NPC-100 and
NPC-120
Series
NovaSensor
Disposable Medical
Pressure Sensors
Features
• Solid state, high reliability
• Media compatibility
• High performance
•
Factory filled with dielectric gel
• Small size
• Fully tested
• Temperature compensated
•
Low cost disposable design
•
Designed to AAMI specifications
Applications
• Medical instrumentation
• Blood pressure measurement
• Infusion pumps
• Kidney dialysis machines
Amphenol
Advanced Sensors
NPC-100 and NPC-120
Series Specifications
The NovaSensor NPC-100 and NPC-120 Series pressure
sensor are specifically designed for use in disposable
medical applications. The devices are compensated and
calibrated per the Association for the Advancement of
Medical Instrumentation (AAMI) guidelines for industry
acceptability. The sensors integrate a high-performance,
pressure sensor die with temperature compensation
circuitry and gel protection in a small, low-cost package.
The NPC-100 and NPC-120 Series are manufactured
in a class 1000 clean-room to minimize possible
sources of contamination. A specially designed silicon
micromachined sensing element is used to meet or
exceed all industry requirements while minimizing
assembly and test cost for maximum customer
value. Thick-film laser-trimming is employed for final
compensation and calibration. Sensitivity is maintained to
±1% and linearity is better than 1% in the physiological
operating pressure range.
The NPC-100 and NPC-120 Series are batch-
manufactured in ceramic plate form and shipped as
an intact array or in Tape and Reel for easy customer
automation. This assembly method draws from well-
established manufacturing techniques used in the
electronics industry in order to produce a quality, high
volume product.
0.0875 in to 0.0905 in
(2.22 mm to 2.30 mm)
0.049 in to 0.052 in
(1.24 mm to 1.32 mm)
0.393 in to 0.397 in
(9.98 mm to 10.08 mm)
0.392 in to 0.397 in
(9.96 mm to 10.08 mm)
Parameter
General
Pressure Ranges
Overpressure
Value
Units
Notes
–50 to 300
125
mmHg
psi
–0.97 psi to 5.8 psi
minimum
Electrical @ 72°F (22°C) Unless Otherwise Stated
Input Excitation
Dielectric Breakdown
Risk Current
Input Impedance
Output Impedance
Environmental
Temperature
Compensated
Operating
Storage
Humidity
Light Sensitivity
Operating Product Life
Shelf Life
Mechanical
Weight
Media Interface
Gel Tube Interface
Material
Parameter
Performance Parameters*
Offset
Sensitivity
Calibration
Symmetry
Linearity (–30 to 100 mmHg)
Linearity (100 to 200 mmHg)
Linearity (200 to 300 mmHg)
Thermal Coefficient Offset
Thermal Coefficient Span
Frequency Response
Phase Shift
Offset Drift
mmHg
µV/V/mmHg
mmHg
%
mmHg
% output
% output
mmHg/°C
%/°C
Hz
degrees
mmHg/8 hrs
–20
4.95
97.5
—
—
—
—
—
—
1200
—
—
0
5
100
—
—
—
—
—
—
—
—
—
20
5.05
102.5
±5
1
1
1.5
±0.3
±0.1
—
5
1
6
6
6
3
3
5
5
4,5
2
<0.0044
lb
(<2 g)
15°C to 40°C
15°C to 40°C
–25°C to 70°C
10 to 90
1
168
5
°C
°C
°C
%
mmHg
hours
years
maximum (per AAMI BP22)
(59° to 104°F)
(59° to 104°F)
(–13° to 158°F)
1 to 10
10,000
2
1800 to 3300
285 to 315
VDC
VDC
µA
Ω
Ω
Calibrated for 6 VDC
5
Maximum (per AAMI), 5
Medical grade, dielectric gel
Polycarbonate
Units
Minimum
Type
Maximum
Notes
0.236 in to 0.246 in
(5.99 mm to 6.25 mm)
0.189 in to 0.192 in
(4.80 mm to 4.88 mm)
Radius 0.005 in
(0.13 mm) maximum
0.142 in to 0.145 in
(3.61 mm to 3.68 mm)
Material is clear Polycarbonate.
0.282 in to 0.290 in
(7.16 mm to 7.37 mm)
NPC-100 Series pressure sensor port interface
0.0875 in to 0.0905 in
(2.22 mm to 2.30 mm)
0.049 in to 0.052 in
(1.24 mm to 1.32 mm)
0.124 in to 0.126 in
(3.14 mm to 3.20 mm)
Radius 0.005 in
(0.13 mm) maximum
0.142 in to 0.145 in
(3.61 mm to 3.68 mm)
Material is clear Polycarbonate.
0.282 in to 0.290 in
(7.16 mm to 7.37 mm)
0.236 in to 0.246 in
(5.99 mm to 6.25 mm)
* 1. All values measured at 6 VDC and 71.6°F (22°C) and after five second warm up unless
otherwise specified.
2. Output of sensor with no pressure applied and a 150 kΩ resistor shorted across +IN to +OUT.
3. Over a temperature range of 59°F to 104°F (15°C to 40°C).
4. Normalized offset/bridge voltage—8 hours after 20 seconds. warm-up.
5. Previously qualified, not tested in production.
6. Deviation from straight line drawn through zero and 100 mmHg data points.
7. These sensors are compatible with ETO (ethylene oxide) sterilization. The user is responsible for
establishing and validating suitable process conditions to prevent damage to the sensor.
–
0.160 in to 0.170 in
(4.06 mm to 4.32 mm)
Pressure
port
Dielectric gel
Vent hole
Polycarbonate
gel tube2
Contact pads
0.320 in +0.010/-0.002
(8.13 mm +0.25/-0.05)
0.415 in +0.010/-0.002
(10.54 mm +0.25/-0.05)
1
Gel height
+0.001 in/-0.035 in
(+0.025 mm/-0.89 mm)
NPC-120 Series pressure sensor port interface
0.109 in to 0.119 in
(2.77 mm to 3.02 mm)
1
0.017 in
(0.43 mm)
0.010 in
(0.25 mm)
1. Tolerances reflect typical flaring during the singulation process.
NPC-100 Series package diagram
+In
Pinout
1 +In
2 +Out
3 -Out
4 -In
+Out
-Out
.069 in (1.75 mm)
1
2
3
4
0.000 in (0 mm)
1. Tolerance on electrical pad location is ±0.010 in (±0.25 mm).
2. Contact pads are suitable for soldering.
3. Split pads on +Out and –Out must be connected for proper operation.
4. All dimensions assume a nominal 0.01 in (0.25 mm) distance between the
edge of the solder pads to the edge of the ceramic. This distance will vary
slightly from part to part.
NPC-100 Series electrical interface
4.35 in (110.5 mm)
0 in (0 mm)
4.35 in (110.5 mm)
0 in (0 mm)
1. Sensors are shipped as 120 UP snapstrates that must be singulated by
the
customer.
2. Each reel or plate may include units that have failed visual or electrical
parameters as well as good units. Bad units are identified with a dot on the
backside of the cell location.
3. Reels and plates are shipped in dust free anti-static containers to
prevent contamination of the gel surface.
NPC-100 Series shipping
Configuration
0.262 in (6.65 mm)
0.152 in (3.86 mm)
0.168 in (4.27 mm)
0.201 in (5.10 mm)
0.223 in (5.66 mm)
0.097 in (2.46 mm)
0.119 in (3.02 mm)
0.058 in (1.47 mm)
0.000 in (0 mm)
NPC-100
NPC-120
Code
T
NPC-120 - T
-In
NPC-100 Series schematic diagram
Ordering Information
The code number to be ordered may be specified as follows:
Blank: Sold in 120-Up Snapstrate
Tape and reel
Typical model number
Storage Conditions
After the original packaging has been opened, reels and
trays should be stored in a nitrogen dry box or moved to a
clean room for assembly. The storage environment should
be free of: sulfur bearing materials or other corrosive
compounds such as paper products, card board, acid
batteries, elastomer or rubber products, vulcanized
rubber, adhesives or plastics that release sulfur over time.
Warranty
NovaSensor warrants its products against defects in material and workmanship for 12 months from the date of
shipment. Products not subjected to misuse will be repaired or replaced. NovaSensor reserves the right to make
changes without further notice to any products herein. NovaSensor makes no warranty, representation or guarantee
regarding the suitability of its products for any particular application. NovaSensor does not assume any liability arising
out of the application or use of any product or circuit and specifically disclaims, and all liability, without limitation
consequential or incidental damages. The foregoing warranties are exclusive and in lieu of all other warranties, whether
written, oral, implied or statutory. No implied statutory warranty or merchantability or fitness for particular purpose shall
apply.
Amphenol
Advanced Sensors
www.amphenol-sensors.com
© 2018 Amphenol Corporation. All Rights Reserved. Specifications are subject to change without notice.
Other company names and product names used in this document are the registered trademarks or
trademarks of their respective owners.
AAS-920-302G