XPC/XPCL Series and
XPXL Series
Compensated/Calibrated and
Uncompensated/Uncalibrated
Board Mount Pressure Sensors
DESCRIPTION
The XPC/XPCL Series and XPXL Series sensors integrate
silicon micromachined sensing technology, temperature
compensation, and calibration in a complete family of low-cost
packages. This series offers a cost-effective solution for design
requirements.
These piezoresistive pressure sensors use micromachined
silicon chips mounted on ceramic and protected with a plastic
cap. Several tube arrangements with nylon housings are
available for various pressure applications.
FEATURES
Low cost, small size
Temperature compensated
Zero and span calibrated
Millivolt output
Differential, gage and absolute pressure
Constant voltage excitation
High impedance - low current
POTENTIAL APPLICATIONS
Medical applications
Applications requiring small size
Applications requiring vacuum and positive pressure
reference or both
Although the sensors are designed for use with noncorrosive,
nonionic pressure media, they accommodate many gases that
are used in medical applications.
On devices of 5 psi and above, the top side of the chip is
protected against humidity by a Silgel® coating.
XPC/XPCL Series and XPXL Series
Table 1. Electrical Specifications
Characteristic
XPC/XPCL at 12 Vdc, 25 °C [77 °F]
Min.
Typ.
Max.
3.0
12.0
16.0
-1.0
0
1.0
±1.0
XPXL at 5 Vdc, 25 °C [77 °F]
Min.
Typ.
Max.
3.0
5.0
12.0
-50
0
50
–
–
–
–
–
-25 [-13]
-40 [-40]
–
±0.5
-1800
PPM/ C
0.25
3.0
3.0
–
–
–
–
–
1.0
–
–
85 [185]
125 [257]
50
Unit
Vdc
mV
mV
Excitation voltage
Null
1
Offset temperature shift
–
–
0 C to 25 C [32 F to 77 F]
25 C to 70 C [77 F to 158 F]
1
Full scale temperature shift
–
–
0 C to 25 C [32 F to 158 F]
25 C to 70 C [77 F to 158 F]
2
–
0.25
Linearity, hysteresis error
Input resistance
5.0
–
Output resistance
–
3.0
Operating temperature
-25 [-13]
–
Storage temperature
-40 [-40]
–
Common mode pressure
–
–
Notes:
1. Shift is relative to 25 °C [77 °F].
2. Measured at 1/2 full scale rated pressure using BFSL.
Table 2. Absolute Output
FS Pressure
±2
1.0
–
–
85 [185]
125 [257]
50
% span
% span
kOhm
kOhm
C [ F]
C [ F]
psi
5 psi
15 psi
30 psi
60 psi
100 psi
XPC at 12 Vdc, 25 °C [77 °F]
Output Full Scale Span (mV)
Min.
Max.
Max.
57
60
63
85
90
95
85
90
95
85
90
95
95
100
105
1
Overpressure
(psi)
Max.
15
45
90
180
250
Table 3. Gage/Differential Output
FS Pressure
XPC/XPCL at 12 Vdc, 25 °C [77 °F]
Output Full Scale Span (mV)
Min.
Typ.
Max.
4 inH
2
O
23
25
27
10 inH
2
O
19
20
21
1 psi
17
18
19
5 psi
57
60
63
15 psi
85
90
95
30 psi
85
90
95
60 psi
85
90
95
100 psi
95
100
105
150 psi
85
90
95
Note:
1. Differential common mode pressure should not exceed 50 psi.
Figure 1. Mounting Dimensions (For reference only: mm (in).)
D/A Housing
GFS/AFS Housing
(Differential/Absolute)
(Gage/Absolute Offset Axial)
XPXL at 5 Vdc, 25 °C [77 °F]
Output Full Scale Span (mV)
Min.
Typ.
Max.
50
68
86
45
78.5
112
40
75
110
112
168.5
225
168
253
338
168
253
338
189
263.5
338
210
295
380
187
262.5
338
Overpressure
(psi)
Max.
3
3
3
15
45
90
180
250
250
GBFS Housing
(Gage Backside Offset Axial)
2
www.honeywell.com/sensing
Figure 2. Equivalent Circuit
Pinout
1. -V Excitation
2. + Output Signal
3. +V Excitation
4. - Output Signal
Order Guide
1
Media compatibility:
Clean, dry gases only.
P1 Port:
Media must be compatible with epoxy-based adhesive and silicon gel.
P2 Port:
Media must be compatible with nylon housing, epoxy-based adhesive and
silicon.
XPCL
Series
XPC - compensated and calibrated (psi)
XPCL - compensated and calibrated (inH
2
0)
XPXL - uncompensated and uncalibrated (inH
2
0)
04
G
B
F
C
Accuracy Grade
(XPC, XPCL only)
C - Commercial
H - High Accuracy
Pressure Range
04 inH
2
0, 10 inH
2
0
01, 05, 15, 30, 60, 100, 150 psi
Pressure Reference
G - Gage
D - Differential
A - Absolute > 5 psi
Port Option
Port Operation
2
B - Backside
F - Axial
T - Radial
FS - Offset Axial
2
Notes:
1. Not all combinations may be available. Contact Honeywell representative for details.
2. Option available in ported gage version only.
WARNING
PERSONAL INJURY
DO NOT USE these products as safety or emergency stop
devices or in any other application where failure of the
product could result in personal injury.
Failure to comply with these instructions could result in
death or serious injury.
WARRANTY/REMEDY
Honeywell warrants goods of its manufacture as being free of
defective materials and faulty workmanship. Honeywell’s
standard product warranty applies unless agreed to otherwise
by Honeywell in writing; please refer to your order
acknowledgement or consult your local sales office for specific
warranty details. If warranted goods are returned to Honeywell
during the period of coverage, Honeywell will repair or replace,
at its option, without charge those items it finds defective.
The
foregoing is buyer’s sole remedy and is in lieu of all other
warranties, expressed or implied, including those of
merchantability and fitness for a particular purpose. In no
event shall Honeywell be liable for consequential, special,
or indirect damages.
While we provide application assistance personally, through
our literature and the Honeywell web site, it is up to the
customer to determine the suitability of the product in the
application.
Specifications may change without notice. The information we
supply is believed to be accurate and reliable as of this
printing. However, we assume no responsibility for its use.
WARNING
MISUSE OF DOCUMENTATION
The information presented in this product sheet is for
reference only. Do not use this document as a product
installation guide.
Complete installation, operation, and maintenance
information is provided in the instructions supplied with
each product.
Failure to comply with these instructions could result in
death or serious injury.
SALES AND SERVICE
Honeywell serves its customers through a worldwide network
of sales offices, representatives and distributors. For
application assistance, current specifications, pricing or name
of the nearest Authorized Distributor, contact your local sales
office or:
E-mail:
info.sc@honeywell.com
Internet:
www.honeywell.com/sensing
Phone and Fax:
+65 6355-2828
+65 6445-3033 Fax
Europe
+44 (0) 1698 481481
+44 (0) 1698 481676 Fax
Latin America +1-305-805-8188
+1-305-883-8257 Fax
USA/Canada +1-800-537-6945
+1-815-235-6847
+1-815-235-6545 Fax
Asia Pacific
Sensing and Control
Honeywell
1985 Douglas Drive North
Golden Valley, MN 55422
www.honeywell.com/sensing
008046-6-EN IL50 GLO Printed in USA
July 2011
© 2011 Honeywell International Inc. All rights reserved.