 888-597-9255 •
website:
www.wallindustries.com
• e-mail:
sales@wallindustries.com
Page 4 of 5
Rev C
DCMOP06 SERIES
6 Watt Medical DC/DC Power
Converters
MODEL NUMBER SETUP
DCMOP
Series Name
06
Output Power
06:
6 Watts
-
48
Input Voltage
5:
5 VDC
12:
12 VDC
24:
24 VDC
48:
48 VDC
S
Output Quantity
S:
Single Output
05
Output Voltage
33:
3.3 VDC
05:
5 VDC
12:
12 VDC
15:
15 VDC
24:
24 VDC
Pin
Connection
A:
A Type
B:
B Type
B
-
None:
P
(1)
Remote ON/OFF Option
No Remote
ON/OFF
P:
Remote ON/OFF
T
(1)
Trim Option
None
No Trim
:
T:
Trim
D:
Dual Output
05:
±5 VDC
12:
±12 VDC
15:
±15 VDC
(1) Remote ON/OFF Control and Trim options are only available for “B” type pin connection models.
COMPANY INFORMATION
Wall Industries, Inc. has created custom and modified units for over 50 years. Our in-house research and development
engineers will provide a solution that exceeds your performance requirements on-time and on budget. Our ISO9001
certification is just one example of our commitment to producing a high quality, well-documented product for our customers.
Our past projects demonstrate our commitment to you, our customer. Wall Industries, Inc. has a reputation for working closely
with its customers to ensure each solution meets or exceeds form, fit and function requirements. We will continue to provide
ongoing support for your project above and beyond the design and production phases. Give us a call today to discuss your
future projects.
Contact
Wall Industries
for further information:
Phone:
Toll Free:
Fax:
E-mail:
Web:
Address:
(603)778-2300
(888)597-9255
(603)778-9797
sales@wallindustries.com
www.wallindustries.com
37 Industrial Drive
Exeter, NH 03833
©2019 Wall Industries, Inc. Specifications subject to change without notice. Wall Industries is not responsible for typographical errors. The information contained herein is for
informational purposes only. This information is provided by Wall Industries and we make no representations or warranties of any kind, express or implied, about the completeness,
accuracy, reliability, suitability or availability with respect to the information contained in this document for any purpose. All product and manufacturer names are trademarks or
registered trademarks of their respective companies.
5/21/2019
Wall Industries, Inc. • Tel: 603-778-2300 • Toll Free: 888-597-9255 •
website:
www.wallindustries.com
• e-mail:
sales@wallindustries.com
Page 5 of 5