GLD140 Gold Performance Medical Switchers
140 Watt Multiple Output
FEATURES:
Compact size (4.5" x 8.5" x 1.95")
Power factor corrected to IEC 1000-3-2 Class A
Documented FMEA, WCA and EMC test results
EMI compliance to CISPR11, FCC Class B, IEC601-1-2
Approved to UL2601, IEC601-1 and
CSA 22.2 No. 601.1-M90 and EN60601: 1990
•
marked to LVD
•
•
•
•
•
SPECIFICATIONS:
Ac Input
90-264 Vac, 47-63 Hz single phase.
Output Power
Continuous output power 140 W with unrestricted convection
cooling, 180 W with 150 LFM of air.
Input Current
Active Power Factor Correction circuitry assures compliance
with IEC1000-3-2, Class A. Maximum input current at minimum
input voltage and full load is 3.0 A.
Efficiency
73 – 85 % at full rated load, nominal input voltage, depending on
model and load distribution.
Hold-up Time
26 ms total. 16 ms minimum from loss of ac input at full load until
activation of the power fail signal. Output voltages remain within
specified regulation limits for an additional 9 ms minimum after
power fail activates.
Overload Protection
Fully protected against short circuit and output overload. Short
circuit protection is cycling type power limit.
Output Noise
0.5% rms, 1% pk-pk, 20 MHz Bandwidth, differential mode.
Measured with noise probe directly across output terminals of
the power supply.
Transient Response
500 µs typical response time for return to within 0.5% of final
value for a 45% load step from any load greater than 10% of full
rated load,
∆i/∆t<
0.2 A µs. Maximum voltage deviation is 3.5%.
Remote Sense
Provided as a standard feature on V1 and V2.
Overvoltage Protection
Built in on V1, 2 and 3 outputs.
Voltage Adjustment
Factory set to specified voltage with user adjustable potentiom-
eter on V1, 2 and 3. User can adjust V1, 2 and 3 at least ±5% of
nominal output voltage.
Output Regulation
Regulation for all outputs is the maximum deviation from initial
set point under all line and load conditions. Initial set tolerance is
measured with all outputs at 50 % of full rated load.
No Load Turn-On/Standby
A minimum load of 3 A on V1 is required for proper regulation. If
not met, no degradation of reliability will occur.
Temperature Coefficient
0.03% / °C typical on all outputs.
Input Protection
Internal ac fuses provided on both lines on all units.
Inrush Current
Inrush is limited by internal thermistors. The inrush at 240 Vac,
averaged over the first ac half-cycle under cold start conditions
will not exceed 37 A.
Thermal Shutdown
Provided as a standard feature. Designed to protect unit from
prolonged over temperature.
Power Fail
TTL / CMOS compatible output goes low (<0.5 V) 10 ms before
output voltage drops more than 4% below nominal voltage upon
loss of ac power.
Power Good
TTL / CMOS compatible output rises high 100 to 300 ms after V1
reaches regulation and should assure that sufficient energy is
stored in the input section to provide normal power fail /
shutdown.
Inhibit
Inhibit signal when pulled to the V1 output common will inhibit all
output voltages.
Fan Output
An additional thermally controlled 12 Vdc, 250 mA output suitable
for powering a dc fan is included in all models (Factory installed
cover with fan makes this output unavailable).
EMI/EMC Compliance
All models include built-in EMI filtering to meet the following
requirements:
EMC SPECIFICATION
COMPLIANCE LEVEL
Conducted Emissions
EN55011, Class B; FCC Class B
Static Discharge
EN61000-4-2, 6 kV contact 8 kV air
RF Field Susceptibility
EN61000-4-3, 3V/meter
Fast Transients / Bursts
EN61000-4-4, 2 kV, 5 kHz
Surge Susceptibility
EN61000-4-5, 1 kV diff., 2 kV com.
Conducted RF Susceptibility EN61000-4-6, 3V
Voltage Sags & Surges
EN61000-4-11
Line Frequency Harmonics
EN61000-3-2 Class A
EMC Compliance for Gold Series
Magnetic Emissions - Emissions will not exceed the limits of the
Army curve in MIL-STD-461D, RE101 when measured at 7 cm
from 30 Hz - 100 kHz over all nominal inputs and at full rated load.
Leakage Current
70 µA under normal conditions 120 Vac @ 60 Hz
Single fault conditions 280 µA, 264 Vac @ 50 Hz.
SAFETY AGENCY APPROVALS:
All models are Certified to be in compliance with the
applicable requirements of UL2601-1, CSA-C22.2 No. 601.1,
IEC 601-1/60601-1.
DESIGN VERIFICATION DOCUMENTS
The “Gold” series has undergone rigorous review and design
analysis. The following product documentation is available upon
request: 1. Failure Mode and Effects analysis (FMEA), 2. DVT
Data, 3. EMC / Susceptibility test results.
58
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
GLD140 Medical Switchers
140 Watt Multiple Output
Medical
Model
GLD140A
Output
N o.
1
2
3
4
1
2
3
4
1
2
3
4
1
2
3
4
1
2
3
4
1
2
3
4
1
2
3
4
Output
Vo l t a g e
+5.0 V
+12 V
12 V (C)
-12 V
+5.0 V
+12 V
5 V (C)
-12 V
+5.0 V
+12 V
15 V (C)
-15 V
+5.0 V
+24 V
12 V (C)
-12 V
+5.0 V
+24 V
15 V (C)
-15 V
+5.0 V
+15 V
12V (C)
-12 V
+5.0 V
+3.3 V
12 V (C)
-12 V
Output
Current (A)
16 A
6A
4A
0.75 A
16 A
6A
4A
0.75
16 A
6A
4A
0.75 A
16 A
4A
4A
0.75
16 A
4A
4A
0.75 A
16 A
5A
4A
0.75 A
16 A
10 A
4A
0.75 A
Output
Current (B)
20 A
8A
5A
1.2 A
20 A
8A
5A
1.2 A
20 A
8A
5A
1.2 A
20 A
6A
5A
1.2 A
20 A
6A
5A
1.2 A
20 A
7A
5A
1.2 A
20 A
12 A
5A
1.2 A
Set
To l e r a n c e
1%
1%
1%
2.2%
1%
1%
1%
2.2%
1%
1%
1%
2.2%
1%
1%
1%
2.2%
1%
1%
1%
2.2%
1%
1%
1%
2.2%
1%
1%
1%
2.2%
Line
Load
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
OV P
Tr i p
6.2 ± 0.3 V
14 ± 0.7 V
14 ± 1.2 V
-------
6.2 ± 0.3 V
14 ± 0.7 V
6.3 ± .5 V
-------
6.2 ± 0.3 V
14 ± 0.7 V
18 ± 1.2 V
-------
6.2 ± 0.3 V
27 ± 1 V
14 ± 1.2 V
-------
6.2 ±0.3 V
27 ± 1 V
18 ± 1.2 V
-------
6.2 ± 0.3 V
18 ± 1.2 V
14 ± 1.2 V
-------
6.2 ± 0.3 V
4.2 ± 0.4 V
14 ± 1.2 V
-------
Ripple
and Noise
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
1%
GLD140B
GLD140C
GLD140D
GLD140E
GLD140F
GLD140G
Output Current for Individual Outputs:
A. Output current for unrestricted natural convection.
B. Output current with 150 LFM forced air convection or peak current rating.
C. Isolated (floating) output may be referenced positive or negative.
GLD140
MECHANICAL SPECIFICATIONS
INPUT: JI
MOLEX P.C.B. HEADER P/N 39-30-2056
PIN 1) AC GROUND
PIN 2) N/C
PIN 3) AC NEUTRAL
PIN 4) N/C
PIN 5) AC LINE
MATING CONNECTOR MOLEX P/N:
HOUSING 39-01-4051, CONTACT 39-00-0164
SIGNALS: J2
AMP P
.C.B. HEADER P/N: 641215-6
PIN 1) +SENSE
PIN 2) -SENSE
PIN 3) POWER FAIL
PIN 4) POWER GOOD
PIN 5) COMMON
PIN 6) INHIBIT
MATING CONNECTOR AMP P/N:
HOUSING 770602-6, CONTACT 770666-2
OUTPUT: J3
MOLEX P
.C.B. HEADER P/N: 39-29-9206
PINS 1) +V3out
PINS 2) -V4out
PINS 3,12,13) COMMON
PINS 4) +V2 SENSE
PINS 5,14,15) +V2out
PINS 6-8, 16) COMMON
PINS 9,10,19,20) +V1out
PINS 11) +V3 RTN
PINS 17,18) COMMON
MATING CONNECTOR MOLEX P/N:
HOUSING 39-01-2200, CONTACT 39-00-0164
FAN: J5
AMP P.C.B. HEADER P/N 641215-2
PINS 1) FAN RTN
PINS 2) +FAN
MATING CONNECTOR AMP P/N:
HOUSING 770602-2, CONTACT 770666-2
Environmental
Specification
Temperature (A)
Humidity (A)
Shock (B)
Altitude
Vibration (C)
Operating
0 to 50
o
C
0 to 95% R H
20 g
pk
-500 to 10,000 ft
1.5 g
rms
, 0.003 g
2
/Hz
Non-operating
-40 to +85
o
C
0 to 95% RH
40 g
pk
-500 to 40,000 ft
5 g
rms
, 0.026 g
2
/Hz
A. Units should be allowed to warm up/operate under non-condensing conditions
before application of power.
B. Random vibration—10 to 2000Hz, 6dB/octave roll-off from 350 to 2000Hz, 3
orthogonal axes. Tested for 10 min./axis operating and 1 hr./axis non-operating.
C. Shock testing—half-sinusoidal, 10 ± 3 ms duration, ± direction, 3 orthogonal axes,
total 6 shocks.
Condor D.C. Power Supplies, Inc., 2311 Statham Parkway, Oxnard, CA 93033
800-235-5929 • 805-486-4565 • FAX 805-487-8911 • www.condorpower.com
59