The AD592 is a two terminal monolithic integrated circuit tem-
perature transducer that provides an output current propor-
tional to absolute temperature. For a wide range of supply
voltages the transducer acts as a high impedance temperature
dependent current source of 1
µA/K.
Improved design and laser
wafer trimming of the IC’s thin film resistors allows the AD592
to achieve absolute accuracy levels and nonlinearity errors previ-
ously unattainable at a comparable price.
The AD592 can be employed in applications between –25°C
and +105°C where conventional temperature sensors (i.e., ther-
mistor, RTD, thermocouple, diode) are currently being used.
The inherent low cost of a monolithic integrated circuit in a
plastic package, combined with a low total parts count in any
given application, make the AD592 the most cost effective tem-
perature transducer currently available. Expensive linearization
circuitry, precision voltage references, bridge components, resis-
tance measuring circuitry and cold junction compensation are
not required with the AD592.
Typical application areas include: appliance temperature sens-
ing, automotive temperature measurement and control, HVAC
(heating/ventilating/air conditioning) system monitoring, indus-
trial temperature control, thermocouple cold junction compen-
sation, board-level electronics temperature diagnostics,
temperature readout options in instrumentation, and tempera-
ture correction circuitry for precision electronics. Particularly
useful in remote sensing applications, the AD592 is immune to
voltage drops and voltage noise over long lines due to its high
impedance current output. AD592s can easily be multiplexed;
the signal current can be switched by a CMOS multiplexer or
the supply voltage can be enabled with a tri-state logic gate.
The AD592 is available in three performance grades: the
AD592AN, AD592BN and AD592CN. All devices are pack-
aged in a plastic TO-92 case rated from –45°C to +125°C. Per-
formance is specified from –25°C to +105°C. AD592 chips are
also available, contact the factory for details.
1. With a single supply (4 V to 30 V) the AD592 offers
0.5°C temperature measurement accuracy.
2. A wide operating temperature range (–25°C to +105°C)
and highly linear output make the AD592 an ideal sub-
stitute for older, more limited sensor technologies (i.e.,
thermistors, RTDs, diodes, thermocouples).
3. The AD592 is electrically rugged; supply irregularities
and variations or reverse voltages up to 20 V will not
damage the device.
4. Because the AD592 is a temperature dependent current
source, it is immune to voltage noise pickup and IR
drops in the signal leads when used remotely.
5. The high output impedance of the AD592 provides
greater than 0.5°C/V rejection of supply voltage drift and
ripple.
6. Laser wafer trimming and temperature testing insures
that AD592 units are easily interchangeable.
7. Initial system accuracy will not degrade significantly over
time. The AD592 has proven long term performance
and repeatability advantages inherent in integrated cir-
cuit design and construction.
378
343
I
OUT
– µA
1µA/
o
K
273
248
–45
–25
0
+70
+105 +125
TEMPERATURE –
o
C
REV.
B
Information furnished by Analog Devices is believed to be accurate and
reliable. However, no responsibility is assumed by Analog Devices for its
use, nor for any infringements of patents or other rights of third parties
which may result from its use. No license is granted by implication or
otherwise under any patent or patent rights of Analog Devices.
One Technology Way, P.O. Box 9106, Norwood, MA 02062-9106, U.S.A.
Tel:
781/329-4700
Fax:
781/461-3113
AD592* PRODUCT PAGE QUICK LINKS
Last Content Update: 02/23/2017
COMPARABLE PARTS
View a parametric search of comparable parts.
DISCUSSIONS
View all AD592 EngineerZone Discussions.
DOCUMENTATION
Application Notes
•
AN-892: Temperature Measurement Theory and Practical
Techniques
Data Sheet
•
AD592: Low Cost, Precision IC Temperature Transducer
Data Sheet
SAMPLE AND BUY
Visit the product page to see pricing options.
TECHNICAL SUPPORT
Submit a technical question or find your regional support
number.
DOCUMENT FEEDBACK
DESIGN RESOURCES
•
AD592 Material Declaration
•
PCN-PDN Information
•
Quality And Reliability
•
Symbols and Footprints
Submit feedback for this data sheet.
This page is dynamically generated by Analog Devices, Inc., and inserted into this data sheet. A dynamic change to the content on this page will not
trigger a change to either the revision number or the content of the product data sheet. This dynamic page may be frequently modified.
AD592–SPECIFICATIONS
(typical @ T = +25 C, V = +5 V, unless otherwise noted)
A
S
Model
ACCURACY
Calibration Error @ +25°C
1
T
A
= 0°C to +70°C
Error over Temperature
Nonlinearity
2
T
A
= –25°C to +105°C
Error over Temperature
3
Nonlinearity
2
OUTPUT CHARACTERISTICS
Nominal Current Output
@ +25°C (298.2K)
Temperature Coefficient
Repeatability
4
Long Term Stability
5
ABSOLUTE MAXIMUM RATINGS
Operating Temperature
Package Temperature
6
Forward Voltage (+ to –)
Reverse Voltage (– to +)
Lead Temperature
(Soldering 10 sec)
POWER SUPPLY
Operating Voltage Range
Power Supply Rejection
+4 V < V
S
< +5 V
+5 V < V
S
< +15 V
+15 V < V
S
< +30 V
AD592AN
Min Typ Max
1.5
1.8
0.15
2.0
0.25
2.5
3.0
0.35
3.5
0.5
AD592BN
Min Typ Max
0.7
0.8
0.1
0.9
0.2
1.0
1.5
0.25
2.0
0.4
AD592CN
Min Typ Max
0.3
0.4
0.05
0.5
0.1
0.5
0.8
0.15
1.0
0.35
Units
°C
°C
°C
°C
°C
298.2
1
0.1
0.1
–25
–45
+105
+125
44
20
300
4
30
0.5
0.2
0.1
4
–25
–45
298.2
1
0.1
0.1
+105
+125
44
20
300
30
0.5
0.2
0.1
4
–25
–45
298.2
1
0.1
0.1
+105
+125
44
20
300
30
0.5
0.2
0.1
µA
µA/°C
°C
°C/month
°C
°C
V
V
°C
V
°C/V
°C/V
°C/V
NOTES
1
An external calibration trim can be used to zero the error @ +25°C.
2
Defined as the maximum deviation from a mathematically best fit line.
3
Parameter tested on all production units at +105°C only. C grade at –25°C also.
4
Maximum deviation between +25°C readings after a temperature cycle between –45°C and +125°C. Errors of this type are noncumulative.
5
Operation @ +125°C, error over time is noncumulative.
6
Although performance is not specified beyond the operating temperature range, temperature excursions within the package temperature range will not damage the device.
Specifications subject to change without notice.
Specifications shown in boldface are tested on all production units at final electrical test. Results from those tests are used to calculate outgoing quality levels. All min
and max specifications are guaranteed, although only those shown in boldface are tested on all production units.
METALIZATION DIAGRAM
66MILS
V+
TEMPERATURE SCALE CONVERSION EQUATIONS
42MILS
V–
C =
5
( F –32)
9
K =
°C
+273.15
°R
=
°F
+459.7
F =
9
C +32
5
–2–
REV.
B
Typical Performance Curves–AD592
Typical @ V
S
= +5 V
+2.0
+1.5
+1.0
+2.0
+1.5
+1.0
TOTAL ERROR –
o
C
+0.5
0
–0.5
–1.0
–1.5
–2.0
–25
0
+25
TEMPERATURE –
o
C
+70
+105
TOTAL ERROR –
o
C
+0.5
0
–0.5
–1.0
–1.5
–2.0
–25
0
+25
TEMPERATURE –
o
C
+70
+105
AD592CN Accuracy Over Temperature
AD592BN Accuracy Over Temperature
+2.0
+1.5
+1.0
TOTAL ERROR –
o
C
0.75
0.50
+0.5
0
–0.5
–1.0
TOTAL ERROR –
o
C
0.25
0
–0.25
–0.50
–1.5
–2.0
–0.75
0
500
1000
TIME – Hours
1500
2000
–25
0
+25
+70
+105
TEMPERATURE –
o
C
AD592AN Accuracy Over Temperature
Long-Term Stability @ +85
°
C and 85% Relative Humidity
0.75
0.50
TOTAL ERROR –
o
C
0.25
0
–0.25
–0.50
–0.75
0
500
1000
TIME – Hours
1500
2000
Long-Term Stability @ +125
°
C
REV.
B
–3–
AD592
THEORY OF OPERATION
TOTAL ERROR –
o
C
The AD592 uses a fundamental property of silicon transistors
to realize its temperature proportional output. If two identical
transistors are operated at a constant ratio of collector current
densities, r, then the difference in base-emitter voltages will be
(kT/q)(ln r). Since both k, Boltzman’s constant and q, the
charge of an electron are constant, the resulting voltage is
directly Proportional To Absolute Temperature (PTAT). In the
AD592 this difference voltage is converted to a PTAT current
by low temperature coefficient thin film resistors. This PTAT
current is then used to force the total output current to be pro-
portional to degrees Kelvin. The result is a current source with
an output equal to a scale factor times the temperature (K) of
the sensor. A typical V-I plot of the circuit at +25°C and the
temperature extremes is shown in Figure 1.
resistor. Note that the maximum error at room temperature,
over the commercial IC temperature range, or an extended
range including the boiling point of water, can be directly read
from the specifications table. All three error limits are a combi-
nation of initial error, scale factor variation and nonlinearity de-
viation from the ideal 1
µA/K
output. Figure 2 graphically
depicts the guaranteed limits of accuracy for an AD592CN.
+1.0
+0.5
MAXIMUM ERROR
OVER TEMPERATURE
TYPICAL ERROR
0
CALIBRATION
ERROR LIMIT
–0.5
–1.0
MAXIMUM ERROR
OVER TEMPERATURE
378
I
OUT
– µA
+105
o
C
+25
o
C
–25
o
C
UP TO
30V
–25
0
+25
TEMPERATURE –
o
C
+70
+105
298
248
Figure 2. Error Specifications (AD592CN)
0
1
2
3
4
SUPPLY VOLTAGE – Volts
5
6
Figure 1. V-I Characteristics
NONLINEARITY –
o
C
Factory trimming of the scale factor to 1
µA/K
is accomplished
at the wafer level by adjusting the AD592’s temperature reading
so it corresponds to the actual temperature. During laser trim-
ming the IC is at a temperature within a few degrees of 25°C
and is powered by a 5 V supply. The device is then packaged
and automatically temperature tested to specification.
FACTORS AFFECTING AD592 SYSTEM PRECISION
The AD592 has a highly linear output in comparison to older
technology sensors (i.e., thermistors, RTDs and thermo-
couples), thus a nonlinearity error specification is separated
from the absolute accuracy given over temperature. As a maxi-
mum deviation from a best-fit straight line this specification rep-
resents the only error which cannot be trimmed out. Figure 3 is
a plot of typical AD592CN nonlinearity over the full rated tem-
perature range.
+0.2
+0.1
TYPICAL NONLINEARITY
0
The accuracy limits given on the Specifications page for the
AD592 make it easy to apply in a variety of diverse applications.
To calculate a total error budget in a given system it is impor-
tant to correctly interpret the accuracy specifications, non-
linearity errors, the response of the circuit to supply voltage
variations and the effect of the surrounding thermal environ-
ment. As with other electronic designs external component se-
lection will have a major effect on accuracy.
CALIBRATION ERROR, ABSOLUTE ACCURACY AND
NONLINEARITY SPECIFICATIONS
–0.1
–0.2
–25
0
+25
TEMPERATURE –
o
C
+70
+105
Figure 3. Nonlinearity Error (AD592CN)
TRIMMING FOR HIGHER ACCURACY
Three primary limits of error are given for the AD592 such that
the correct grade for any given application can easily be chosen
for the overall level of accuracy required. They are the calibra-
tion accuracy at +25°C, and the error over temperature from
0°C to +70°C and –25°C to +105°C. These specifications cor-
respond to the actual error the user would see if the current out-
put of an AD592 were converted to a voltage with a precision
Calibration error at 25°C can be removed with a single tempera-
ture trim. Figure 4 shows how to adjust the AD592’s scale fac-
The value of multi-channel conversion is about 0.2V smaller than that of single-channel conversion, and the more channels there are, the smaller the value is. When converting four channels, the value ...
I have a question for you seniors. Does the wiring on the PCB board have to be done according to the connection of the flying wire? Can I connect it according to my own ideas instead of following the ...
The life of a power supply often depends on the life of an electrolytic capacitor, but what changes occur in the characteristics of the capacitor as the electrolytic capacitor ages? Please feel free t...
In some cases, the application code may suddenly run away or generate an abnormal interrupt, and the cause of the problem may be difficult to find. When debugging such problems, you can consider using...
The most significant event in the technology circle during this period is undoubtedly the news that ZTE was banned by the United States. Once the news came out, it shocked the entire Chinese techno...[Details]
Guangdong Traffic Regulation No. 355
To all municipal transportation bureaus (commissions) above the prefecture level, provincial transportation groups, Guangdong Power Grid Compa...[Details]
Recently, the Provincial Development and Reform Commission and the Provincial Poverty Alleviation Office jointly organized the application for a number of centralized photovoltaic poverty alleviation ...[Details]
In the war without gunpowder between China and the United States,
ZTE
has become a new victim. The seven-year chip ban by the United States has pushed this communications giant to a life-or...[Details]
I recently upgraded MDK to V5.25 and found that there are many libraries in the Managing Run-Time Environment. Compared with the past, it is much better, from the bottom-level driver to the upper-le...[Details]
0. Preparation First use st-link to connect the stm32 core board and PC for burning St-link Stm32 3.3V 3.3V GND GND SWDIO PART SWCLK DCLK Then connect with USB serial port board to v...[Details]
introduction STM32 has many registers, which are difficult to remember, so the official packaged two sets of library functions. One is the standard library, but the official has not updated it on F7,...[Details]
1. Determination of STM32 ADC sampling frequency 1. First, read some information to determine the ADC clock: (1) The ADCCLK clock provided by the clock controller is synchronized with the PCLK2 (A...[Details]
According to MEMS Consulting, Japan's
Toshiba
will further develop its automotive
LiDAR
(laser radar) semiconductor business, mainly focusing on: (1)
light receiving elements use...[Details]
Artificial intelligence
There is no inherent bias in AI. It does not "think" something is true or false for reasons that cannot be explained by logic. Unfortunately, human bias exists in mach...[Details]
China Energy Storage Network News:
The State Grid Corporation of China, this elephant, is planning to turn around and enter the emerging power service market.
The rallying cry for the tr...[Details]
There has been a lot of news about Huawei
phones
recently
. At the
Huawei
Global Analyst Conference
on the 17th
,
Huawei
announced that it will launch
5G
phones
in the seco...[Details]
I am going to do a project recently, which requires counting the frequency and duty cycle of the external PWM signal. The chip used is STM32F207 Looking at the datasheet, I found that one of the ma...[Details]
The watchdog timer is used to prevent the program from running away due to strong interference noise caused by power supply, space electromagnetic interference or other reasons. Many microcontrollers...[Details]
*This article is based on the author's report at the "Embedded System Association Theme Discussion (Total 22nd) - Current Situation and Development Prospects of Internet of Things Operating Systems...[Details]