GPC40 Overview

PWR SPLY MEDICAL TRI OUT 5 -+12

GPC40 Commercial/GPM40 Medical
40 WATT GLOBAL PERFORMANCE SWITCHERS
GLOBAL PERFORMANCE SWITCHERS
Summary:
• Wide-range ac input 85-264 Vac
• 2-year warranty
• Conducted EMI exceeds FCC Class B and CISPR 22
Class B (Commercial models) and CISPR 11
Class B (Medical models)
• Single and multiple outputs
• Commercial Approved to UL1950,
EN60950 E3:2000 and CSA-C22.2 No. 234
• Medical Approved to UL2601-1, EN60601-1 and
CSA22.2 No. 601
• Single and multiple outputs
• RoHS compliant models available (G suffix)
DC Power Supplies
marked to LVD
Output Noise
0.5% rms, 1% pk-pk, 20 MHz bandwidth, differential mode. Measured with noise
probe directly across output terminals of the power supply.
Transient Response
Main output—500 μs typical response time for return to within 0.5% of final value
for a 50% load step change. ∆i/ ∆t<0.2 A/μs. Maximum voltage deviation is 3.5%.
Startup/ shutdown overshoot less than 3%.
Remote Sense
Provided as a standard feature on single-output models.
Voltage Adjustment
Built-in potentiometer adjusts main output voltage ±5%.
EMI/EMC Compliance
All models include built-in EMI filtering to meet the
following emissions requirements:
EMI SPECIFICATIONS
Conducted Emissions
Conducted Emissions
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Surge Susceptibility
Line Frequency Harmonics
COMPLIANCE LEVEL
EN55022 Class B; FCC Class B
EN55011 Class B; FCC Class B
EN61000-4-2, 6 kV contact, 8 kV air
EN61000-4-3, 3 V/meter
EN61000-4-4, 2 kV, 5 kH
EN61000-4-5, 1 kV diff., 2 kV com.
EN61000-3-2
SPECIFICATIONS
Ac Input
85-264 Vac, 47-63 Hz single phase.
Input Current
Maximum input current at 120 Vac, 60 Hz with full rated output load: 1.3 A
Hold-Up Time
20 ms minimum from loss of ac input at full load, nominal line (115 Vac).
Output Power
40 W continuous, 50 W peak. Peak ratings are for 60 s maximum duration,
10% duty cycle. During peak load condition, output regulation may exceed
total regulation limits.
Overload Protection
Fully protected against short circuit and output overload.
Short circuit protection is cycling type power limit on outputs 1 & 2; foldback
type on output 3. Recovery after fault is automatic. See output ratings chart
for additional notes or conditions.
Overvoltage Protection.
Main outputs: 124% ± 12%.
Efficiency
70% at full rated load, nominal input voltage, depending on model and load
distribution.
Turn-on Time
Less than 1 second at 120 Vac, 25°C (inversely proportional to input voltage
and thermistor temperature).
Input Protection
Internal ac fuse provided. Designed to blow only if a catastrophic failure oc-
curs in the unit.
Inrush Current
Inrush is limited by internal thermistors. Inrush at 240 Vac under cold start
conditions will not exceed 34 A.
Temperature Coefficient
0.03%/°C typical on all outputs.
Environmental
Designed for 0 to 50°C operation at full rated output power; derate output
current and total output power by 2.5% per °C above 50°C.
Commercial Safety
Approved to UL1950, CSA22.2 No. 950, IEC950, EN60950. All dc outputs are SELV
under normal and single fault conditions.
Medical Leakage Current
The maximum leakage current is as follows:
Test Condition
Normal Single
120 Vac @ 60 Hz input
25 μA
264 Vac @ 60 Hz input
40 μA
Fault
35 μA
60 μA
Medical Safety
Approved to UL2601-1, CSA22.2 No. 601 Level 3 and EN60601-1. UL file E116994;
CSA #LR46516. The output(s) are intended for safety earthed Signal Output
and Intermediate Circuits only. The output(s) are not acceptable for patient con-
nection without additional isolation. All dc outputs are SELV under normal and
single fault conditions.
All specifications are typical at nominal input, full load at 25
O
C unless otherwise stated
Regarding the irregular table issues related to PDA R&D display
Regarding the problem of displaying irregular tables on PDA R&D, there are a lot of data tables that need to be displayed on PDA. Each table and data is different. How to display them on PDA? I use vs...
wodi520 Embedded System
About AD sampling storage
A question about AD sampling storage: A 4-channel 16-bit AD sampling chip, the sampling frequency is set to 50KHZ, all the collected data is stored in a 16-bit RAM. The size of the RAM is known to be ...
tmmdh FPGA/CPLD
CCS error, please help me
CCS error, please help me...
dxx6253316 Microcontroller MCU
Why can't the D3DM sample programs run under WINCE60+S3C6410 using the d3dm_sedrv.dll provided by Samsung?
Why can't I run the D3DM sample programs under WINCE60+S3C6410 using d3dm_sedrv.dll provided by Samsung? I keep getting Invalid CPUID for D3DM Driver!!!...
purmoon Embedded System
VMM Image Measurement Software
Could you please tell me which cameras VMM image measurement software can support? Does it support driverless cameras?...
UUC Industrial Control Electronics
IAR 5.2 External Interrupts
Hey guys, please help me with a program. I am using the IAR5.2 compiler and I don't know how to write an external interrupt. I don't know if it works this way. Also, how do I add it when starting the ...
白菜 Embedded System

Technical ResourceMore

EEWorld
subscription
account

EEWorld
service
account

Automotive
development
circle

Robot
development
community

Index Files: 1592  1094  894  483  2126  33  23  19  10  43 
Datasheet   0 1 2 3 4 5 6 7 8 9 A B C D E F G H I J K L M N O P Q R S T U V W X Y Z
Room 1530, 15th Floor, Building B, No. 18 Zhongguancun Street, Haidian District, Beijing Telephone: (010) 82350740 Postal Code: 100190
Copyright © 2005-2026 EEWORLD.com.cn, Inc. All rights reserved 京ICP证060456号 京ICP备10001474号-1 电信业务审批[2006]字第258号函 京公网安备 11010802033920号