GPC130 Commercial/GPM130
Medical
130 Watt Global Performance Switchers
FEATURES:
• Wide-range ac input 85-264 Vac
• 2-year warranty
• Commercial Approved to UL1950, IEC950,
EN60950 and CSA22.2-234 L3
• Medical Approved to UL2601-1, IEC601-1 and
CSA22.2 No. 601
• Conducted EMI exceeds FCC Class B and CISPR 22
Class B (Commercial models) and CISPR 11 Class B
(Medical models)
•
marked to LVD
SPECIFICATIONS:
Ac Input
85-264 Vac, 47-63 Hz single phase.
Input Current
Maximum input current at 120 Vac, 60 Hz with full rated output
load: 3.5 A
Hold-Up Time
20 ms minimum from loss of ac input at full load, nominal line
(115 Vac).
Output Power
100 W continuous, 165 W peak; 130 W with 26 cfm air flow. Peak
ratings are for 60 s maximum duration, 10% duty cycle. During
peak load condition, output regulation may exceed total
regulation limits.
Overload Protection
Fully protected against short circuit and output overload. Short
circuit protection is cycling type power limit on outputs 1 & 2;
foldback type on outputs 3 & 4. Recovery after fault is automatic.
See output ratings chart for additional notes or conditions.
Overvoltage Protection
Main outputs: 124% ± 12% typical.
Efficiency
70% at full rated load, nominal input voltage, depending on model
and load distribution.
Turn-on Time
Less than 1 second at 120 Vac, 25°C (inversely proportional to
input voltage and thermistor temperature).
Input Protection
Internal ac fuse provided. Designed to blow only if a catastrophic
failure occurs in the unit.
Inrush Current
Inrush is limited by internal thermistors. Inrush at 240 Vac under
cold start conditions will not exceed 34 A.
Temperature Coefficient
0.03%/
°
C typical on all outputs.
Thermal Shutdown
Provided as a standard feature. Designed to protect unit from
prolonged overtemperature.
Environmental
Designed for 0 to 50°C operation at full rated output power;
derate output current and total output power by 2.5% per °C
above 50°C. See Environmental and Packaging Specifications
on next page.
Power Fail
TTL- or CMOS-compatible output goes low (< 0.5V) 5 ms
before output voltage drops more than 4% below nominal
voltage upon loss of ac power. The signal is factory set to
trip on 84 to 94Vac brown-out depending upon incoming line
impedance and distortion. Other settings are available to the
user through adjustment of built-in potentiometer.
Output Noise
0.5% rms, 1% pk-pk, 20 MHz bandwidth, differential mode.
Measured with noise probe directly across output terminals of
the power supply.
Transient Response
Main output—500µSec typical response time for return to within
0.5% of final value for a 50% load step change.
∆i/∆t<0.2
A/µs.
Maximum voltage deviation is 3.5%. Startup/shutdown overshoot
less than 3%.
Remote Sense
Provided as a standard feature on output 1.
Voltage Adjustment
Built-in potentiometer adjusts voltage ±5% on outputs 1 & 2.
Overload Protection
Factory set to begin power limiting at approximately 170 W.
EMI/EMC Compliance
All models include built-in EMI filtering to meet the following
emissions requirements:
EMI SPECIFICATIONS
COMPLIANCE LEVEL
Conducted Emissions GPC130 EN55022 Class B; FCC Class B
Conducted Emissions GPM130 EN55022 Class B; FCC Class B
Static Discharge
EN61000-4-2, 6 kV contact, 8 kV air
RF Field Susceptibility
EN61000-4-3, 3 V/meter
Fast Transients/Bursts
EN61000-4-4, 2 kV, 5 kHz
Surge Susceptibility
EN61000-4-5, 1 kV diff., 2 kV com.
Commercial Leakage Current
0.7 mA 254 Vac @ 60 Hz input.
Commercial Safety
Approved to UL1950, CSA22.2 No. 234 Level 3, IEC950 and
EN60950. UL file #E135803 commercial; CSA #LR46516 all
models. The output(s) are intended for safety earthed Signal
Output and Intermediate Circuits only. All dc outputs are SELV
under normal and single fault conditions.
Medical Leakage Current
35 µA 254 Vac @ 60 Hz input.
Medical Safety
Approved to UL2601, CSA22.2 No. 601 Level 3 and IEC601. UL
file E116994; CSA #LR46516. The output(s) are intended for
safety earthed Signal Output and Intermediate Circuits only. The
output(s) are not acceptable for patient connection without
additional isolation. All dc outputs are SELV under normal and
single fault conditions.
54
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
GPC130 Commercial/GPM130
Medical
130 Watt Multiple Output
Commercial
Model
GPC130A
Medical
Model
GPM130A
Output No.
1
2
3
4
1
2
3
4
1
2
3
4
1
2
3
4
Output
+5 V
+12 V
-12 V
+12 V
+5V
+12 V
-5 V
-12 V
+5 V
+12 V
-15 V
+15 V
+5 V
+24 V
-12 V
+12 V
Output
Minimum
3A
0A
0A
0A
3A
0A
0A
0A
3A
0A
0A
0A
3A
0A
0A
0A
Output
Maximum (B)
16 A
5A
1.2 A
1.2 A
16 A
5A
1.2 A
1.2 A
16 A
5A
1.2 A
1.2 A
16 A
3.5 A
1.2 A
1.2 A
Output
Maximum (C)
20 A
5A
1.2 A
1.2 A
20 A
5A
1.2 A
1.2 A
20 A
5A
1.2 A
1.2 A
20 A
3.5 A
1.2 A
1.2 A
Output Peak
25 A
8A
1.5 A
1.5 A
25 A
8A
1.5 A
1.5 A
25 A
8A
1.5 A
1.5 A
25 A
5A
1.5 A
1.5 A
Noise P-P
50 mV
120 mV
120 mV
120 mV
50 mV
120 mV
50 mV
120 mV
50 mV
120 mV
150 mV
150 mV
50 mV
240 mV
120 mV
120 mV
Total
Regulation (A)
2%
2%
3%
3%
2%
2%
3%
3%
2%
2%
3%
3%
2%
2%
3%
3%
2%
2%
3%
3%
GPC130B
GPM130B
GPC130C
GPM130C
GPC130D
GPM130D
1
+5 V
3A
16 A
20 A
25 A
50 mV
2
+24 V
0A
3.5 A
3.5 A
5A
240 mV
3
-15 V
0A
1.2 A
1.2 A
1.5 A
150 mV
4
+15 V
0A
1.2 A
1.2 A
1.5 A
150 mV
A.Total regulation is defined as the maximum deviation from the nominal voltage for all steady-state conditions of initial voltage setting, input line voltage and output load.
B. Unrestricted natural convection cooling.
C. Requires 26 cfm moving air.
GPC130E
GPM130E
GPC130/GPM130
MECHANICAL SPECIFICATIONS
Environmental
Specification
Temperature (A)
Humidity (A)
Shock (B)
Altitude
Vibration (C)
Operating
See individual specs
0 to 95% RH
20 g
pk
-500 to 10,000 ft
1.5 g
rms
, 0.003 g
2
/Hz
Non-operating
-40 to +85
o
C
0 to 95% RH
40 g
pk
-500 to 40,000 ft
5 g
rms
, 0.026 g
2
/Hz
A. Units should be allowed to warm up/operate under non-condensing conditions
before application of power.
B. Random vibration—10 to 2000Hz, 6dB/octave roll-off from 350 to 2000Hz, 3
orthogonal axes. Tested for 10 min./axis operating and 1 hr./axis non-operating.
C. Shock testing—half-sinusoidal, 10 ± 3 ms duration, ± direction, 3 orthogonal axes,
total 6 shocks.
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
55