Process Indicator
K3HB-X
CSM_K3HB-X_DS_E_11_2
A Process Indicator Ideal for Discriminating and
Displaying Measurements for Voltage/Current
Signals
• Easy recognition of judgement results using color display that can be
switched between red and green.
*
• Equipped with a position meter for monitoring operating status trends.
• External event input allows use in various measurement and discrimina-
tion applications.
• Series expanded to include DeviceNet models.
• Short body with depth of only 95 mm (from behind the front panel), or 97
mm for DeviceNet models.
• UL certification approval (Certification Mark License).
• CE Marking conformance by third party assessment body.
• Water-resistant enclosure conforms to NEMA 4X (equivalent to IP66).
• Capable of high-speed sampling at 50 times per second (20 ms)
• Easy-to-set two-point scaling allows conversion and display of any user-
set values.
*
Visual confirmation of judgement results is not supported on models that do not have
an output or models that do not support DeviceNet.
You can change the display color by setting it, but you cannot switch it based on the
judgement results.
For the most recent information on models that have been certified for
safety standards, refer to your OMRON website.
Refer to
Safety Precautions for All Digital Panel
Meters.
Model Number Structure
■
Model Number Legend
Base Units and Optional Boards can be ordered individually or as sets.
Base Units
Base Units with Optional Boards
K3HB-X@
1
5
1. Input Sensor Code
VD: DC voltage input
AD: DC current input
VA: AC voltage input
AA: AC current input
K3HB-X@-@@@
1 2 3 4
5
2. Sensor Power Supply/Output Type Code
None: None
CPA: Relay output (PASS: SPDT) + Sensor power supply
(12 VDC +/−10%, 80 mA) (See note 1.)
L1A: Linear current output (0 to 20 or 4 to 20 mA DC) + Sensor power supply
(12 VDC +/−10%, 80 mA) (See note 2.)
L2A: Linear voltage output (0 to 5, 1 to 5, or 0 to 10 VDC) + Sensor power supply
(12 VDC +/−10%, 80 mA) (See note 2.)
A: Sensor power supply (12 VDC +/−10%, 80 mA)
FLK1A: Communications (RS-232C) + Sensor power supply
(12 VDC +/−10%, 80 mA) (See note 2.)
FLK3A: Communications (RS-485) + Sensor power supply
(12 VDC +/−10%, 80 mA) (See note 2.)
5. Supply Voltage
100-240 VAC: 100 to 240 VAC
24 VAC/VDC: 24 VAC/VDC
Note: 1.
CPA can be combined with relay outputs only.
2.
Only one of the following can be used by each Digital Indicator:
RS-232C/RS-485 communications, a linear output, or
DeviceNet communications.
Optional Board
Sensor Power Supply/Output Boards
3. Relay/Transistor Output Type Code
None: None
C1: Relay contact (H/L: SPDT each)
C2: Relay contact (HH/H/LL/L: SPST-NO each)
T1: Transistor (NPN open collector: HH/H/PASS/L/LL)
T2: Transistor (PNP open collector: HH/H/PASS/L/LL)
BCD
*:
BCD output + transistor output (NPN open collector: HH/H/PASS/L/
LL)
DRT: DeviceNet (See note 2.)
*
A Special BCD Output Cable (sold separately) is required.
K33-@
2
Relay/Transistor Output Boards
K34-@
3
Event Input Boards
4. Event input Type Code
None: None
1: 5 inputs (M3 terminal blocks), NPN open collector
2: 8 inputs (10-pin MIL connector), NPN open collector
3: 5 inputs (M3 terminal blocks), PNP open collector
4: 8 inputs (10-pin MIL connector), PNP open collector
K35-@
4
Note:
The following combinations are not possible.
• Communications (FLK@A) + DeviceNet (DRT)
• Communications (FLK@A) + BCD output (BCD)
• Linear current/voltage (L@A) + DeviceNet (DRT)
1
K3HB-X
Accessories (Sold Separately)
K32-DICN: Special Cable (for event inputs, with 8-pin connector)
K32-BCD: Special BCD Output Cable
Watertight Cover
Model
Y92A-49N
Rubber Packing
Model
K32-P1
Note:
Rubber packing is provided with the Controller.
Specifications
■
Ratings
Power supply voltage
Allowable power supply voltage range
Power consumption
(See note 1.)
Current consumption
Input
A/D conversion method
External power supply
Event inputs (See
note 2.)
Timing input
100 to 240 VAC (50/60 Hz), 24 VAC/VDC, DeviceNet power supply: 24 VDC
85% to 110% of the rated power supply voltage, DeviceNet power supply: 11 to 25 VDC
100 to 240 V: 18 VA max. (max. load)
24 VAC/DC: 11 VA/7 W max. (max. load)
DeviceNet power supply: 50 mA max. (24 VDC)
DC voltage, DC current, AC voltage, AC current
Delta-Sigma method
12 VDC ±10%, 80 mA (models with external power supply only)
NPN open collector or no-voltage contact signal
ON residual voltage: 3 V max.
ON current at 0
Ω:
17 mA max.
Max. applied voltage: 30 VDC max.
OFF leakage current: 1.5 mA max.
Startup compensa-
NPN open collector or no-voltage contact signal
tion timer input
ON residual voltage: 2 V max.
ON current at 0
Ω:
4 mA max.
Hold input
Max. applied voltage: 30 VDC max.
Reset input
OFF leakage current: 0.1 mA max.
Forced-zero input
Bank input
Output ratings (de- Relay output
pends on the mod-
el)
Transistor output
Linear output
250 VAC, 30 VDC, 5 A (resistive load)
Mechanical life expectancy: 5,000,000 operations, Electrical life expectancy: 100,000 operations
Maximum load voltage: 24 VDC, Maximum load current: 50 mA, Leakage current: 100
μA
max.
Linear output 0 to 20 mA DC, 4 to 20 mA DC:
Load: 500
Ω
max, Resolution: Approx. 10,000, Output error: ±0.5% FS
Linear output 0 to 5 VDC, 1 to 5 VDC, 0 to 10 VDC:
Load: 5 kΩ max, Resolution: Approx. 10,000, Output error: ±0.5% FS
(1 V or less: ±0.15 V; no output for 0 V or less)
Negative LCD (backlit LED) display
7-segment digital display (Character height: PV: 14.2 mm (green/red); SV: 4.9 mm (green)
Scaling function, measurement operation selection, averaging, previous average value comparison, forced-ze-
ro, zero-limit, output hysteresis, output OFF delay, output test, teaching, display value selection, display color
selection, key protection, bank selection, display refresh period, maximum/minimum hold, reset
−10
to 55°C (with no icing or condensation)
25% to 85%
−25
to 65°C (with no icing or condensation)
2,000 m max.
Watertight packing, 2 fixtures, terminal cover, unit stickers, instruction manual. DeviceNet models also include
a DeviceNet connector (Hirose HR31-5.08P-5SC(01)) and crimp terminals (Hirose HR31-SC-121) (See note
3.)
Display method
Main functions
Ambient operating temperature
Ambient operating humidity
Storage temperature
Altitude
Accessories
Note: 1.
DC power supply models require a control power supply capacity of approximately 1 A per Unit when power is turned ON. Particular
attention is required when using two or more DC power supply models. The OMRON S8VS-series DC Power Supply Unit is recommend-
ed.
2.
PNP input types are also available.
3.
For K3HB-series DeviceNet models, use only the DeviceNet Connector included with the product. The crimp terminals provided are for
Thin Cables.
2
K3HB-X
■
Characteristics
Display range
Sampling period
−19,999
to 99,999
20 ms (50 times/second)
Comparative output response time
DC input: 100 ms max.; AC input: 300 ms max.
(The time until the comparative output is output when there is a forced sudden change in the input signal
from 15% to 95% or 95% to 15%.)
Linear output response time
DC input: 150 ms max.; AC input: 420 ms max.
(The time until the final analog output value is reached when there is a forced sudden change in the output
signal from 15% to 95% or 95% to 15%.)
20 MΩ min. (at 500 VDC)
2,300 VAC for 1 min between external terminals and case
100 to 240 VAC models:
±1,500
V at power supply terminals in normal or common mode
(waveform with 1-ns rising edge and pulse width of 1
μs/100
ns)
24 VAC/VDC models:
±1,500
V at power supply terminals in normal or common mode
(waveform with 1-ns rising edge and pulse width of 1
μs/100
ns)
Frequency: 10 to 55 Hz; Acceleration: 50 m/s
2
, 10 sweeps of 5 min each in X, Y, and Z directions
150 m/s
2
(100 m/s
2
for relay outputs) 3 times each in 3 axes, 6 directions
Approx. 300 g (Base Unit only)
Front panel
Rear case
Terminals
Memory protection
Applicable standards
Conforms to NEMA 4X for indoor use (equivalent to IP66)
IP20
IP00 + finger protection (VDE0106/100)
EEPROM (non-volatile memory)
Number of rewrites: 100,000
UL61010-1, CSA C22.2 No.61010-1-04
EN61010-1 (IEC61010-1): Pollution degree 2/Overvoltage category II
EN61326-1
EMI: EN61326-1 Industrial electromagnetic environment
Electromagnetic radiation interference
CISPR 11 Group 1, Class A
Terminal interference voltage
CISPR 11 Group 1, Class A
EMS: EN61326-1 Industrial electromagnetic environment
Electrostatic Discharge Immunity
EN61000-4-2: 4 kV (contact), 8 kV (in air)
Radiated Electromagnetic Field Immunity
EN61000-4-3: 10 V/m 1 kHz sine wave amplitude modulation (80 MHz to 1 GHz, 1.4 to 2 GHz)
Electrical Fast Transient/Burst Immunity
EN61000-4-4: 2 kV (power line), 1 kV (I/O signal line)
Surge Immunity
EN61000-4-5: 1 kV with line (power line), 2 kV with ground (power line)
Conducted Disturbance Immunity
EN61000-4-6: 3 V (0.15 to 80 MHz)
Power Frequency Magnetic Immunity
EN61000-4-8: 30 A/m (50 Hz) continuous time
Voltage Dips and Interruptions Immunity
EN61000-4-11: 0.5 cycle, 0°/180°, 100% (rated voltage)
Insulation resistance
Dielectric strength
Noise immunity
Vibration resistance
Shock resistance
Weight
Degree of
protection
EMC
3
K3HB-X
■
Input Range (Measurement Range and Accuracy) CAT II
Input type
Range
Set value
a Ud
b Ud
c Ud
d Ud
a ad
b ad
c ad
d ad
a Ua
b Ua
c Ua
d Ua
a aa
b aa
c aa
d aa
Measurement range
Maximum
measurement range
-199.99 to 219.99 V
-19.999 to 21.999 V
-1.9999 to 2.1999 V
0.5000 to 5.5000 V
199.99 to 219.99 mA
19.999 to 21.999 mA
1.9999 to 2.1999 mA
2.000 to 22.000 mA
0.0 to 440.0 V
0.00 to 219.99 V
0.000 to 21.999 V
0.0000 to 1.9999 V
0.000 to 11.000 A
0.0000 to 2.1999 A
0.00 to 219.99m A
0.000 to 21.999m A
(0.5 VA CT)
(See note 3.)
(0.5 VA CT)
(See note 3.)
1
Ω
max.
10
Ω
max.
±0.5%rdg ± 10
digits max.
2A
1
Ω
max.
10
Ω
max.
33
Ω
max.
10
Ω
max.
1 MΩ min.
±0.3%rdg ± 5
digits max.
±0.5%rdg ± 10
digits max.
±0.5%rdg ± 20
digits max.
700 V
400 V
20 A
±0.1%rdg ± 1
digit max.
±400 mA
±200 mA
Input
impedance
10 MΩ min.
1 MΩ min.
Accuracy
Allowable
instantaneous
overload (30 s)
±400 V
±200 V
K3HB-XVD
DC voltage
A
B
C
D
K3HB-XAD
DC current
A
B
C
D
K3HB-XVA
A
AC voltage
B
(See note 4.)
C
D
K3HB-XAA
AC current
A
B
C
D
±199.99 V
±19.999 V
±1.9999 V
1.0000 to 5.0000 V
±199.99 mA
±19.999 mA
±1.9999 mA
4.000 to 20.000 mA
0.0 to 400.0 V
0.00 to 199.99 V
0.000 to 19.999 V
0.0000 to 1.9999 V
0.000 to 10.000 A
0.0000 to 1.9999 A
0.00 to 199.99 mA
0.000 to 19.999 mA
±0.1%rdg ± 1
digit max.
Note: 1.
The accuracy is for an input frequency range of 40 Hz to 1 kHz (except for AD current input A and B ranges of 50 to 60 Hz) and an ambient
temperature of 23 ±5°C. The error, however, increases below 10% of the maximum input value.
DC voltage input (all ranges): 10% or less of max. input = ±0.15% FS
DC current input (all ranges): 10% or less of max. input = ±0.1% FS
AC voltage input (A: 0.0 to 400.0 V): 10% or less of max. input = ±0.15% FS
AC voltage input (B: 0.00 to 199.99 V): 10% or less of max. input = ±0.2% FS
AC voltage input (C: 0.000 to 19.999 V; D: 0.0000 to 1.9999 V): 10% or less of max. input = ±1.0% FS
AC current input (A: 0.000 to 10.000 A): 10% or less of max. input = ±0.25% FS
AC current input (B: 0.0000 to 1.9999 A): 10% or less of max. input = ±0.5% FS
AC current input (C: 0.00 to 199.99 mA; D: 0.000 to 19.999 A): 10% or less of max. input = ±0.15% FS
When DC voltage input models are used with a ±1.9999 V range, make sure that the connections between input terminals are not open.
If the input terminals are open, the display will show large variations. Connect resistance of approximately 1 MΩ between the input ter-
minals if they are open.
2.
The letters “rdg” mean “reading” and refer to the input error.
3.
The value (0.5 VA CT) is the VA consumption of the internal CT (current transformer).
E2
E3
±1.9999
V
+
1 MΩ
E4
E5
E6 COM
−
4.
The K3HB-XVA@@ complies with UL standards when the applied input voltage is within the range 0 to 150 VAC.
If the input voltage is higher than 150 VAC, install an external transformer or take other measures to drop the voltage to 150 VAC or lower.
4
K3HB-X
Common Specifications
■
Event Input Ratings
Input type
Contact
No-contact
S-TMR, HOLD, RESET, ZERO, BANK1, BANK2,
BANK4
ON: 1 kΩ max., OFF: 100 kΩ min.
ON residual voltage: 2 V max.
OFF leakage current: 0.1 mA max.
Load current: 4 mA max.
Maximum applied voltage: 30 VDC max.
---
ON residual voltage: 3 V max.
OFF leakage current: 1.5 mA max.
Load current: 17 mA max.
Maximum applied voltage: 30 VDC max.
TIMING
■
Output Ratings
Contact Output
Item
Resistive loads
(250 VAC, cosφ=1;
30 VDC, L/R=0 ms)
5 A at 250 VAC
5 A at 30 VDC
Inductive loads
(250 VAC, closed
circuit, cosφ=0.4;
30 VDC, L/R=7 ms)
1 A at 250 VAC
1 A at 30 VDC
Transistor Output
Maximum load voltage
Maximum load current
Leakage current
24 VDC
50 mA
100
μA
max.
Rated load
Mechanical life
5,000,000 operations
expectancy
Electrical life
expectancy
100,000 operations
Linear Output
Item
Allowable load impedance
Resolution
Output error
0 to 20 mA
500
Ω
max.
Approx. 10,000
±0.5%FS
±0.5%FS (1 V or less: ±0.15 V; no output for 0 V or less.)
4 to 20 mA
0 to 5 V
5 kΩ min.
1 to 5 V
0 to 10 V
Serial Communications Output
Item
Synchronization method
Baud rate
Transmission code
Data length
Stop bit length
Error detection
Parity check
RS-232C, RS-485
Start-stop synchronization
9,600, 19,200, or 38,400 bps
ASCII
7 bits or 8 bits
2 bits or 1 bit
Vertical parity and FCS
Odd, even
Communications method
Half duplex
BCD Output I/O Ratings
(Input Signal Logic: Negative)
I/O signal name
Inputs
REQUEST
HOLD
MAX
MIN
RESET
Item
Input signal
Input current for
no-voltage input
Rating
No-voltage
contact input
10 mA
Signal ON voltage 1.5 V max.
level
OFF voltage 3 V min.
Maximum load
voltage
Maximum load
current
Leakage current
Maximum load
voltage
Maximum load
current
Leakage current
24 VDC
10 mA
100
μA
max.
24 VDC
50 mA
100
μA
max.
Note:
For details on serial and DeviceNet communications, refer to
the
Digital Indicator K3HB Communications User's Manual
(Cat.No. N129).
Outputs
DATA
POLARITY
OVER
DATA VALID
RUN
HH
H
PASS
L
LL
Note:
For details on serial and DeviceNet communications, refer to
the
Digital Indicator K3HB Communications User's Manual
(Cat.No. N129).
5