Installation Instructions for the
22PC Series Pressure Sensors
WARNING
PERSONAL INJURY
DO NOT USE these products as safety or
emergency stop devices, or in any other
application where failure of the product could
result in personal injury.
Failure to comply with these instructions
could result in death or serious injury.
GENERAL INFORMATION
22PC Series pressure sensors are four active
element piezoresistive bridges. When pressure is
applied, resistance changes and the 22PC
provides an output signal proportional to the input
pressure.
Gage
pressure is measured with respect to
ambient pressure. When applied pressure
increases, the differential voltage V2 - V4
increases. As pressure decreases, differential
voltage V2 - V4 decreases.
Differential
sensors provide a differential
voltage proportional to the pressure differential
between port P2 and P1. As DP = P2 – P1
increases, differential voltage V2 - V4 increases.
As DP = P2 – P1 decreases, differential voltage
V2 - V4 decreases.
PERFORMANCE CHARACTERISTICS
at 10.0
±
0.01 VDC Excitation, 25
°C
Min.
Typ.
Max.
12
1.0
6.0 k
6.0 k
Units
Vdc
ms
Ohm
Ohm
gram
ISSUE 2
PK 80085
Note: For media compatibility specifications refer to
catalog or web site:
www.honeywell.com/sensing
22PC CIRCUIT TERMINATION
1
+
V
S
2
+
V
-
O
4
-
3
Pin 1 = Vs (+)
Pin 2 = Output + (V2)
Pin 1 is notched
Pin 2 is next to Pin 1, etc.
Pin 4 = Output - (V4)
Pin 3 = Ground (-)
CONSTANT CURRENT EXCITATION
SCHEMATIC
1.5 mA constant current excitation significantly
reduces sensitivity shift over temperature.
Constant
Current
Source
R1
+
Output Voltage
R4
R2
R5
R3
Shunt Resistor for
High Temperture
Corrections
-
Excitation
—
10
Response Time
—
—
Input Resistance*
4.0 k
5.0 k
Output Resistance*
4.0 k
5.0 k
Weight
2
*
Measured using a 1 mA current
Noncompensated pressure sensors, excited by
constant current instead of voltage, exhibit temperature
compensation of Span.
Constant current excitation has an additional benefit of
temperature measurement. When driven by a constant
current source, a silicon pressure sensor’s terminal
voltage will rise with increased temperature. The rise in
voltage not only compensates for the Span, but is also an
indication of sensor temperature.
SOLDERING
Limit soldering to 315
°C
[600
°F]
maximum, with
duration of 10 seconds maximum.
CLEANING
Proper cleaning fluids should be selected, based
on the type of contaminants to be removed.
Honeywell recommends alcohols or fluorinated
solvents. Do not immerse the sensor.
ENVIRONMENTAL SPECIFICATIONS
Operating
Temperature
Storage
Temperature
Shock
Vibration
-40
°C
to 85
°C
[-40
°F
to 185
°F]
-55
°C
to 100
°C
[-67
°F
to 212
°F]
Qualification tested to 150 G
Qualification tested: 0 kHz to
2 kHz, 20 G sine
Automation and Control Products
Honeywell 2002, All Rights Reserved
22PC Series Pressure Sensors
2
CATALOG
LISTING
STAMP
A
F
6
G
XX
XX
G
6
F
A
2
2002
DATE STAMP
CODE
ISSUE 2
PK 80085
22PC LASER BRANDING SCHEME
(Note: Alternate form is the enire catalog listing)
= Special (Sequentially numbered)
= Pressure measurement: G = Gage, D = Differential
= Termination style: 1 = 1 x 4 (short), 2 = 2 x 2, or
6 = 1 x 4 (long)
= Media seal: F = Fluorosilicone, E = EPM, N = Neoprene,
S = Silicone, V= Viton
= Range (psi) : A = 1, C = 15, F = 100
= Type: (2 = Uncompensated)
10
(Note: Alternate date code is 102002)
10
= Week in the year
2002
= Year
ADDITIONAL PORT VARIATIONS
(dimensions for reference only) mm/in
B Barbed
D Modular
H M5 Thread*
J Needle
M ¼-28 Thread
S Manifold
Accessory
PC-10182 — Steel lockring
* Recommended torque for sealing is 4 in-lb. Do not exceed 6 in-lb of torque. Use size 007 O-Ring. O-Ring
counterbore dimensions are 0.04
±
.005 in D x 0.300
±
.003 in Dia.
** Recommended torque for sealing is 8 in-lb. Do not exceed 12 in-lb. Use size 009 O-Ring. O-Ring counterbore
dimensions are .040
±
.002 in D x 0.360
±
.003 in Dia.
2
Honeywell
•
Automation and Control Products
22PC Series Pressure Sensors
MOUNTING DIMENSIONS
(dimensions for reference only) mm/in
GAGE SENSOR
1 x 4 Termination (Style 6) Port Style A, Straight
ISSUE
2
PK 80085
2 x 2 Termination (Style 2) Port Style A, Straight
Pin 1 is notched, Pin 2 is next to Pin 1, etc.
Pin 1 is notched, Pin 1 is shown at lower right corner.
Pins 2, 3, 4 are clockwise.
DIFFERENTIAL SENSOR 1 X 4 Termination (Style 6) Port Style A, Straight (Only)
Honeywell
•
Automations and Control Products
3
22PC Series Pressure Sensors
WARRANTY/REMEDY
Honeywell warrants goods of its manufacture as
being free of defective materials and faulty
workmanship. Contact your local sales office for
warranty information. If warranted goods are
returned to Honeywell during the period of
coverage,
Honeywell will repair or replace without charge
those items it finds defective. The foregoing is
Buyer’s sole remedy and is
in lieu of all other
warranties, expressed or implied,
merchantability and fitness for a particular
purpose.
Specifications may change without notice. The
information we supply is believed to be accurate
and reliable as of this printing. However, we
assume no responsibility for its use.
ISSUE 2
PK 80085
While we provide application assistance
personally, through our literature and the
Honeywell web site, it is up to the customer to
determine the suitability of the product in the
application.
For application assistance, current specifications,
or name of the nearest Authorized Distributor,
check the Honeywell web site or call:
1-800-537-6945 USA
1-800-737-3360 Canada
1-815-235-6847 International
FAX
1-815-235-6545 USA
INTERNET
www.honeywell.com/sensing
info.sc@honeywell.com
Automation and Control Products
Honeywell Inc.
11 West Spring Street
Freeport, Illinois 61032
PK 80085-2-EN IL50 GLO 0902 Printed in USA