NFS110 Medical Series
Single and quad output
LOW TO MEDIUM POWER AC/DC POWER SUPPLIES
•
•
•
•
•
•
7.0 x 4.25 x 1.8 inch package
Medical, dental and laboratory applications
Overvoltage and short circuit protection
110W with 20CFM
UL, VDE and CSA safety approvals
EN60601-1 and UL2601 medical approvals
80-110W AC/DC Universal Input Switch Mode Power Supplies
1
The NFS110 medical series is a 80 to 110W universal input AC/DC power supply on a 7
x 4.25 inch card with a maximum component height of 1.8 inches for use in medical
applications. The NFS110 medical series has the same generic feature set as the
standard NFS110 series but have been designed with lower safety ground leakage and
higher isolation as required for medical safety approval. The NFS110 provides 80W of
output power with free air convection cooling which can be boosted to 110W with
20CFM of air. Standard features include overvoltage and short-circuit protection. The
series, with full medical safety approval to EN60601 and UL2601, meets conducted
emissions EN55022 level A. The NFS110 medical series is designed for use in low
power medical, dental and laboratory applications such as dialysis machines,
monitoring equipment, instrumentation and infusion pump controls.
2 YEAR WARRANTY
All specifications are typical at nominal input, full load at 25°C unless otherwise stated
SPECIFICATIONS
OUTPUT SPECIFICATIONS
Voltage adjustability
+5.1V output on multi’s
5.1V single
12V single
15V single
24V single
LL to HL, FL
All outputs on all units
At turn-on no load
All outputs
Multi output 5.1V only
5.1V single
12V single
15V single
24V single
Primary power
limited
±3.0%
±3.0%
12V to 14V
15V to 18V
24V to 30V
±0.1% max.
0%
±0.02%/°C
6.25V±0.75V
6.25V±0.75V
15.75V±1.0V
22V±1.5V
33V±2.5V
Pin max. 160W
Pout min. 110W
Burst mode operation
GENERAL SPECIFICATIONS
Hold-up time
110VAC
110VAC
230VAC
230VAC
@
@
@
@
80W
110W
80W
110W
35ms
17ms
140ms
100ms
70%
70%
72%
75%
typ.
typ.
typ.
typ.
Efficiency
Line regulation
Overshoot/undershoot
Temperature coefficient
Overvoltage protection
Multiple outputs
+5.1V single
12V and 15V singles
24V single
Input/output
Input/chassis
At 100 Watts output
At zero load
Isolation voltage
Switching frequency
Approvals and
standards
(See Note 12)
Weight
MTBF
4000VAC
1500VAC
20 to 70kHz
100 to 250kHz
Output power limit
Short circuit protection
INPUT SPECIFICATIONS
VDE0750, IEC60601
IEC1010, UL2601
CSA C22.2 No. 125
Singles
Multiple outputs
MIL-HDBK-217E
550g (19.4oz)
600g (21.2oz)
125,000 hours
ENVIRONMENTAL SPECIFICATIONS
Input voltage range
Input frequency range
Input surge current
Safety ground
leakage current
110VAC, 50Hz
230VAC, 50Hz
132VAC
264VAC
90 to 253VAC
127 to 357VDC
47Hz to 440Hz
17A
35A
50µA
100µA
Thermal performance
Operating, see curve
0°C to +70°C
Non-operating
–40°C to +85°C
0°C to +50°C,
80W
amb. convection cooled
+50°C to +70°C,
Derate 2W/°C
amb. convection cooled
0°C to +50°C,
110W
20CFM forced air
+50°C to +70°C,
Derate
20CFM forced air
2.75W/°C
Peak, 0°C to +50°C,
110W
max. 60 seconds
Non-condensing
Operating
Non-operating
5Hz to 500Hz
5% to 95% RH
10,000 feet max.
40,000 feet max.
2.4G approx.
EMC CHARACTERISTICS
Conducted emissions
Radiated emissions
ESD air
ESD contact
Surge
Fast transients
Radiated immunity
Conducted immunity
EN55022, FCC part 15
EN55022, FCC part 15
EN61000-4-2, level 3
EN61000-4-2, level 4
EN61000-4-5, level 3
EN61000-4-4, level 3
EN61000-4-3, level 3
EN61000-4-6, level 3
Level A
Level A
criteria 1
criteria 1
criteria 1
criteria 1
criteria 2
criteria 2
Relative humidity
Altitude
Vibration (See Note 11)
Perf.
Perf.
Perf.
Perf.
Perf.
Perf.
File Name: nfs110m.pdf Rev: 29 Aug 2005
NFS110 Medical Series
Single and quad output
LOW TO MEDIUM POWER AC/DC POWER SUPPLIES
80-110W AC/DC Universal Input Switch Mode Power Supplies
2
For the most current data and application support visit www.artesyn.com/powergroup/products.htm
OUTPUT
VOLTAGE
+5.1V
+12.0V
–12.0V
–5.0V
+5.1V (V
A
)
+24.0V (V
B
)
+12.0V
–12V
+5.1V
+15.0V
-15V
-5V
+5.1V
12V
15V
24V
Notes
1
2
3
4
5
Convection cooled, 80W maximum.
Peak outputs lasting less than 60 seconds with duty cycle less than 10%.
Total peak power must not exceed 110W.
Forced air, 20CFM at 1 atmosphere, 110W maximum.
Figure is peak-to-peak. Output ripple is measured across a 50MHz
bandwidth using a 12 inch twisted pair terminated with a 47µF capacitor.
Total regulation is defined at the static output regulation at 25°C, including
initial tolerance, line voltage within stated limits and output voltages
adjusted to their factory settings. Also for NFS110-7902P, for 24V output
stated regulation I
A
/ I
B
≤ 5. This output will maintain
±
5.0% regulation if
I
A
≤ 5A, where I
A
= +5.1V output current and I
B
= +24V output current.
Single output models have floating outputs which may be referenced as
either positive or negative. Higher voltage supplies, may be adjusted over
a wide output voltage range, as long as the total output power does not
exceed 80 Watts (natural convection) or 110 Watts (forced air).
Power fail detect not available on single output models.
Derating curve is application specific for ambient temperatures > 50°C, for
optimum reliability no part of the heatsink should exceed 90
o
C and no
semiconductor case temperature should exceed 100
o
C.
Caution: Allow a minimum of 1 second after disconnecting the power when
making thermal measurements.
The user should read the PSU installation instructions in conjunction with
the relevant national safety regulations in order to ensure compliance.
Three orthogonal axes, random vibration, 10 minute test for each axis.
This product is only for inclusion by professional installers within other
equipment and must not be operated as a stand alone product.
NOTICE: Some models do not support all options. Please contact your
local Artesyn representative or use the on-line model number search tool at
http://www.artesyn.com/powergroup/products.htm to find a suitable
alternative.
(6)
OUTPUT CURRENTS
MAX
(1)
RIPPLE
FAN
(3)
(4)
PEAK
(2)
TOTAL
REGULATION
±2.0%
±3.0%
±3.0%
±3.0%
±2.0%
+10/–5.0%
±3.0%
±3.0%
±2.0%
±3.0%
±3.0%
±3.0%
±2.0%
±2.0%
±2.0%
±2.0%
(5)
MODEL NUMBERS
(13)
NFS110-7901P
8.0A
4.5A
0.5A
0.5A
8.0A
3.5A
4.5A
0.5A
8.0A
4.0A
0.5A
0.5A
16.0A
7.0A
5.0A
3.5A
20.0A
9.0A
1.5A
1.5A
20.0A
4.5A
9.0A
1.5A
20.0A
7.5A
1.5A
1.5A
22.0A
9.0A
7.3A
4.5A
10.0A
5.0A
1.0A
1.0A
10.0A
4.5A
5.0A
1.0A
10.0A
5.0A
1.0A
1.0A
20.0A
9.0A
7.3A
4.5A
50mV
120mV
120mV
50mV
50mV
240mV
120mV
120mV
50mV
150mV
150mV
50mV
50mV
120mV
150mV
240mV
TRANSIENT RESPONSE
NFS110-7901P
NFS110-7902P
NFS110-7904P
NFS110-7905
NFS110-7912
NFS110-7915
NFS110-7924
(6,7)
(6,7)
(6,7)
(6,7)
(6)
(6)
(6)
+5.1V (7.5A to 10A)
+12V (2.5A to 5A)
-12V (0.5A to 1A)
-5V (0.5A to 1A)
150mV peak,
1ms recovery
100mV peak,
0.5ms recovery
100mV peak,
0.5ms recovery
100mV peak,
0.5ms recovery
150mV peak,
1ms recovery
100mV peak,
0.5ms recovery
100mV peak,
0.5ms recovery
300mV peak,
1ms recovery
150mV peak,
1ms recovery
100mV peak,
0.5ms recovery
100mV peak,
0.5ms recovery
100mV peak,
0.5ms recovery
250mV peak,
1ms recovery
360mV peak,
1ms recovery
450mV peak,
1ms recovery
720mV peak,
1ms recovery
NFS110-7902P
+5.1V (7.5A to 10A)
+12V (2.5A to 5A)
-12V (0.5A to 1A)
24V (1.5A to 3A)
6
7
8
9
10
11
12
13
NFS110-7904P
+5.1V (7.5A to 10A)
+15V (2.5A to 5A)
-15V (0.5A to 1A)
-5V (0.5A to 1A)
NFS110-7905
NFS110-7912
NFS110-7915
NFS110-7924
+5.1V (10A to 20A)
+12V (4.5A to 9A)
+15V (3.65A to 7.3A)
+24V (2.25A to 4.5A)
File Name: nfs110m.pdf Rev: 29 Aug 2005
NFS110 Medical Series
Single and quad output
LOW TO MEDIUM POWER AC/DC POWER SUPPLIES
80-110W AC/DC Universal Input Switch Mode Power Supplies
3
For the most current data and application support visit www.artesyn.com/powergroup/products.htm
AC (J1) mating connector
Molex 09-50-3051 or Molex 09-91-0500 mating connector with 2478 or
equivalent crimp terminals.
AC INPUT
OPTIONAL POWER FAIL DETECT
TIMING DIAGRAM
DC (J2) mating connector
Molex 09-50-3131 or Molex 09-91-1300 mating connector with 2478 or
equivalent crimp terminals.
5V OUTPUT
4.75V
4.75V
HIGH
3.5V MIN
PFD SIGNAL
LOW
0-0.4 MAX
T3
T1
T2
Power fail detect signal (See Note 7)
50ms ≤ T1 ≤ 200ms
T2 will vary with line and load
T3 ≥ 3ms
Pout: 110W
PFD output is an open collector
which will sink ≤ 40mA in the low state
0.475
±0.010
(12.07)
DERATING CURVE
(See Notes 8, 9)
Output Power (Watts)
1.80 (45.72)
MAX
0.050
±0.020
(1.27)
±0.005
(25.4)
1.55
±0.005
(39.37)
1.00
110W
20 CFM FORCED AIR COOLING
80W
3.35
±0.015
(85.09)
6-32 UNC
(4PL)
PIN 1
C
Voltage Adjust
Optional
Power Fail
Detect Circuit
(Key = Pin 2)
J2
4.250
±0.020
(107.95)
3.750
±0.005
(95.25)
T1
Pin 1
L1
J1
C6
F1
PIN 1
0.156
DIA. HOLE
(4PL)
NATURAL
CONVECTION
COOLING
55W
40W
11W
0W
11W MINIMUM LOAD REQUIRED AT 230VAC
0.250
±0.010
(6.35)
0.250
±0.010
(6.35)
6.50
±0.005
(165.1)
7.00
±0.020
(177.8)
5A, 250 VAC
0
°
C 10
°
C
20
°
C
30
°
C
40
°
C
50
°
C
60
°
C
70
°
C
PIN CONNECTIONS
ALL DIMENSIONS IN INCHES (mm)
J1
Mechanical Notes
A
Metallic or non-metallic stand-offs (maximum diameter 5.4mm) can be
used in all four mounting holes without affecting safety approval.
B
The ground pad of the mounting hole near J1 allows system grounding
through a metal stand-off to the system chassis.
C
The heatsink is grounded, and allows system grounding by mechanical
connection to the system chassis.
D
The supply must be mechanically supported using the PCB mounting holes
and may be additionally supported by the heatsink mounting holes.
E
It is always advisable to attach the power supply heatsink to another
thermal dissipator (such as a chassis or finned heatsink etc). The resulting
decrease in heat sink mounted component temperatures will improve
power supply lifetime.
F
A standard L-bracket and cover is available for mounting which contains all
screws connectors and necessary mounting hardware. Two different kits
are available, order part number ‘NFS110 COVER KIT’ or ‘NFS110C’.
–7901P
AC Ground
AC Neutral
AC Line
–7902P
AC Ground
AC Neutral
AC Line
–7904P
AC Ground
AC Neutral
AC Line
SINGLES
AC Ground
AC Neutral
AC Line
Pin 1
Pin 2
Pin 3
J2
Pin 1
Pin 2
Pin 3
Pin 4
Pin 5
Pin 6
Pin 7
Pin 8
Pin 9
+5.1V
+5.1V
+5.1V
Return
Return
Return
Return
+12V
+12V
PFD
–12V
+5.1V
+5.1V
+5.1V
Return
Return
Return
Return
+12V
+12V
PFD
–12V
+5.1V
+5.1V
+5.1V
Return
Return
Return
Return
+15V
+15V
PFD
–15V
V
out
V
out
V
out
Return
Return
Return
Return
V
out
V
out
N/C
N/C
International Safety Standard Approvals
VDE0750/EN60601-1/IEC601/IEC1010 File No. 10401-3336-1049
licence No. 2874
UL2601 File No. E147937
CSA C22.2 No. 125 File No. LR41062C
Pin 10
Pin 11
Pin 12
Pin 13
Removed for Key
–5V
+24V
–5V
N/C
N/C = no connection.
Datasheet © Artesyn Technologies® 2005
The information and specifications contained in this datasheet are believed to be correct at time of publication. However, Artesyn Technologies accepts no responsibility for consequences arising
from printing errors or inaccuracies. Specifications are subject to change without notice. No rights under any patent accompany the sale of any such product(s) or information contained herein.
www.artesyn.com
File Name: nfs110m.pdf Rev: 29 Aug 2005