GLC65 Commercial/GLM65 Medical
MULTIPLE OUTPUT GLOBAL PERFORMANCE SWITCHERS
Summary:
• 65 Watt Wide Range Input
• High Efficiency (Up to 90%)
• Universal input 90-264 Vac
• 2-year warranty
• Also available in single outputs
• Conducted EMI complies with FCC Class B and
CISPR 22 Class B (Commercial models) and
CISPR 11 Class B (Medical models)
• Exempt from line harmonics standard EN61000-3-2
• Commercial Approved to UL1950, CSA-C22.2 No. 950,
• Medical Approved to UL2601-1, CSA 22.2 No. 601.1, and
EN60601-1• Medical Approved to UL2601-1, CSA 22.2 No. 601.1,
• RoHS Compliant (G suffix)
•
marked to LVD
65 Watt Multiple Output Global Performance Switchers
SPECIFICATIONS
Ac Input
90-264 Vac, 47-63 Hz single phase.
Output Power
Total continuous output power is 60 W, 75 W peak for 60 s 10% duty cycle.
Total continuous rating with 150 LFM of air is 70 W.
Input Current
Units are exempt from input current harmonic requirements of EN61000-3-2
when output power is less than 62W. Maximum input current at minimum
input voltage and output overload will be less than 2.0 A.
Inrush Current
Inrush is limited by internal thermistor. The inrush at 240 Vac, averaged over
the first ac half-cycle under cold start conditions will not exceed 37 A.
Input Protection
Internal AC fuse provided on all units. Designed to blow only if a catastrophic
failure occurs in the unit -- Fuse does not blow on short circuit or unsustained
overload.
Holdup Time
Output voltage stays within regulation for 20 ms from loss of ac input at 65 W
load, from 120 Vac input.
Efficiency
80% minimum.
Overload Protection
Fully protected against short circuit and output overload. Short circuit
protection is cycling type power limit and will automatically recover after
removal of fault.
Overvoltage Protection
Built in with firing point set. OVP firing reduces output to less than 50% of
nominal voltage in less than 50 ms.
Output Noise
0.5% rms, 1% pk-pk, 20 MHz Bandwidth, differential mode. Measured with
scope probe directly across output terminals of the power supply with load
terminated with 1μF capacitor.
Transient Response
Main Output - 500 μs typical response time for return to within 0.5% of final
value for a 50% load step within the regulation limits of minimum and maxi-
mum load, ∆i/∆t< 0.2 A/μs. Maximum voltage deviation is 3.5%.
Minimum Load
No minimum load required on any output. However, regulation limits may be
exceeded if extreme conditions are applied. Contact factory for assistance.
Temperature Coefficient: 0.03%°C typical on all outputs.
Voltage Adjustment
Units provided with one potentiometer capable of adjusting all 3 outputs a mini-
mum of +/- 5% from nominal setting.
EMI / EMC Compliance:
All models include built-in EMI filtering to meet EMC requirements of IEC601-1-2.
PERFORMANCE
Conducted Emissions GLC65
Conducted Emissions GLM65
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Surge Susceptibility
Conducted RF Susceptibility
Voltage Sags & Surges
Line Frequency Harmonics
EMC STANDARD
EN55022 Class B; FCC Class B
EN55011 Class B; FCC Class B
EN61000-4-2, Level 3
EN61000-4-3, Level 3
EN61000-4-4, Level 3
EN61000-4-5, Level 3
EN61000-4-6, 3V, 80%
EN61000-4-11
EN61000-3-2 Class A (>62Wout)
Exempt (<62Wout)
GLC65 Commercial - Safety
SL Power Electronics Corp declares under our sole responsibility that all models
are in conformity with the applicable requirements of EN60950 following the
provisions of the Low Voltage Directive 73/23/EEC. All models are approved to
UL1950 (with no D3 deviations), CSA22.2 No. 950-95 3rd edition, Level 3, IEC950,
EN60950. CB certificate available.
Commercial Leakage Current
Less then 500 μA @ 120 Vac. Less then 1.2 mA @ 240 Vac.
GLM65 Medical - Safety
Approvals: All models are certified to be in compliance with the applicable require-
ments of UL2601-1, CSA 22.2 No. 601.1, IEC 601-1 (1988), EN 60601-1: 1990.
Medical Leakage Current
80 μA under normal conditions (132 Vac @ 60 Hz). Maximum under single fault
conditions (264 Vac @ 60 Hz) is 180 μA.
All specifications are typical at nominal input, full load at 25
O
C unless otherwise stated
Commercial
Model
GLC65A
Medical
Model
GLM65A
Output No.
1
2
3
1
2
3
1
2
3
1
2
3
1
2
3
1
2
3
Output
+5.0 V
+12 V
-12 V
+5.1 V
+15 V
-15 V
+5.0 V
+24 V
-12 V
+5.0 V
+24 V
+12 V
+5.0 V
+3.3 V
+12 V
+3.3 V
+5.0 V
+12 V
Output
Maximum (A)
7A
3A
2.5 A
7A
2.5 A
2A
7A
1.5 A
2.5 A
7A
1.5 A
2.5 A
5A
4A
2.5 A
5A
4A
2.5 A
Output
Maximum (B)
9A
5 A
4A
9A
4A
3A
9A
2.5 A
4A
9A
2.5 A
4A
8A
5A
4A
8A
6A
3A
Total
Regulation (E)
2%
5%
6%
2%
4%
5%
2%
3%
6%
2%
3%
6%
3%
3%
6%
3.5%
4.5%
6%
V1 Adjustment
Note (C)
± 5%
—
—
± 5%
—
—
± 5%
—
—
± 5%
—
—
± 5%
—
—
± 5%
—
—
OVP Set-
point
6.2 ± 0.6 V
—
—
6.2 ± 0.6 V
—
—
6.2 ± 0.6 V
—
—
6.2 ± 0.6 V
—
—
6.2 ± 0.6 V
—
—
4.3 ± 0.8 V
—
—
Ripple/
Noise
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
GLC65B
GLM65B
GLC65D
GLM65D
GLC65E
GLM65E
GLC65G
GLM65G
GLC65H
GLM65H
Notes:
A. Continuous individual output ratings for unrestricted convection cooling. Combination of individual output loads must not exceed total power rating.
B. Peak rating for 60 s 10% duty cycle or continuous rating 150 LFM forced air cooling.
C. Adjustment on V1 varies all outputs simultaneously (1% on V1 @ 1% on V2 & V3).
D. Total combined current of V1 & V2 not to exceed 12 A. Combination of individual output loads must not exceed total power rating.
E. Total regulation is defined as maximum deviation from the initial set point. With all other outputs at 50% load, output under test can be varied from 0 to 100% load
and varied to any ac line voltage. Initial set point is 1% on V1, 2% on V2 and 3% on V3
F. Add “G” suffix to model number for RoHS compliant model.
GLC65 MECHANICAL SPECIFICATIONS
1.20
30.28
ENVIRONMENTAL SPECIFICATIONS
Temperature (A)
Humidity (A)
Shock (B)
Altitude
Vibration (C)
OPERATING
0 TO 50°C
0 to 95% RH
20 g
pk
-500 to 10,000 ft
1.5 g
rms’
0.003 g
2
/Hz
NON-OPERATING
-40 to +85°C
0 to 95% RH
40 g
pk
-500 to 40,000 ft
5 g
rms’
0.026 g
2
/Hz
A. Units should be allowed to warm up/operate under non-condensing con-
ditions before application of power. Derate output current and total output
power by 2.5% per °C above 50°C. For operation in a confined space, moving
air may be required. Under all conditions, the cooling vs. load profile should
be such that heat sinks and/or heatsink temperatures do not exceed 90 °C
for extended periods.
B. Shock testing—half-sinusoidal, 10 ± 3 ms duration, ± direction, 3 orthogo-
nal axes, total 6 shocks.
C. Random vibration—10 to 2000Hz, 6dB/octave roll-off from 350 to 2000Hz,
3 orthogonal axes. Tested for 10 min./axis operating and 1 hr./axis non-op-
erating.
SL Power Electronics Corp., 6050 King Drive, Bldg. A, Ventura, CA 93003, USA. Phone:(805) 486 4565 Fax:(805) 487 8911 www.slpower.com Rev. 09/10/08. Data Sheet ©
2008 SL Power Electronics Corp. The information and specifications contained in this data sheet are believed to be correct at time of publication.
However, SL Power accepts no responsibility for consequences arising from reproduction errors or inaccuracies. Specifications are subject to change without notice.