GPFM600 Medical
600 Watt Global Performance Switchers
FEATURES:
•
•
•
•
•
Small package (12.00” x 5.50” x 3.09”)
Power factor corrected to EN61000-3-2 Class A
Full load rating to 60°C
85 to 264 Vac Input
Medically Approved to UL2601-1, IEC601-1, CSA
C22.2 No. 601.1.2
• 2 year warranty
•
marked to LVD
SPECIFICATIONS:
Ac Input
85-264 Vac, 47-63 Hz single phase.
Output Power
Continuous output power, 600 watts.
Hold-upTime
20 ms minimum from loss of ac input until output reaches
3% below nominal.
Line and Load Regulation
±1%
over the entire input voltage and frequency range,
and over the entire output current range.
Minimum Load
No minimum load required.
Overload Protection
Current limit 105% minimum, 125% maximum of the
specified full load current.
Output Noise
50 mV pk-pk maximum.
Transient Response
Maximum of 5% change in Vout for a load change of 1.25 A.
Recovers to ±1% within 500u sec.
Overvoltage Protection
Built in on all models. OVP crowbar reduces output
voltage below nominal rating in less than 50 ms. Output
voltage decay is dependent upon loading.
DC Power Good
Optocoupler isolated output indicates that the DC output
is equal to or greater than 21.6 (-0 + 0.25) volts. This
signal shall be driven when in the active ‘DC Good’ state.
Efficiency
Minimum of 75% under full load and input of 120 Vac 60 Hz.
Inrush Current
Limit inrush current to 125% of maximum operating
current. Turn on time of the power supply not to exceed 3
seconds over the full input voltage and frequency range.
EMI/EMC Compliance
All models include built-in EMI filtering to meet emissions
requirements:
EMI SPECIFICATIONS
Conducted Emissions
Static Discharge
RF Field Susceptibility
Fast Transients/Bursts
Surge Susceptibility
COMPLIANCE LEVEL
EN55011 Class B; FCC Class B
EN61000-4-2, 6 kV contact, 8 kV air
EN61000-4-3, 3 V/meter
EN61000-4-4, 2 kV, 5 kHz
EN61000-4-5, 1 kV diff., 2 kV com.
Vibration and Shock
All models are to meet MIL-STD 810E, method 514.4
Category.
Environmental
Designed for 0 to 60°C operation at full rated output
power; derate output current and total output power by
2.5% per °C above 50°C.
Medical Leakage Current
150 µA under normal conditions. Maximum under single
fault conditions, 250 µA. Patient sink leakage with 120
Vac, 60 Hz applied to any output with respect to the
chassis will be less than 500 µA.
Medical Safety Agency
Approved to UL2601-1, EN60601-1-2, CSA C22.2 No.
601.1.2.
112
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
GPFM600 Medical
600 Watt Global Performance Switchers
Commercial Model
GPFM600-24
Output
24 Vac
Output
Maximum(A)
25 A
OVP Setpoint
25.8 ± 0.6 V
Total Regulation
1%
Noise P-P
100 mV
GPFM600
MECHANICAL SPECIFICATIONS
INPUT: J1
MOLEX P/N 43160-31-3
PIN 1: AC LINE
PIN 2: AC NEUTRAL
PIN 3: EARTH GROUND
OUTPUT: J2
MOLEX P.N. 39-29-1147
14 PIN RIGHT ANGLE HEADER
PIN 1: +24 V
PIN 2: +24 V
PIN 3: +24 V RETURN
PIN 4: +24 V RETURN
PIN 5: N/A
PIN 6: AC GOOD
PIN 7: AC GOOD RETURN
PIN 8: +24 V
PIN 9: +24 V
PIN 10: +24 V RETURN
PIN 11: +24 V RETURN
PIN 12: N/A
PIN 13: DC GOOD
PIN 14: DC GOOD RETURN
Environmental
Specification
Temperature (A)
Humidity (A)
Shock (B)
Altitude
Vibration (C)
Operating
0 to 60ºC
0 to 97% RH
20 g
pk
-500 to 10,000 ft
1.5 g
rms
, 0.003 g
2
/Hz
Non-operating
Dimensions:
-18 to +60ºC
0 to 97% RH
40 g
pk
-500 to 40,000 ft
5 g
rms
, 0.026
g
2
/Hz
Inches
Millimeters
A. Units should be allowed to warm up/operate under non-condensing conditions
before application of power. Cooling provided by internal fan— heatsink
temperatures should not exceed 90ºC for extended periods in the installation.
B. Random vibration—10 to 2000Hz, 6dB/octave roll-off from 350 to 2000Hz, 3
orthogonal axes. Tested for 10 min./axis operating and 1 hr./axis non-operating.
C. Shock testing—half-sinusoidal, 10 ± 3 ms duration, ± direction, 3 orthogonal axes,
total 6 shocks.
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
113